@stefbeer,
Good spot. I totally missed that important specification detail ! Or more accurately I misread it as a statement that the +-2.0C applied at above 0C ! To be truthful I wasn't buying for amazing accuracy so it wasn't of that much interest.
It makes it very clear that the camera has a specified accuracy only above 0C. FLIR do not calibrate the E4 below 0C, as others have suggested.
My industrial Raytek IR thermometer actually refuses to measure any temperature below 0C as it would not be accurate at such temperatures.
From the lack of any accuracy detail below 0C in the specs of the E4, it would appear that any temperature shown is basically a 'best guess' and not to be trusted.
I have also been looking into the behaviour of PVC electrical tape and that is also a source of error in my temperature test. PVC tape changes in emissivity as its temperature changes ! Life is never simple. A true clack body is needed to carry out serious testing of the E4. Sadly I no longer have access to one at work. I may do some more testing on accuracy when time permits.
My FLIR PM695 manual suggests measuring a target surface with a thermocouple and and then adjusting the PM695 emissivity setting whilst looking at the same spot with the camera. this accurately sets the emissivity on that specific surface.
The E4 is still very useful to me as I tend to work with targets at temperatures above freezing
.
I will use my accurate, if somewhat cumbersome, PM695's for any work requiring good accuracy between -40C and +500C.