Author Topic: U1271A/U1272A/U1273A/U1273AX models/likes/dislikes/tips  (Read 23004 times)
Offline LaurentRTopic starter
« on: April 21, 2015, 06:00:28 am »
      For any prospective buyer or happy/unhappy owner, here is an attempt at listing all the good and bad things about the U1271A/U1272A/U1273A/U1273AX. Feel free to add/comment.
      I am very happy about my U1273A (I love the screen and the local and remote logging capabilities) but the thing has definite quirks and could be improved. It's more than 4 years old by now, so let's just hope for an eventual refresh  >:D
      Models:
      • U1271A ($347): U1272A minus some features (No Zlow, no SmartOhm, no AC+DC TrueRMS, 20kHz AC BW vs 100kHz, no 30mV range, no 30Ohm or 300MOhm ranges, no auto diode test, 200 entry logging vs. 10k)
      • U1272A ($378): Main model.
      • U1273A ($404): Same as U1272A but with OLED screen.
      • U1273AX ($451): Same as U1273A but with extended operating temperature range to -40C
      Latest Firmware:
      • U1271A/U1272A: 2.04 (2013-07-05)
      • U1273A/U1273AX: 1.95 (2013-08-05)
      Reviews:
      Likes
      • Solid. IP54 rated.
      • U1273A/U1273AX OLED screen indoor. Very readable large digits. Scales down for dual display. Better GUI especially for setup. (LaurentR)
      • Lots of logging options with good memory size. Good Bluetooth modules and apps/sw (LaurentR).
      • Arbitrary V/A conversion factor (electr_peter).
      • SmartOhm (electr_peter).
      • Continuity threshold adjustable with Range button. Backlight continuity alert on U1271A/U1272A (Zbig).
      • Calibration certificate with calibration results for all ranges.
      • Leads are high quality silicone. 15A rated and AWG16 (higher than all Fluke leads). Two-parts for more flexibility.
      Dislikes
      • Continuity is scratchy. It is latching, but still scratchy while a firm contact is not made. There is a funky multi-tone option for continuity that plays a melody depending on quality of the contact. Continuity threshold can be changed with the Range button. The U1272A also has a really nice screen flashing option (backlight turns on), which actually works better than the buzzer.
      • Connecting the leads to mains (or any high voltage) for enough time changes measurements on some ranges (e.g. the lowest resistance range). This is possibly due to self-heating. Recovery takes several minutes.
      • Capacitance measurements seem marginal. I don't own a DMM I can trust with capacitance tests, so I am not sure how true this is.
      • Zlow mode shows ~2-3VAC with shorted inputs.
      • The U1273A/U1273AX OLED screen is hard to see in bright daylight. It is gorgeous indoor (and the dot-matrix gives it a better interface than that of the U1272A), but it's really bad outdoor, to the point of just not being able to see anything. See for a comparison of the 2 screens outdoor.
      • The U1273A/U1273AX have reduced battery life due to the OLED display. Rated at 60hr Alkaline / 100hr Lithium vs. 300hr Alkaline for the U1272A.
      • APO is fully settable but beeps right as it turns off, leaving no time to prevent it.
      • Off position is not the left-most one on the rotary selector. There is a "Zlow" selection completely to the left. Minor annoyance.
      • Beep on power-on. The frequency of the beep can be picked (or the beep can be turned off), but this option changes the beeps of all functions, including continuity. The most ridiculous is the U1273A that has a melody option on power-up...
      • The screen seems to scratch very easily (on the U1273A at least). See this thread.
      • Provided leads are good quality but the handle is in two parts, which makes them quite long.
      Tips
      • For faster stabilization of readings, turn off DC LPF in settings (that's a change in the latest firmware).
      • After doing mains work, wait several minutes before doing precision electronics measurements (watching Ohms of shorted leads is a good indication).
      • Turn off unused features in setup (dB, mA scale, melody).
      • If using rechargeable batteries, set the battery type to "SEC" in setup.
      Accessories:
      • U1177A IR-Bluetooth adapter (10m range class 2)
      • U1117A IR-Bluetooth adapter (100m range class 1)
      • U1173A/B IR-USB adapter
      • Free iPhone, Android and Windows apps
« Last Edit: April 24, 2015, 05:10:21 am by LaurentR »
The following users thanked this post: PushUp
Offline retiredcaps
Re: U1272A/U1273A/U1273AX quirks and traits
« Reply #1 on: April 21, 2015, 06:06:27 am »
Martin's review of the U1272A.
« Reply #2 on: April 21, 2015, 06:18:13 am »
I had watched but somehow didn't list it :(
Thanks!
Online electr_peter
  • Country: lt
Re: U1271A/U1272A/U1273A/U1273AX quirks and traits
« Reply #3 on: April 21, 2015, 06:38:05 am »
Continuity is scratchy. It is latching, but still scratchy while a firm contact is not made. There is a funky multi-tone option for continuity that plays a melody depending on quality of the contact. I find it better than the plain continuity buzzer, but it's funky. The U1272A also has a really nice screen flashing option (backlight turns on), which actually works better than the buzzer. The U1273A has no such option.
U1272A has nice solid buzzer plus it can show full resistance range (in the mega Ohms) in continuity mode if range button is pressed.
At least they protect and inform you about the danger which is the most important thing. Crap DMMs only show high voltage symbol when you select high voltage AC/DC range independent of actual voltage - this is utterly useless gimmick.
Such type of DMMs cannot accurately measure capacitance on low ranges. If you are desperate with low value caps, try using very short leads - it helps reducing overall system capacitance.
Annoying, but not quite, other manufacturers also do this if Low Z mode is available. Actually, LowZ mode has much strong de-tent on purpose so lighter turn of knob to the left will land you in off position.
Also, both off and LowZ positions have a low impedance of ~1.7kOhm, be aware of this.
U1272 beep cannot be fully turned off - beeper is not completely disconnected and you can hear faint clicks in silent room. You only have option of beeps or silence (meaning only one setting for turn on beep, button beeps, continuity beep). Separate settings would be nice.
Provided leads are good quality but the handle is in two parts, which makes it quite long. Nothing a good set of Fluke TL71 can't fix...
Get single piece construction Agilent leads if you don't like two piece design.
Nice and accurate meter overall, has arbitrary amps/volts conversion factor and smart-Ohms measurement.
Offline Muttley Snickers
  • Cursed: 679 times
« Reply #4 on: April 21, 2015, 06:55:56 am »
I have a couple of U1272A's and only use them periodicallyand out of all the meters that I use its the only one that I find that a quick refresher course is required every time that they come out, some of the hold down features are not that intuitive unless you are using them frequently.
I keep meaning to print up and laminate a small cheat sheet and stick it in behind the bail, maybe tomorrow.
Also nearly forgot, on my 189 and also on a twenty year old Dick Smith meter I have the ability to latch the continuity buzzer with a momentary contact on the probes, this is handy when working in large complex's and factories, maybe there is a trick to getting it to work on the U1272A's.
Otherwise one of my favourite bits of kit...... :-+
Muttley
« Last Edit: February 13, 2016, 02:42:09 am by Muttley Snickers »
Offline wirbelwind
« Reply #5 on: April 21, 2015, 11:49:24 am »
Don't forget the short battery life on the U1273A and in the LowZ shows the meter 2V  - 3V without any voltage.
Offline 128er
« Reply #6 on: April 21, 2015, 01:08:17 pm »
and out of all the meter's that I use it's the only one that I
that they come out, some of the hold down feature's are
Even if the U1273A is my regular DMM for private work. I always have to look up some features in the manual, that I don't use very often. Or I try to remember and start poking around the knobs and get the function quite fast  ::)
sheet and stick in in behind the bail, maybe tomorrow.
Very good Idea. Would be handy.
Otherwise one of my favourite bit's of kit...... :-+
Confirm that  :-+
Like that too!
Things like the beep at start up, or Zlow at the most left position on the switch are not sooo annoying to me. I got used to it quite fast. But the start up melody I changed very quickly.
The crapy display cover however is shit. Nothing that should happen on a industrial grade meter.
About the Zlow range. Yes, my meter also shows usually 3,3 VAC. Even if the voltage that is present is for example 2.0 VAC.
And the Datasheet says:
-AC V and AC ?A/mA/A specifications are AC coupled, true RMS and are valid from 5% of range to 100% of range.
- After ZLOW voltage measurements, wait at least 20 minutes for thermal impact to cool before proceeding with any other measurement.
And the specification for Zlow range is 2% of range + 40 counts of LSD at 45 to 65 Hz.
Zlow is definitely not made for rocket science.
Offline Zbig
« Reply #7 on: April 21, 2015, 01:38:22 pm »
  • Continuity is scratchy. It is latching, but still scratchy while a firm contact is not made. There is a funky multi-tone option for continuity that plays a melody depending on quality of the contact. I find it better than the plain continuity buzzer, but it's funky. The U1272A also has a really nice screen flashing option (backlight turns on), which actually works better than the buzzer. The U1273A has no such option.
I feel it should be mentioned, to the meter's advangtage, that the continuity treshold can be selected with manual ranging button, if needed.
EDIT: Perhaps that could be the reason for the more snappy response of the 1272's LCD backlight (as compared to the buzzer), i.e. backlight - fixed-treshold, simple analog comparator, buzzer - software-selectable treshold after ADC?
  • Provided leads are good quality but the handle is in two parts, which makes them quite long. Nothing a good set of Fluke TL71 can't fix...
The upside being, I was able to get a pair of female shrouded banana to male unshrouded banana adapters from eBay which allows me to connect the meter direclty to my bench supply, MonsieurAdaptateur or any other gear that won't otherwise accept a shrouded plug.*)
*) Yes, I do realize I can just plug an unshrouded banana plug at the meter side just as well ;)
« Last Edit: April 21, 2015, 01:50:57 pm by Zbig »
Online macboy
« Reply #8 on: April 21, 2015, 01:40:47 pm »
handy when working in large complex's and factories.
Maybe there's a trick to getting it to work on the U1272A's.
Muttley
Please don't pluralize word's with apostrophe's. That makes it difficult for other's to read your writing.  :P
Use of U1272A's is acceptable. The rest are not and serve to confuse your meaning.
Offline Zbig
« Reply #9 on: April 21, 2015, 02:32:56 pm »
I will type whatever I like in any manner that I like until
a moderator say's otherwise, your the first to bitch.
Just FWIW, it annoys me no end as well and is just incorrect - plain and simple.
EDIT:
http://theoatmeal.com/comics/apostrophe  :P
« Last Edit: April 21, 2015, 02:37:59 pm by Zbig »
Offline Wytnucls
« Reply #10 on: April 21, 2015, 02:48:24 pm »
Foreigners make an effort to write in English properly. I don't think it is too much to ask of native speakers to do the same, if only as a sign of respect for their own language.
The upside is that over time, beginners learn the queen's English (or American) and it improves overall written communication on the site.
Offline Zbig
« Reply #11 on: April 21, 2015, 02:50:19 pm »
Not all of us went to school, some of us grew up in institution's
in fact, so if you dont like it then dont read it and fuck off.
Classy.
EDIT:
When I was a kid, I had to walk to school and it was uphill both ways. I guess that's the perfectly valid reason for me, as a grownup, to insult others when they point out the errors in my use of my native language, right?
« Last Edit: April 21, 2015, 02:57:06 pm by Zbig »
  • Cursed: 679 times
« Reply #12 on: April 21, 2015, 03:45:55 pm »
If you think that my grammer is bad wait till you see my cooking.. :palm:
You are welcome to go over every one of my posts, not one insult.
So I hereby retract and apologise.
« Reply #13 on: April 21, 2015, 06:45:50 pm »
Being a non-native English speaker, I'll do my best to watch my spelling  :-/O
I'll collate the positives that have been already mentioned and add them to a new section in the original post.
If you have more likes and dislikes about those DMMs, keep posting!
Offline Zbig
« Reply #14 on: April 21, 2015, 07:15:38 pm »
I had to invite my friends over for a cookout so I could finally see if my kitchen works. My task was beer supply ;) Peace! O0
Offline Zbig
« Reply #15 on: April 22, 2015, 06:03:32 pm »
And when we're at it, 230V across 1.7kOhm amounts to ~30W and 1000V - almost... 600W of dissipated power!? :scared: No way. What's really going on there? Is there a PTC in series with the shunt? If that's the case, shouldn't they be mentioning the widely varying input impedance in the datasheet? :-//
EDIT:
I'll answer this myself: yup, there is! That's probably what these several PTCs shown in Dave's teardown are doing. I've measured the Keysight's input resistance right after it was measuring mains voltage for a minute with another DMM. Sure enough, as the PTCs cooled down, the resistance dropped from several tens of kOhms to - wait for it - some 1.5kOhms, which is actually less than the initial 1.7kOhms! Go figure. I wonder if that recovery period doesn't affect other kinds of measurements.
« Last Edit: April 22, 2015, 06:20:21 pm by Zbig »
Offline Zbig
« Reply #16 on: April 22, 2015, 09:51:47 pm »
There's one more thing I find mildly annoying. My cheap and portable UNI-T UT33A emits a warning beep some time before it goes to auto-sleep so you have a chance to prevent it by pressing a button. Keysight doesn't do that - it just beeps the exact moment it's switching off. Fortunately, you can either override the APO temporarily when turning the meter on or permanently - from the menu.
Re: U1271A/U1272A/U1273A/U1273AX models/likes/dislikes/tips
« Reply #17 on: April 24, 2015, 05:03:39 am »
Updated the original post with a bunch of new material including what has been said in this thread.
Some other comments:
- For safety reasons I am not fully understanding, the new recent firmware has a default LPF on DC measurements. I have found that this noticeably slows down the display rate and stabilization time of DC measurements. I turned it off in setup.
- I found that sometimes the mVDC range has a substantial offset (typically ~20uV, sometimes up to 50uV) that slowly drifts back to 0 in most cases. The spec clearly says the offset (20 counts == 20uV) is only with offset Nulled out, but I am surprised the offset sometimes stays for a while. I though it would be some thermal effect that goes away by itself, but no.
I remember the U1272A doing the faint clicks in silent mode, but the U1273A doesn't do it. I wonder if it's linked to the U1272A's backlight continuity feature (even if it's turned off).
Offline 128er
« Reply #18 on: April 24, 2015, 06:57:23 am »
http://cp.literature.agilent.com/litweb/pdf/U1273A-01-S.pdf
Online electr_peter
  • Country: lt
« Reply #19 on: April 24, 2015, 08:02:19 am »
I actually prefer meter not giving pre-emptive warning with sound for APO. Case in point - Amprobe AM-240. It has APO turned on by default. After some time it emits super annoying beeping sound, then waits few minutes, beeps again and turns off. I hate this behaviour - just beep once before actual turn off. There is no need to inform annoy anyone that "Hey, I have reached half timer mark for APO".
Is DMM connected to something during measurement? (open leads -> random readings) Is filters turned on? Low range is sensitive to all sorts of interference pick-up, especially with long leads.
In case of U1272A, buzzer circuit is not completely turned of in HW or SW. U1273A model probably has different buzzer circuitry because it can play melodies.
Offline echen1024
  • 15 yo Future EE
« Reply #20 on: April 26, 2015, 04:07:28 am »
Let me think...
I really like the huge resistance range, up to 300MOhm.
I really wish it would work with Agilent's amazing BenchVue software, since that is where I mostly use my DMM. I understand that BenchVue is likely meant for a bench dmm. I also dislike how it does not have a dedicated enter key or markings, which make the menus harder to navigate at first. And the ability to bisplay measurements in dB is also a really nice feature.
I'm not saying we should kill all stupid people. I'm just saying that we should remove all product safety labels and let natural selection do its work.
https://www.youtube.com/user/echen1024
« Reply #21 on: April 26, 2015, 04:19:00 am »
Let me think...
Interestingly, the new version of the HH DMM logger software has been reskinned to looked pretty much exactly like BenchVue. I use both and it's too bad they're different products. I guess BenchVue is Visa-based, but still (the HH DMMs actually use very SCPI-like commands through a serial port, so it's not that far).
Offline VK5RC
« Reply #22 on: April 26, 2015, 06:14:03 am »
Whilst a little off a U1273, I am a big fan of oleds mainly because of the viewing angle and readability (in doors).  Happy U1253b owner (5yrs).
Offline nowlan
« Reply #23 on: April 26, 2015, 07:43:58 am »
Wasnt there an issue with the oled lifetime. Some warranty stink on the forums last year.
Offline VK5RC
« Reply #24 on: April 26, 2015, 12:29:46 pm »
There was a guy with a u1253a who had it fade. I have had some contact with other u1253b guys and not heard of others having similar issues, I think MTTF of OLED is thought to be higher 2-5x compared to LCD but that is still pretty infrequent.
