News from the nixies: The 5245L is counting!
Refilled the cans, see photos below.
I fixed the decimal counter assembly with a silicon. The problem was the cap C21 I measured OK broke its leg, the tiny gap was non-obvious. After resoldering the shortened leg, the "carry bit" now works. While experimenting with the board, I managed to kill the other 2N2455 Q10 on the board
Fixed with a silicon PNP too. The needle pulses I was missing are clearly visible with a faster scope (500MHz instead of 100MHz).
The high speed display assembly 05232-6010 with the dead nixie is counting, but the photocoupler is bad. Digits 0 and 6 are not working OK, digit 3 is somehow blurred. Sometimes the 4 is displayed for a short time instead of the 6. Swapped a good nixie from another board to this for tests. Will try to repair it later, currently this one has no display.
All other display boards are counting and displaying OK.
Several decimal neon bulbs are dead (broken legs at the glass suspected to corrosion), have to remove and refit new neons to it. Will do that in the next days, I hope.
The scaler output at the back has some glitches when scaling down to 100, 10 or 1cps. Have to look into it.
The fan is alarmingly loud, have to remove and rework it.
The oven is working but the frequency is off by several Hz, this has to be adjusted as the last step.