Just for the records:
calibration is not adjusting.
Calibration means: comparing the DUT to a reference and checking, if it is within specs.
If it is within specs, the differences will be documented.
If the DUT is out of specs, then an adjustment or repair of the DUT can be considered.
So if you send a meter away for calibration, it will come back to you with the same settings as you sent it with?
Yes. The most valuable part of such a meter is its documented history, because based on this one can judge the drifting of a meter and its quality. It can be, that if a meter is drifting out of spec it will not be adjusted, because this means breaking the history.
I was under the distinct impression that if the meter was in need of an adjustment, it would be made and then returned to you, with a demonstrable calibration.
I've sent my 34401A to Keysight in Böblingen for calibration and they stated clearly, that they will document only.
If adjustment or repair is needed, they will contact me, asking, how to proceed. Not only because of the additional costs but
mainly because of the history of the instrument.
When I read the service manual for my BlackStar 4503, it gives precise instructions on how to check each range. Then how to enter the calibration mode and make the required corrections to its parameters and saving those corrections to memory, which is what I did after repairing the damage done by the leaking battery.
I think, there are different rules for different manufactures. I think, Keysight and Fluke are handling it that way I've described above.