One of my contacts at Sourcetronic was kind enough to share how to do the Dissipation Factor test. He said:
For the D test, Cs-D fundamentally has the same information as Cs-Rs, as the loss factor is directly connected to the ESR Rs.
That also means that you can fake the D by putting actual resistors in series to the capacitor, and you get different sizes of D depending on the resistance you put up.
So I tried it out, and it works. I threw different value resistors in series, and it gave DF values similar to what was expected using math (eww, gross)...just kidding, I used a
calculator. 🤣
Measured Rs of the cap was about 29Ω at 1kHz. So using some standard R values, I can do something like this:
29Ω = 0.00018 (cap only)
79Ω = 0.00049 (29Ω + 50Ω)
162Ω = 0.00101 (29Ω + 133Ω)
797Ω = 0.00506 (29Ω + 768Ω)
1609Ω = 0.01010 (29Ω + 1.58k)
16029Ω = 0.10071 (29Ω + 16k)
160029Ω = 1.00530 (29Ω + 160k)
I'm sure those numbers aren't perfectly accurate, but they should be close enough to test it on a new test board. 😉
Thanks,
Josh