Hi,
The attached shows a burst mode transient voltage applied to our mains connected equipment.
This comes from a Schaffner BEST  burst mode tester.
I set it to 2kV Peak...but as can be seen, its only 1600Vpk. (the zero ref for the "yellow"
channel is at -1600Vdc).
Why did it not get up to 2kV?
I measured it with a 100:1 probe connected from true mains earth (not the earth coming out of the Schaffner)
and Line. (Line that  comes out of the Schaffner)
The Schaffner was set up as follows....
_2kv peak
_No of burst pulses = 1
_Phase angle of burst = 90 degrees
_Transient coupled to Line only.

So why did it only get up to 1600V?
Was it due to the MOV etc?
« Last Edit: July 27, 2024, 09:29:58 pm by Faringdon »
'Perfection' is the enemy of 'perfectly satisfactory'
 

Offline joeqsmith

  • Super Contributor
  • ***
  • Posts: 12060
  • Country: us
That's some beginner stuff there.  The generators are spec'ed for open circuit voltage and short circuit current waveforms (combo).    They have a specified source impedance as well.  Drive it into a dead short and the voltage will be less.   

You really would be better off renting some time at a lab. 

Offline FaringdonTopic starter

  • Super Contributor
  • ***
  • Posts: 2118
  • Country: gb
Quote
The generators are spec'ed for open circuit voltage and short circuit current waveforms (combo).
Thanks,  say we have a Boost PFC'd power supply with 800uF of electrolytics after the booster. Then i take it that we can set 2kV, but will get way, way under that.
As you kindky discuss, it will clamp at the current limit. I must admit, i didnt see from the manual what that was. I think those pages are missing.
(theres no manual online for the Schaffner BEST)
Thanks but We cant afford to pay for lab time.
'Perfection' is the enemy of 'perfectly satisfactory'
 

Offline joeqsmith

  • Super Contributor
  • ***
  • Posts: 12060
  • Country: us
Quote
(theres no manual online for the Schaffner BEST)

As you stated in your previous thread:
https://www.eevblog.com/forum/testgear/schaffner-best-transient-tester/

Quote
Thanks but We cant afford to pay for lab time.
Can you afford not to?

I would just run it open circuit, and then with some resistive loads until you at least understand some of the basics.  With it being so old an unsupported, are you sure it even meets the standards you are required to adhere to?  Or is this more just for fun and you really are not trying to meet any of the current standards?

Offline floobydust

  • Super Contributor
  • ***
  • Posts: 7237
  • Country: ca
I am familiar with Teseq ones, super easy to use so techs can do the testing. It's a very high cost of ownership with commercial mains transient generators.
A new standard rolls out, or an update to a standard and you have to send it in for hardware mods- or the unit just goes obsolete. Steady churn of these units on the surplus market.
Calibration is a constant cash drain, for test-lab requirements. New, they are the price of a car.
That's why I wanted joeq to give details about the insides of his homebrew one ;)

If output voltage is low, make sure the unit is capable of that in the first place. It could be for older (standards).
It could need repairs, have some damage and only be making 1.6kV not 2kV, or have high output resistance. Or there is an arc in the DUT.
I'd test it with a simple resisitive load to make sure it's working properly.

Most important is safe test leads and the blast shield, on an insulated bench. You have to know what you are doing. We wouldn't want a dead Treez.
 

Offline Gyro

  • Super Contributor
  • ***
  • Posts: 9760
  • Country: gb
... We wouldn't want a dead Treez.

A nasty fright to make you actually listen and think for yourself however.  >:D
Best Regards, Chris
 

Offline floobydust

  • Super Contributor
  • ***
  • Posts: 7237
  • Country: ca
... We wouldn't want a dead Treez.

A nasty fright to make you actually listen and think for yourself however.  >:D

If you buy used test equipment and the safety hardware is missing, it can accelerate Darwin's process if self-reflection and intellect is also absent from the kit.
Icarus mentioned in the manual's missing pages. He thought he was pretty good at SMPS design, a bit cheap on the bee's wax though.
 

Offline joeqsmith

  • Super Contributor
  • ***
  • Posts: 12060
  • Country: us
Calibration is a constant cash drain, for test-lab requirements. New, they are the price of a car.
That's why I wanted joeq to give details about the insides of his homebrew one ;)

Sorry but I don't follow. People here like automotive analogies.  If a new McLaren was a quarter mil and I show a home made wind up toy, how does showing how the wind up toy was designed help?   The design constraints between a generator designed for the IEC standard and what I came up with to test multimeters is completely different.   I've been asked about transient testing AC mains products.  It shows just how far off some of the thoughts are about what was shown.  It doesn't seem to matter how much I try to educate viewers, people see and hear what they want.  In this case, OP I assume is wanting to test using an actual combo generator to some IEC standard.

 
Quote
Most important is safe test leads and the blast shield, on an insulated bench.
For mains products like the OP is asking about, the leads may very well be the supplied line cord.  I would imagine in many cases, there is no protective shield used.   Do use some sort of shield where you work?  If so, and you are able to discuss, may be worth a new topic. 



Offline floobydust

  • Super Contributor
  • ***
  • Posts: 7237
  • Country: ca
Calibration is a constant cash drain, for test-lab requirements. New, they are the price of a car.
That's why I wanted joeq to give details about the insides of his homebrew one ;)

Sorry but I don't follow. People here like automotive analogies.  If a new McLaren was a quarter mil and I show a home made wind up toy, how does showing how the wind up toy was designed help?   The design constraints between a generator designed for the IEC standard and what I came up with to test multimeters is completely different.   I've been asked about transient testing AC mains products.  It shows just how far off some of the thoughts are about what was shown.  It doesn't seem to matter how much I try to educate viewers, people see and hear what they want.  In this case, OP I assume is wanting to test using an actual combo generator to some IEC standard.

I'd asked what component you used for the power switch in your transient generator, curious what could take that much voltage and current. Not for formal McLaren kind of building or scrutiny. It was simple wondering, towards that thread getting around using a relay.


Most important is safe test leads and the blast shield, on an insulated bench.
For mains products like the OP is asking about, the leads may very well be the supplied line cord.  I would imagine in many cases, there is no protective shield used.   Do use some sort of shield where you work?  If so, and you are able to discuss, may be worth a new topic.

It's a 1/4" thick plexiglass U-cover that keeps sparks, molten metal, component bodies from flying across the lab. It might have been made by another engineer, or is a Teseq accessory, not sure. TO-247 explodes and epoxy flies across the lab. Eye protection is good but getting beaned is still unpleasant.
When you have noobs and people walking around, pretty much mandatory in a corporate lab environment.

The cover I suppose is similar to the need for your multimeter test enclosure with the door and window, except this was sized for larger exposed PC boards under test.
 