Thanks, Joeqsmith.
I continue to believe that the clamp is just a zener and not a high voltage TVS, because it has to clamp at very low voltage which is unusual for TVS. At anything over a few volts, it will need to dissipate, and survive after, several watts. Remember the applied voltage may be any value below 1000V; at 100V, there is a current of just under 100mA, and the PTC would take longer to cut this down. At some even lower voltage, the current will get to be too small to heat the PTC, yet be large enough to cook the zener, unless it clamps at just a few volts. The zener is 500mW at best and has no lead conducted cooling. More, it has done its job through the catastrophe when the switch arced over, and is none the worse for it after. I would not hesitate to test it all out, but Keysight want to investigate further in their own way and would need the evidence in tact.
Agree that we still don't know what caused the switch to arc over in the first place.
There is data on 1N4574, but the marking is 457A. 4574 is a special temperature compensated reference zener at 6.4V.