I have made some more tests with spectrum analyzer.
Here in these pictures signal from SDG5082 is of course sinewave. Level is in all pictures 0dBm and naturally 50ohm. Signal go around 2.5m in coaxial to HP8568B spectrum input. (some amount crap cable but no matter in this test)
Note about spectrum analyzer. Freq related some calibrations are not perfect but in this case this do not mean "anything" just "cosmetics" in numbers.
I have used 10Hz RBW and 100Hz RBW.
In some image signal is 20dB over top reference line! (note this as you think trace levels - as told signal is always 0dBm level)
Note for this displayed traces. In all pictures mode have been MAX hold. So this level is more like peak noise level than avarage noise level!
Many times in specs is use average noise levels or displayed average noise levels or something like it. With human eye it may somehow imagine where is average.
Then, most specs today give dBc/Hz
Oh well, it is nice but it is not what you see.
How to convert displayed noise level for dBc/Hz.
Ok, example. If carrier (this c) is level 0dBm and then example 1kHz we can see trace level is
-50dBm. We can say its level is -50dBc (dB related to carrier level) Then RBW (resolution bandwidth) is 100Hz. It is 100 times more wide than 1Hz RBW. Ok, it is 20dB. So this -50dBc with 100Hz filter responds -70dBc/Hz (70dB below carrier if RBW is 1Hz)
Why I use 100Hz RBW is just becouse my timetable was busy and narrrow RBW sweep takes forever.
Image A
80MHz carrier is just left and there is 1kHz span and 10Hz RBW
Reference level is same as signal (0dBm)
note for dBc/Hz -10dB due to 10Hz RBW
(if you think reference level is -10dB this give right level for dBc/Hz)
Image B
10MHz carrier is just left and there is 1kHz span and 10Hz RBW
Image C
1MHz carrier is just left and there is 20kHz span and 100Hz RBW
Reference level is -20dB under carrier level.
note for dBc/Hz -20dB due to 100Hz RBW
(if you think reference level is -40dB this give right level for dBc/Hz)
Image D
50MHz carrier is just left and there is 200kHz span and 100Hz RBW