Re: The Siglent SDG2042X Thread
« Reply #1700 on: September 06, 2023, 02:51:19 pm »
When you say *INSTR, don't you mean *IDN?  ?

I don't recognize *INSTR as a valid query and, even if it were, it would require the question mark in order to be a query.
*INSTR is used as a parameter to the Visa FindRsrc command.  See Chapter 5 of the NI Visa Programmer's Guide. 

An excerpt:  The viFindRsrc() operation matches the value specified in the expr parameter with the
resources available for a particular interface. A regular expression is a string consisting of
ordinary characters as well as special characters. You use a regular expression to specify
patterns to match in a given string; in other words, it is a search criterion. The viFindRsrc()
operation uses a case-insensitive compare feature when matching resource names against the
regular expression specified in expr. For example, calling viFindRsrc() with
“VXI?*INSTR” would return the same resources as invoking it with “vxi?*instr”.


Use this function within instrument control applications to scan for GPIB, LAN, RS232 or USB devices. 
 
The following users thanked this post: tv84, KungFuJosh

Offline tautech

  • Super Contributor
  • ***
  • Posts: 29492
  • Country: nz
  • Taupaki Technologies Ltd. Siglent Distributor NZ.
    • Taupaki Technologies Ltd.
Re: The Siglent SDG2042X Thread
« Reply #1701 on: September 11, 2023, 11:05:39 am »
New firmware for SDG2000X models.

Version: V2.01.01.37R8
38.6 MB
https://int.siglent.com/upload_file/zip/firmware/Signal_generator/SDG2000X_V2.01.01.37R8_EN.zip

Release notes
1. Solve the problem of slow USB drive detection during startup
2. Solve the problem of outputting abnormal pulses during burst mode
3. Add instructions for saving and reading status(XML)files
4. Add the instruction to clear the statistics of the counter
Avid Rabid Hobbyist.
Some stuff seen @ Siglent HQ cannot be shared.
 
The following users thanked this post: KungFuJosh, Johnny B Good, Bad_Driver, gjvdheiden, Wrenches of Death

Offline NoisyBoy

  • Frequent Contributor
  • **
  • Posts: 505
  • Country: us
Re: The Siglent SDG2042X Thread
« Reply #1702 on: September 13, 2023, 03:04:03 am »
Thanks Tautech,

I upgraded mine, absolutely straight forward update.
 

Offline Vit G

  • Contributor
  • Posts: 40
  • Country: kz
Re: The Siglent SDG2042X Thread
« Reply #1703 on: September 18, 2023, 06:15:54 am »
Hi, Tautech!

    Can I hope that you will someday correct the phase shift between channels that cannot be eliminated by the EqPhase button?
-9 Degrees error on 100MHz (abt 0.25nS)
    Phase Coupling procedure is not a solution that should be used. it is only needed to compensate for the length of the cables if it is different. with the same length there should be enough EqPhase button!
Studying metrology and LNA
 

Offline tautech

  • Super Contributor
  • ***
  • Posts: 29492
  • Country: nz
  • Taupaki Technologies Ltd. Siglent Distributor NZ.
    • Taupaki Technologies Ltd.
Re: The Siglent SDG2042X Thread
« Reply #1704 on: September 18, 2023, 07:57:24 am »
Hi, Tautech!

    Can I hope that you will someday correct the phase shift between channels that cannot be eliminated by the EqPhase button?
-9 Degrees error on 100MHz (abt 0.25nS)
    Phase Coupling procedure is not a solution that should be used. it is only needed to compensate for the length of the cables if it is different. with the same length there should be enough EqPhase button!
Do you have the latest firmware installed ?

I have seen this before with identical cables but with EqPhase it is much reduced then to add some small Phase adjustment can be nulled completely out. However as you change frequencies phase can move slightly and you need make further adjustments.

Attached from a SDG6022X, 20 MHz, 500mV and identical cables with EqPhase and channel copy applied where any phase difference is tiny then have added 5o to show a difference between channels.
Avid Rabid Hobbyist.
Some stuff seen @ Siglent HQ cannot be shared.
 

Offline Vit G

  • Contributor
  • Posts: 40
  • Country: kz
Re: The Siglent SDG2042X Thread
« Reply #1705 on: September 19, 2023, 06:57:44 am »
Nothing changed after the last update. The same difference of 250 pS between channels. If the EqPhase button still doesn't work, then replace it with something more useful =) The cables I use are identical in length. I even swapped them, nothing changed. Look how Rigol did it. when you press the EqPhase button, their device very accurately phases the outputs
Studying metrology and LNA
 

Online MathWizard

  • Super Contributor
  • ***
  • Posts: 1628
  • Country: ca
Re: The Siglent SDG2042X Thread
« Reply #1706 on: October 23, 2023, 06:29:51 am »
How much better is the SDG6022X than a SDG2142X ? (which I own , but I would like closer to 0.5-1GHz)

What else is on par with the 6022X ? I'm sort of locked into Siglent tho, since my scopes are Siglent, and they can work together with the AWG.
 

Offline tautech

  • Super Contributor
  • ***
  • Posts: 29492
  • Country: nz
  • Taupaki Technologies Ltd. Siglent Distributor NZ.
    • Taupaki Technologies Ltd.
Re: The Siglent SDG2042X Thread
« Reply #1707 on: October 23, 2023, 07:19:28 am »
How much better is the SDG6022X than a SDG2142X ? (which I own , ...)
SDG6000X is a 500 MHz capable design and they are both all but identical to operate.

Quote
but I would like closer to 0.5-1GHz
SDG7000A models within budget ?

Would an RF gen like the 2.1 GHz SSG3021X meet your HF needs ?
Avid Rabid Hobbyist.
Some stuff seen @ Siglent HQ cannot be shared.
 

Online coromonadalix

  • Super Contributor
  • ***
  • Posts: 6746
  • Country: ca
Re: The Siglent SDG2042X Thread
« Reply #1708 on: October 23, 2023, 07:31:27 am »
and the SSG3021X  can be more ....
 

Offline tautech

  • Super Contributor
  • ***
  • Posts: 29492
  • Country: nz
  • Taupaki Technologies Ltd. Siglent Distributor NZ.
    • Taupaki Technologies Ltd.
Re: The Siglent SDG2042X Thread
« Reply #1709 on: October 23, 2023, 07:35:57 am »
and the SSG3021X  can be more ....
Correct, more economical than SDG7000A series when you only need higher BW sinewaves.

SSG3021X has been my choice for such needs.
Avid Rabid Hobbyist.
Some stuff seen @ Siglent HQ cannot be shared.
 

Offline tautech

  • Super Contributor
  • ***
  • Posts: 29492
  • Country: nz
  • Taupaki Technologies Ltd. Siglent Distributor NZ.
    • Taupaki Technologies Ltd.
Re: The Siglent SDG2042X Thread
« Reply #1710 on: October 23, 2023, 08:43:30 am »
and the SSG3021X  can be more ....
Yes mine is SSG3032X now......
Avid Rabid Hobbyist.
Some stuff seen @ Siglent HQ cannot be shared.
 
The following users thanked this post: NoisyBoy

Offline RoV

  • Regular Contributor
  • *
  • Posts: 189
  • Country: it
Re: The Siglent SDG2042X Thread
« Reply #1711 on: October 25, 2023, 08:39:16 pm »
My SDG2042X is updated with FW V2.01.01.37R8. Today I have discovered a weird behaviour in sweep mode: output amplitude shows a step around ~293 kHz, being correct at >293 kHz and something like 10% higher than normal below that frequency. See attachments. This happens regardless of waveform, sweep time, log/lin sweep, channel 1 or 2. At the same frequencies, in normal (non sweep) mode, the instrument works fine.
By searching in this same eevblog topic, I have discovered this msg from Tjuurko that claims a different anomaly:

In new firmware (V2.01.01.37R*), button "SWEEP" increases the amplitude of the output signal: CH1 +25%, CH2 +5%.
Without cables connected, the output overvoltage protection is activated.
Tested for SDG2042X (SV: 2.01.01.35R3B2, HV: 02-01-00-40-00).

At this point my suspect is that there is a bug and calibration is bypassed in sweep mode, so the instrument works in uncalibrated mode and different things can happen to different machines.
Can you try on yours?

I annex a few photos of the SDG sweeping, connected to my scope. Last photo is on a reduced interval around 293 kHz, with a zoom on the sinusoid showing the abrubt amplitude change.

EDIT: I don't understand why, but I see the images rotated by 90° after publication. Original images are correct.
« Last Edit: October 25, 2023, 08:44:18 pm by RoV »
 

Offline RoV

  • Regular Contributor
  • *
  • Posts: 189
  • Country: it
Re: The Siglent SDG2042X Thread
« Reply #1712 on: October 30, 2023, 08:10:05 am »
My SDG2042X is updated with FW V2.01.01.37R8. Today I have discovered a weird behaviour in sweep mode: output amplitude shows a step around ~293 kHz, being correct at >293 kHz and something like 10% higher than normal below that frequency. See attachments. This happens regardless of waveform, sweep time, log/lin sweep, channel 1 or 2. At the same frequencies, in normal (non sweep) mode, the instrument works fine.

I have solved the problem by downgrading the FW:
SDG2000X_V2.01.01.37R8_EN    NO (the most recent one discussed in the original post)
SDG2000X_V2.01.01.37R6_EN    NO
SDG2000X_V2.01.01.37R3_EN    NO
SDG2000X_2.01.01.35R3B2_EN  OK
From 37R3 onwards they support new HW versions, they likely messed with backwards compatibility (I forgot to say that my unit is dated 2017, HW revision 02-01-00-41-00).  |O returning to the 35R3 loses several improvements.
 
The following users thanked this post: tautech, gjvdheiden, phecap

Offline tautech

  • Super Contributor
  • ***
  • Posts: 29492
  • Country: nz
  • Taupaki Technologies Ltd. Siglent Distributor NZ.
    • Taupaki Technologies Ltd.
Re: The Siglent SDG2042X Thread
« Reply #1713 on: October 30, 2023, 08:58:05 am »
Thanks much for your investigations RoV...... now reported.
Avid Rabid Hobbyist.
Some stuff seen @ Siglent HQ cannot be shared.
 
The following users thanked this post: RoV, gjvdheiden, phecap

Offline phecap

  • Contributor
  • Posts: 15
  • Country: nl
Re: The Siglent SDG2042X Thread
« Reply #1714 on: October 30, 2023, 10:28:20 am »
My SDG2042X also has the same problem.
The hardware revision is also 02-01-00-41-00.
I also installed the latest firmware version (SDG2000X_V2.01.01.37R8_EN).
 
The following users thanked this post: Johnny B Good, RoV

Offline maksim36

  • Newbie
  • Posts: 1
  • Country: us
Re: The Siglent SDG2042X Thread
« Reply #1715 on: November 14, 2023, 03:21:04 am »
Does anyone know what display model is used in the sdg2042x generator? My display is broken.
 

Offline Roger Need

  • Regular Contributor
  • *
  • Posts: 75
  • Country: ca
Re: The Siglent SDG2042X Thread
« Reply #1716 on: November 15, 2023, 01:16:25 am »
My SDG2042X also has the same problem.
The hardware revision is also 02-01-00-41-00.
I also installed the latest firmware version (SDG2000X_V2.01.01.37R8_EN).

I have the same problem too with older firmware

Hardware version 02-02-00-40-00
Software version 2.01.01.37R6
 

Offline TheDefpom

  • Frequent Contributor
  • **
  • Posts: 766
  • Country: nz
  • YouTuber Nerd - I Fix Stuff
    • The Defpom's Channel
Re: The Siglent SDG2042X Thread
« Reply #1717 on: November 15, 2023, 08:58:26 am »
Argh... this bug explains something I saw recently when I was doing a review on Micsigs new 150MHz differential probe... I noticed some steps on the bode plot and thought it was the probe doing it, looks like it was the sig gen.
Cheers Scott

Check out my Electronics Repair, Mailbag, or Review Videos at https://www.youtube.com/TheDefpom
 
The following users thanked this post: bicycleguy

Offline ozkarah

  • Regular Contributor
  • *
  • Posts: 96
  • Country: tr
Re: The Siglent SDG2042X Thread
« Reply #1718 on: November 17, 2023, 04:52:08 pm »
Same problem here.

An amplitude increase around 110-120MHz:

 

Offline RoV

  • Regular Contributor
  • *
  • Posts: 189
  • Country: it
Re: The Siglent SDG2042X Thread
« Reply #1719 on: November 17, 2023, 05:08:57 pm »
This is not the same problem (not a step in the response) and, according to me, is probably not a problem at all. The instrument specs give +/-0.4 dB in the 100-120 MHz range which is already close to your results. You didn't say anything about cable, connectors, kind of termination, scope used, so there is a potential margin to justify some deviation on the highest frequencies.
You should check behavior below 1 MHz, where the problem is commonly seen. Please also report HW and FW versions.

Offline Martin72

  • Super Contributor
  • ***
  • Posts: 6873
  • Country: de
  • Testfield Technician
Re: The Siglent SDG2042X Thread
« Reply #1720 on: November 18, 2023, 09:06:34 pm »
Hi,

Output 2 is active at every (re)start, I always have to switch it off manually.
Power on state is last.
How can this be?

(SDG2122X (formerly 2042X), latest firmware)
"Comparison is the end of happiness and the beginning of dissatisfaction."(Kierkegaard)
Siglent SDS800X HD Deep Review
 

Offline bicycleguy

  • Frequent Contributor
  • **
  • Posts: 265
  • Country: us
Re: The Siglent SDG2042X Thread
« Reply #1721 on: November 19, 2023, 12:09:49 am »
@Martin72,
That's funny, I can't get mine to start up with either channel on, no matter the power setting.  Maybe this is as it should be.

At least once in the past the internal files got messed up with firmware revisions.  The fix was setting the to powerOn to default, cycle power and then back to user.

Good luck!
 

Offline bicycleguy

  • Frequent Contributor
  • **
  • Posts: 265
  • Country: us
Re: The Siglent SDG2042X Thread
« Reply #1722 on: November 19, 2023, 12:16:52 am »
@Martin72,

Figured it out.  For each channel, Utility/Output Setup/Page2/PowerOn State/Off  return.
 
The following users thanked this post: tautech, Martin72

Offline Martin72

  • Super Contributor
  • ***
  • Posts: 6873
  • Country: de
  • Testfield Technician
Re: The Siglent SDG2042X Thread
« Reply #1723 on: November 19, 2023, 12:17:15 am »
Quote
I can't get mine to start up with either channel on

I could understand that for "security reasons".
But as soon as the generator has started up, I don't understand that a channel is "online" straight away - I think that's a bug.

EDIT:

Quote
For each channel, Utility/Output Setup/Page2/PowerOn State/Off  return.

Ahh...
"Comparison is the end of happiness and the beginning of dissatisfaction."(Kierkegaard)
Siglent SDS800X HD Deep Review
 

Online mawyatt

  • Super Contributor
  • ***
  • Posts: 3920
  • Country: us
Re: The Siglent SDG2042X Thread
« Reply #1724 on: November 21, 2023, 02:35:30 pm »
Interesting feature within that allows one to construct waveforms based upon fundamental sinewave and harmonics. Here's a result showing the much discussed "Gibbs Ears" on another thread and created by summing odd harmonics up to AWG limit of 15.

Best,