Re: Frequency measurements with oscilloscope - what accuracy to expect?
« Reply #50 on: May 20, 2016, 10:10:41 pm »
Maybe you will find these helpful.

Yes, thank you. Much interesting on your YouTube channel!

I noticed you were using the Standard Deviation value for measuring the jitter. My datasheet (TUSB1310A TI USB 3.0 PHY) above states that if an external oscillator is used, its "absolute p-p" clock jitter should be less than 50 psec. This seems impossible since the p-p jitter is unbounded and the actual built-in oscillator measures around 300 ps p-p. Likely they mean "std dev"? And what is this business with "reference clock TJ with JTF (Jitter Transfer Function) total p-p" of 25ps? The note says calculated as 14.1 x RJ x DJ. That would require (RJ + TJ) to be < (25ps / 14.1) or less than 1.77ps  :palm: All this is like walking in a hall of mirrors...

I'm still reading in on the subject and I will look into the EZ-Jitter Plus software installed on my Agilent scope. But I really would like to understand this analytically.

On a brighter note: I figured out why my oscillator was some 1.5KHz too high. The external load capacitance was incorrectly calculated. When increasing to 33pF, which meets the crystal's total load capacitance of 18pF, the frequency is spot on (within 1PPM). However, it drifts some 200 Hz down when the board heats up from 25 degC to around 50 degC. This is expected since the XTAL used only has a +/- 50PPM temperature stability. I have ordered one with +/- 10 PPM on both frequency and stability over temp spec. Hopefully this will reduce the temp drift.

On a negative note: My PHY still doesn't lock. Next test would be to rework the board to disable the SSC to see if it makes a difference.
« Last Edit: May 20, 2016, 10:13:34 pm by John_ITIC »
Pocket-Sized USB 2.0 LS/FS/HS Protocol Analyzer Model 1480A with OTG decoding.
Pocket-sized PCI Express 1.1 Protocol Analyzer Model 2500A. 2.5 Gbps with x1, x2 and x4 lane widths.
https://www.internationaltestinstruments.com
 

Online tautech

  • Super Contributor
  • ***
  • Posts: 29476
  • Country: nz
  • Taupaki Technologies Ltd. Siglent Distributor NZ.
    • Taupaki Technologies Ltd.
Re: Frequency measurements with oscilloscope - what accuracy to expect?
« Reply #51 on: May 20, 2016, 10:31:44 pm »
On a brighter note: I figured out why my oscillator was some 1.5KHz too high. The external load capacitance was incorrectly calculated. When increasing to 33pF, which meets the crystal's total load capacitance of 18pF, the frequency is spot on (within 1PPM). However, it drifts some 200 Hz down when the board heats up from 25 degC to around 50 degC. This is expected since the XTAL used only has a +/- 50PPM temperature stability. I have ordered one with +/- 10 PPM on both frequency and stability over temp spec. Hopefully this will reduce the temp drift.
Are you using NPO caps?

Just checking.  ;)
Avid Rabid Hobbyist.
Some stuff seen @ Siglent HQ cannot be shared.
 

Offline John_ITICTopic starter

  • Frequent Contributor
  • **
  • Posts: 522
  • Country: us
  • ITIC Protocol Analyzers
    • International Test Instruments Corporation
Re: Frequency measurements with oscilloscope - what accuracy to expect?
« Reply #52 on: May 20, 2016, 11:58:55 pm »
On a brighter note: I figured out why my oscillator was some 1.5KHz too high. The external load capacitance was incorrectly calculated. When increasing to 33pF, which meets the crystal's total load capacitance of 18pF, the frequency is spot on (within 1PPM). However, it drifts some 200 Hz down when the board heats up from 25 degC to around 50 degC. This is expected since the XTAL used only has a +/- 50PPM temperature stability. I have ordered one with +/- 10 PPM on both frequency and stability over temp spec. Hopefully this will reduce the temp drift.
Are you using NPO caps?

Just checking.  ;)

Yes, I'm currently using these capacitors, which I had available in my lab. These have +/- 5% tolerance :
http://www.mouser.com/ProductDetail/Murata-Electronics/GRM1885C1H330JA01D/?qs=E%252b5dfF6xfG3shwmI3hm0tg%3D%3D

For my production boards, I'll likely use these, which are +/- 1%
http://www.digikey.com/product-search/en?keywords=490-10715-1-nd

Regardless, the frequency is now clearly well within the 50PPM specified so not much need to fine-tune further (although it is fun!).
Pocket-Sized USB 2.0 LS/FS/HS Protocol Analyzer Model 1480A with OTG decoding.
Pocket-sized PCI Express 1.1 Protocol Analyzer Model 2500A. 2.5 Gbps with x1, x2 and x4 lane widths.
https://www.internationaltestinstruments.com
 
The following users thanked this post: tautech

Offline John_ITICTopic starter

  • Frequent Contributor
  • **
  • Posts: 522
  • Country: us
  • ITIC Protocol Analyzers
    • International Test Instruments Corporation
Re: Frequency measurements with oscilloscope - what accuracy to expect?
« Reply #53 on: May 27, 2016, 01:12:30 am »
Time Interval Error Jitter Histogram gives vastly different result depending on ns/div setting.

I'm trying to measure oscillator clock jitter with my DSO81304B. The datasheet for the IC (http://www.ti.com/lit/ds/symlink/tusb1310a.pdf ) states on page 32 that a reference clock should have a phase jitter p-p of < 50ps. I have done research and I understand "phase jitter" is the same as TIE jitter? When measuring, I get very different result depending on the horizontal time scale used on the scope. It appears the best (correct) setting is whatever results in a histogram "x scale" of "5.00 ps /". Anything higher will make the reported p-p jitter many times larger. What is going on here?

Thanks.