If you only want R&S and NEED real 50 OHm termination in the scope, then you have to buy RTM3000 or better.
It also have something that resembles SA. But that is 2-3 X your budget
But you are going about this in a wrong way.
If you need RF, you buy SPECTRUM ANALYSER and a simpler scope.
For 2500 € you can get Siglent SVA1015X (and get VNA and SA ) and SDS1104X-E or MSO5000 at the same time.
Or get Siglent SSA3021X Spectrum Analyzer and one the mentioned scopes..
Both combinations will give you 10x better capabillites for RF that even 12000€ RTM3000 with 1GHz bandwidth..
I'm a bit torn between advising getting the spectrum analyser or the VNA. Both are different beasts. While a VNA can be used as a spectrum analyser it is not as good at spectrum analysis compared to a spectrum analyser. A VNA is great if you are into circuit design because it can tell you (by showing the impedances and phase shift) how to fix an RF circuit where a spectrum analyser can only show you something isn't right. OTOH a spectrum analyser allows to dive deep into the frequency spectrum of a signal.
Nico,
that is why I mentioned both. I did try SVA1015X a bit, and it is not great SA, but it is OK. But 1.5GHz VNA is useful.
Like always it really depends what you do. If you do testing of ready made equipment, SA is important. If you're making things, characterising components, cables, connectors... VNA is very important.
Ideally you need both. And I can tell you that even entry level VNA is very useful.
It is like comparison between simple handheld meter and a benchtop 6.5 digit one. There is no question which is better and sometimes you just need all that accuracy.
But even a small handheld will be better than putting a finger on a battery to try to estimate state of charge...