My understanding of the theory is that isolating the voltage ref was going to reduce noise associated to thermal variation inside the case. And that it look like an oversight from the designers (Solartron). That's why I was pointing to the clock which is the best way of reducing noise.
Now it's an interesting theory and it probably worth testing over multiple days.
In any case, I think doing maintenance on old equipment to reach original specs is one thing. Improving the original design is something else. I'm pretty sure the engineers at Solartron were a lot more competent than me on that topic.
I have to say that I'm in no way suggesting that I know better than the highly competent engineers at Solartron, far from it! Having seen how stable my 7075 is when measuring the virtually noise free Weston cell, I hoped the 7061 would be just as good but it seems to have much more noise.
You're absolutely right that some proper data logging with all the environmental data included would show exactly how good or bad the 7061 is. There's no substitute for some proper science!
I'm sure you'll agree though, from looking at the thermal image of the warmed up meter, that it's unlikely there is a stable thermal environment inside the case. Convection currents will roll around the inside constantly due to the uneven distribution of heat from the transformer. The fact that the reference is mounted high off the board on relatively long leads would suggest that local air currents could cause thermocouple effects.
I don't see any noise specifications in the Solartron manuals, so I have nothing to judge what is acceptable for the 7061. I know that the 7075 is about twice as accurate when comparing specs, even though it only has 14 million counts compared to the 7061's 21 million counts. Even so, I was surprised when I saw how it behaved with the Weston cell.
It may be that my suspicions are unfounded, but having the reference mounted high up off the board in a cool corner of a case with uneven internal temperature looks like a noise source to me. As you know, the 70[6,7,8]1 series meters all have an auto zero function which runs every 15 minutes to "correct for thermal drift" as the manual says. I think it's interesting that the older 7075 with its ovenised reference diode does not have this function and manages to deliver incredibly stable measurements to the microvolt level without it. Maybe there's some connection?
One other tool I'll be building to help look at the noise levels on all my precision meters is a xdevs.com KX LTZ1000A reference. I have all the parts now thanks to Mouser's incredible service. All I'm waiting for is the PCBs from OSHPark and some bismuth/tin super low melting point solder paste from China.
Rest assured that I won't let this remain as an unsupported assertion. Once my logging system is working I'll get the meter connected to my Weston cell for some initial testing with and without a tea cosy for the diode. I fear that the temperature sensitivity of the Weston cell may make the results hard to interpret, but if they do, the KX reference should make things clearer.