@med: 585A is one step closer to fixed. I installed one of the new trimmer caps in place of the broken C348 and I've run through the time base cal procedure twice. Everything looks pretty good. Still a little bit of nagging nonlinearity on time base A in the .5, .2, and .1ms/cm settings. R160Z (which is used to adjust the 1ms/cm setting) is out of range to lengthen the sweep any further. These sweep speeds are technically in spec (the manual quotes +/-2.4mm when using the 1cm and 9cm graticules as measuring points), but I think it should be better. I did replace V161 with a NOS 6CL6 which improved the situation here compared to the tube that was in there.