So, the cat's asleep, SWMBO's asleep, so I finally get some peace and quiet to finish the calibration cycle on that HP 66312A with a run through the official "verification" procedure from the manual.
So, results:
Test Description | Minimum Spec | Recorded Results | Maximum Spec |
Voltage Programming and Measurement |
Low Voltage out | -10mV | 779uV | +10mV | PASS |
- front panel measurement error | Vout-3mV | +221uV | Vout+3mV | PASS |
High Voltage Out | 19.980V | 19.999V | 20.020V | PASS |
- front panel measurement error | Vout-9mV | 0mV | Vout+9mV | PASS |
Current Programming and Measurement (High Range) |
Low Current Out | -1mA | 615uA | +1mA | PASS |
High Current Out | 1.998A | 1.999 94A | 2.002A | PASS |
- front panel measurement error | Iout-4.5mA | +143uA | Iout+4.5mA | PASS |
Current Measurement (Low Range) |
Low Current measurement error | Iout-2.5uA | +720nA | Iout+2.5uA | PASS |
High Current measurement error | Iout-22.5uA | +1uA | Iout+22.5uA | PASS |
Current Sink Measurement |
Low Current Sink measurement error | Iout-22.5uA | -4uA | Iout+22.5uA | PASS |
High Current Sink measurement error | Iout-4.85mA | -100uA | Iout+4.85mA | PASS |
Given that this is an HP badged unit, which dates it to the 1999 name change to Agilent or earlier, making this unit at least 21-22 years old, I'd say that was a
very creditable performance. All specs met, most by a decent margin.
I certainly don't exceed the specifications that you'd have expected
me to meet 21 years ago, not by a long chalk. It'll be interesting to repeat this in a year's time and see if it's drifted. At this age, I suspect that all the drift has happened, or at least all the major sources of it.
So, at the cost of £160 all told, a little help from Robert with some spare binding posts, and a bit of (mostly friendly
) hissing and spitting from Anders when he was beaten to it, I'm a very happy man.