Moving on, next test point is the primary winding of the isolation transformer.
The unit isolates its signals by the means of transformers, as one can see from the block diagram, too. There's three transformers per channel: One does the supply job, next transfers the carrier frequency from the digital circuitry to the front end amplifiers and the third transfers the modulated analog signal to the ADC circuits.
So I'm tapping the output transformer, using two channels of the oscilloscope as a poor mans differential probe. Works nicely here.
On the scope screen the modulated signal, it's a very simple scheme: the signal gets inverted with each half period of the carrier. So there's no DC across the transformer, and demodulation is easily done by an synchronous demodulator.
This is just one screenshot, I did one at each input range and input mode for comparison: This signal is absolutely as expected, switching gain through the ranges, no gain error, no offset error.
Conclusion: The isolated side of the current measurement path works fine, both in BNC and shunt measurement mode.
Still there's some large errors (gain and offset) displayed at the instruments screen when switching to shunt mode. What might be wrong now?
I suspect, there's no difference in the analog signal path on the non-isolated side between BNC and shunt mode. One thing left: Digital calibration constants - without knowing how to access these or how to adjust (calibrate) the instrument I'm at a road block now.
For the brave, one might check the ADC input once more, I'll see if I can find a suitable tap.
Edit: Confirmed, there's a perfect signal at the ADC input, no offset or gain issues. So it's a matter of adjusting / calibrating the unit. I couldn't find any calibration menu, nor description of how to do this through the GPIB, so I'm at the end of this journey for now.
Anyway, depending on the method to calibrate the unit, it can be quite challenging to provide the necessary signals.