The new driver IC cured the missing 'a' segment issue in the MSD of the 432C, but it's also looking like I may need to replace the one for the LSD, too - circuit for the bottom segment 'd' appears to be leaky, too, as it stays partially on when it shouldn't be. Looks like I'll be doing more surgery on the display unit.
The other issue I noticed is related to the panaplex display itself - the instrument must have spent a LOT of time either in autorange with no signal present, or on one of the 10X ranges as the glass in front of that decimal point is heavily sputtered and opaque (note the silvered spot roughly in the center of the picture at the corner of the topmost full digit), and a fair bit though not as much time in the 100X range (DP at bottom corner of lower digit) - partially blocked by sputtering. The 1X decimal point (top, adjacent to '1' digit) is nearly clear. The slightly dim center section of the segment that was out before seems to have mostly recovered after I left the meter on all day yesterday giving it some burn-in time (note that it is considerably brighter and more evenly lit in the first picture in this post, which was taken last night after it had been on for about 12 hours)
1X decimal point lit:
10X decimal point lit (hidden behind silvered spot when viewed from the front):
That said, it /sort/ of agrees with my 437B given that I've done nothing yet beyond fixing the display section.
I found replacement SP-351 displays on our favorite wallet-sucking site and have them enroute, but may try to rig up some orange SMD LEDs on the display face for the decimal points first, rather than replacing the otherwise operational panaplex display. Frustrating that it can be rendered inoperative by such a tiny fault.
-Pat
Worth looking out for Heathkit clocks and calculators. They are probably cheaper sources if you don’t mind a gamble. I was never a fan of panaplex displays. All the reliability of light bulbs multiplied by the number of the segments in one inconveniently large package
But do the Heathkit cloc-u-lators use the same size/height/type of display? Also Panaplex displays are used in some pinball machines in high quantities, probably in high demand there too. I'm sure "bigclive" shows the inside of one in a YT video, but it's probably OT and can't be posted here.
I never did add the pictures of mine, in a HP 432B (no auto ranging on 'B' model), it also has the same worn decimal point issue, but not so bad that it isn't visible. If it was replaced you would need to either transfer the over-range decal, or make a new one.
Also have a panel voltmeter that uses similar or the same displays. We had a very large cable crimp pull test machine at work with a similar panel meter, but it got scrapped without my knowledge, not that I could have rescued it anyway, due to the no you can't buy it policy ATM.
They are probably filling more skips, during the shutdown fortnight, as I post this.
Here is my Data Tech panel DVM, almost looks unused in comparison to display in the HP 432B.