I dunno honestly.
It's just that all the instructions for performing the hack simply state "connect via Ethernet"; "turn DHCP on" followed by basically "it's go time".
So reading this my basic thought was connect PC directly to generator and it should work. ![Confused :-//](https://www.eevblog.com/forum/Smileys/default/confused0024.gif)
Unlike Performa, I
will nitpick..
Siglent is
hardly responsible for the fact that instructions how to hack instruments are not friendly to users that don't know much about networking....
And also Siglent in it's manual explains it all:
Note:
⚫ If the generator is connected to the PC directly, set the IP addresses,
subnet masks and gateways for both of the PC and generator. The subnet
masks and gateways of PC and generator must be the same and the IP
addresses of them must be within the same network segment.
⚫ If the generator is connected to the LAN of your PC, please contact with
your network administrator to get an available IP address. For details,
refer to the TCP/IP protocol.
4) DHCP Configuration Mode
In DHCP mode, the DHCP server in the current network assigns LAN
parameters, e.g. IP address, for the generator. Press DHCP to select “On” or
“Off” to turn DHCP mode on or off. The default is “Off”.