Nixie DMM updatePROGRESS has just been made !!!
There is now measurization going on here !!!
![Fighting :box:](https://www.eevblog.com/forum/Smileys/default/fighting0030.gif)
Could not help, could not go to sleep, I was too curious and impatient... I started swapping boards with the working DMM.
Remember I said it looked like the first / LSD digit was counting, but the second one next to it, looked like it was counting only between '0' and '1' and no more ?
Well that was it indeed ! I swapped that board and now all 4 digits are alive and counting !!!
![Fighting :box:](https://www.eevblog.com/forum/Smileys/default/fighting0030.gif)
So I need to fix that board...
There is still a long way to go though, but I am happy that basic operation has now been restored, it basically "works".
So from what I see, remaining problems are :
I fed it with 2.5V from my Chinese Vref.
- On the 5V range (High impedance input, > 1,0000 Meg) = 2,65V, close enough for now...
- On the 5V range ( 5 Meg impedance) = 2,65V also.
- On the 50V range = 4.36V ohhh..... not good.
- On the 500V range = 22.6V !!!
![Face Palm :palm:](https://www.eevblog.com/forum/Smileys/default/facepalm.gif)
So since the reading is "good" (enough) on the 5V range which is the straight path to the A/D converter, that means the entire chain of measurement is basically fine.
It only reads non-sense on the 50V and 500V ranges, which means the problem must be limited to, and located, on the attenuator board. So I will need to look at that...
Also, new problem : the polarity Nixie quit working ! It doesn't light up at all anymore, there is no MINUS or PLUS sign anymore !
![Sad :(](https://www.eevblog.com/forum/Smileys/default/sad.gif)
Plus of course I need to make a new Vref board of course. that won't happen overnight of course, so In the meantime I could swap board with the working DMM to let me move on with my troubleshooting.
So to sum it up for now :
1) A failed cap in the power supply, need to tidy up my repair now.
2) A Vref board to redesign/make
3) Failed polarity Nixie to troubleshoot
4) LSD+1 Nixie board to troubleshoot
5) Attenuator board to troubleshoot.
6) Power switch that's stuck open.. sealed unit it looks like, so not too hopeful
7) Cosmetic restoration / corrosion / pitting to attend to.
So still far from done, but at least now it basically "works", the rest has a clear path to it so not worried...
2H30 AM, I am about to lose consciousness so going to bed, see ya later....