Author Topic: fluke PM2813 problem - please help me  (Read 4092 times)

0 Members and 1 Guest are viewing this topic.

Offline eevblogfanTopic starter

  • Frequent Contributor
  • **
  • Posts: 569
  • Country: 00
fluke PM2813 problem - please help me
« on: May 13, 2013, 08:52:33 pm »
hey

so as some of you know I've purchased the PM2813 , great PS in all means , but ther's some problem with it , the sensing is bad , let me explain

so I've got some binding posts for the 3 channels plus the sensing ports , I've drilled the front panel and all fit nicely , so I turned the PS on and connected one of the Channels to my DC load , set to 20V 3A , draw 2.99 ( 3A will turn the PS into the CC mode ) , and obviously ther's dropp of the wires , so obviously you connect the sensing leads right ?  I did so , but the voltage wasn't 20V . without the Sensing it was say 19.7V and with the sensing it was 19.8V , so what's wrong ? well ther's shorting bars on the rear connectors between the + and S+ and the - and S- , so I've disconnected them , and turned the PS on , BUT there was written " CH1 FAIL " , WTF ? 

well I thought there is no internal 100ohm resistors between the sensing and output , so I put some , but the "FAIL" is still there ( Yes I have restarted the PS )

so please , can someone can spot me the problem  ?

thank you in advance ! :)
 

Offline eevblogfanTopic starter

  • Frequent Contributor
  • **
  • Posts: 569
  • Country: 00
Re: fluke PM2813 problem - please help me
« Reply #1 on: May 14, 2013, 11:24:21 am »
someone ?  :scared:
 

Offline dr_p

  • Frequent Contributor
  • **
  • Posts: 309
  • Country: ro
Re: fluke PM2813 problem - please help me
« Reply #2 on: May 14, 2013, 12:40:27 pm »
is it set to local sense?
 

Offline eevblogfanTopic starter

  • Frequent Contributor
  • **
  • Posts: 569
  • Country: 00
Re: fluke PM2813 problem - please help me
« Reply #3 on: May 14, 2013, 12:44:22 pm »
hey

firstly _ thank you :P

so if it does set to loacl . how shall I set it to remote ?

if you mean the jumpers inside , I have already turn them to the "on" position :(
 

Offline PA4TIM

  • Super Contributor
  • ***
  • Posts: 1164
  • Country: nl
  • instruments are like rabbits, they multiply fast
    • PA4TIMs shelter for orphan measurement stuff
Re: fluke PM2813 problem - please help me
« Reply #4 on: May 14, 2013, 05:24:01 pm »
I have a P2811, the same psu but only one channel ( firmware seems to be equal)

Did you do the calibration ? Mine was not accurate before I did that. ( Voltage was off and -40 mA without sourcing current after following the manual, ( the resistor thing) then I did what looked more logical to me and used my Keithley 2000 to measure current and now it is spot on)

On my 2811 the 4 terminals on the back are connected to 4 bannana busses on the front ( original option in a sub chassis under the psu, I added a powerswitch there
)
There is a switch beside the bannanas that lets me choose between sense and no sense. It just couples the sense to source lines. There are no resistors. It gives an error if I set the switch wrong. ( and sense is not connected) looks like that is your problem. If I understand right you modified it from back connectors only to front connectors ?)

The sense is part of the feedback. It measures the voltage and regulates the output according. I know because that was the problem on mine. Without the sense it sees no output.

If I remember well during repair: If you put an external voltage on the sense lines you get that as a reading on the display. So the display seems to shows the real output and not the set output ( and that is a safe idea) So that could imply you should see a lower reading using sense and long wires. However I would expect it to regulate voltage again to setting by itself. But I have never checked that. Never used the sense. If you see the set value minus the voltdrop there could be a problem in the feedback. Is the output correct using a multimeter ?

I can try some tests on mine if you want.
http://www.pa4tim.nl/?p=3584 my PM2811, so you can see the switch

www.pa4tim.nl my collection measurement gear and experiments Also lots of info about network analyse
www.schneiderelectronicsrepair.nl  repair of test and calibration equipment
https://www.youtube.com/user/pa4tim my youtube channel
 

Offline eevblogfanTopic starter

  • Frequent Contributor
  • **
  • Posts: 569
  • Country: 00
Re: fluke PM2813 problem - please help me
« Reply #5 on: May 14, 2013, 06:14:08 pm »
hey

update :

problem solved ! , one power supply is solved , one more to go ! ( If someone follows me )

I mistakely connected the sens wrong ( output 3 to uutput 1 , and vice versa )  the rear terminals are opposite to the front , hence the mistake ,

I also connected 10ohm resistors from S+ to V+ and S- to V- , due to "fail" written when the sense are not shorted to the output terminals ,

now all fine , I've tested 6V at 10A , all works fine , but when using the sense ther's problem , the PS cannot output more then 1V above the set value ( in order to compensate for wier losses ) , so ther's need to use heafty banana connectors ( mine dropped 200mV just accross the banana connectors )  so I had one big banana plug and he did well there ,

the problem with that PS is when "FAIL" appears on the screen , you cant tell what has fail , you have to know by yourself , it is "smart" supply , why can't they write " eror XX" instead ? , so that on the menual you'll be able to se , "eror 41" will mean , too much drop on wiers , and such ,  :-+

mine HP34401 tells me " 0.021mVRMS" when I measure ACV , I am sure the output ripple is more like 1mVRMS or so ,  but does someone know what is the ripple under 10A ?

my scope tells me 21mVpp ( Ie rughly 10mVRMS ) , can someone confirm ? ( 20Mhz BW limit on )

thank you :)
 