Some news from the fun with TEA department:
Maybe you remember I've got fake AD587 references from china and somewhat later real AD587UQ from Mounty.
I've put them on a simple board (don't have a photo readily available right now) and put them into that white cardboard box. Connected four of them (two fakes, two original) to the HP34401s and log the data:
Here's the first results (another colourful grafana picture):
Traces G and H are the original AD587UQ, E and F are the fakes. At some point (about 11:33) I've switched the HP34401 to slow 6 digit aquisition mode, so this is the reason for the reduced noise. At about 14:07 I've had the serial to USB converter taken out by a power switching transient (nothing impressive, just turned on the computer power circuit by a remote controlled contactor), so I had to restart the aquisition. Another proof USB is just crap.
All plots are normalized to the actual reference output voltage at about 11:40, so the graphs show the difference to each individual voltage at that time.