Author Topic: AFG3101 reports errors  (Read 646 times)
Offline lern01Topic starter
  • Country: cn
« on: December 05, 2023, 01:49:45 pm »
Hi, all
My AFG3101 reports errors during diagnosis and self-calibration, the error codes are 2303 and 1103 respectively, that is  CH1 Output offset failure and CH1 Output offset calibration failure. When referring to the maintenance manual, I have doubts about the following maintenance tree. What I understand should be: No, Replace the Genertor board. Yes, Replace the Amplifier board.  Asking for your help. I hope you all can make suggestions for repairs. Thank you in advance!
Offline alm
The following users thanked this post: lern01
  • Country: cn
Re: AFG3101 reports errors
« Reply #3 on: April 14, 2024, 05:51:52 am »
I have fixed the error. It is a fault of the magnification board, and I have hand-drawn the circuit diagram of the amplifier board. If you need it, please contact me.
« Last Edit: April 17, 2024, 03:11:38 am by lern01 »