@zucca,
Regarding your previous post on my tests using a PM695 as a comparison of accuracy......
The PM695 was within spec during the +7 Degrees C test and the -8 Degrees C test. Readings were +7.3 Degrees C and -8.3 Degrees C. The Camera is rated at +- 2% between -40 and +500 Degrees C. No +-2 Degrees shown as found in the E4 spec.
The E4 was accurate during the + 7 Degrees C test with a reading of +7
BUT when used to read the -8 Degrees C test piece the E4 read
-11.6 degrees C. That is an error of over 30% on a camera that claims +-2% or +-2 Degrees C down to -20 Degrees C.
This phenomena appears to be confirmed by others experience with the camera. Personally I have no problem with this as I do not intend using it as an accurate thermal camera at low temperatures
As has been stated, this is a budget camera that may not even be calibrated at sub zero temperatures. My PM695 can accurately read very low temperatures down to -40 Degrees C because it is an industrial preventative maintenance camera designed to be capable of such ...... but it cost over $50,000 so you might expect better performance in some areas when compared to a $1000 E4
I will do some more E4 temperature accuracy tests when I find time....Christmas is rushing up on me and much to do.
Manual including specs for the PM695 is here.
http://www.flir.com/uploadedFiles/Legacy_-_NEW/Thermography/ThermaCAMPM695.pdfThe PM695 manual contains a useful emissivity guide for many materials which may be used to set up your E4 emissivity option.The manual contains no warnings regarding accuracy loss at sub zero C temperatures.
If there is an issue here, it is that FLIR suggest accurate use down to -20 Degrees C. I have yet to be convinced that such is true. Testing at these low temperatures is challenging as 'frost' quickly forms on the test piece. This ruins the accuracy of the test.