From firmware 1.11 onward an over-voltage protection OVP or over-current protection OCP level reached will lead to switch off of the related channel as tonyalbus described in
Reply #45 .
Sometimes it is desirable to switch off all channels as for example symmetric supply of op-amps should be switched off together in case of one faulty supply.
The trigger function for the digital I/O lines can be used for this purpose:
Three lines are used as an output providing an "Output Off" event for each channel. One line is used as an input to switch off all channels.
The connector is wired according to the attached jpg using the connector screws, no soldering needed.
By the way the connector type is: Phoenix Contact MC 1,5/5-STF-3,81 with screwed connection, Nr. 1827732
The Trigger software setup is done in the menu Utility -> Trigger as following:
Data Line D0 set to "IN"
Ctrled Channel: CH1 CH2 CH3
Trig Type: Rise Edge
Sensitivity: Middle
Response: Output Off
Data Line D1 set to "OUT"
Ctrl Source: CH1
Trig Condition: Output Off
Polarity: Positive
Data Line D2 set to "OUT"
Ctrl Source: CH2
Trig Condition: Output Off
Polarity: Positive
Data Line D3 set to "OUT"
Ctrl Source: CH3
Trig Condition: Output Off
Polarity: Positive
Don't forget to enable the lines!
Now you can switch on the channels and the function is active.
As Pfriemler has discussed in
Reply #97 the trigger output signals are latched, so after an event, the level of the related line remains in its post-trigger mode.
To reactivate the above discussed function after tripping, there are different ways. The simplest I have found is to power cycle the Power Supply. Fortunately the trigger configuration is preserved with one exception, you have to enable all four lines again in the Trigger menu.