Dana 5100 DMM update:
Budget 'scope arrived. I'm back to troubleshooting. The 5100AF manual is very good but it is for a 5100AF. There are small differences. For example the 5100 clock is 1MHz and the 5100AF clock is 2MHz.
Everything was looking OK until I checked the Null Detector output. Flatline.
My guesses are either C2, which is a tantalum or the 1V reference. I'm not going to guess. The high voltage buffer is on the same card as the digitiser. Anything is possible.
For those following along, I've attached the schematic. The 5100 uses a classic dual-slope converter. The unknown voltage charges capacitor C2 for a known time period, which is 50mS in the 5100AF
Then the capacitor is discharged by the reference voltage (in reverse polarity) - Which should trigger the null detect unless it times out.
Either the reference is faulty or something is wrong on this board.