Iso-Tech IDM 505 problem
« on: July 13, 2024, 09:50:43 am »
Sorry! I realise this is in the wrong topic!

I've recently been given an Iso-Tech IDM 505 which has a problem. When I turn it on, the device emits a continuous beep and shows "PROBE" on the display. The only setting that works is the A setting which works fine and is measuring current OK.

I've tried a reset but nothing has changed.

Has anyone come across this before? I have no idea where to start looking!

Thanks.
« Last Edit: July 13, 2024, 10:17:57 am by mattltm »
 

Offline mattltmTopic starter

  • Newbie
  • Posts: 8
  • Country: gb
Re: Iso-Tech IDM 505 problem
« Reply #1 on: July 14, 2024, 02:19:23 pm »
OK. Managed to work this one out for myself. A small reflow job on the probe terminals and it's working fine!
 