Ok then, SORRY mnem I missed that post
My take on the cal mode is here.
1. Briefly press either up or down key to move on the desired preset temperature P1 to P5 etc.
2. Set 380 oC, select P4 profile 400oC press down key for 3 seconds to enter temp adjust mode, then press down button again till display reads 380oC. To increase current temp use the up button till required temp is displayed. This is a temporary adjustment only.
3. To cal the tip, do the following, let’s say that P2 (pre-set at 300) reads on your calibrator as 280oC, select P2 and press both up and down keys together and digits will flash. Press the up key till the display reads 20 and then press both keys together to save the setting which should add 20oC to the stored setting of 300 but in reality is 280 as displayed on your tester.
4. Check the temp again on your calibrator and see if it matches the displayed setting of 300oC.
5. If not repeat the step 3 and use either the up or down key to add or subtract to the stored preset.
6. Check again and repeat until it is correct. Repeat process for other pre-sets if required.
This may or may not off is be right but its my educated guess that it is. Steps 1 and 2 is a temporary temp change only and once switched off is not saved.
A copy of the instructions is below (downloaded pdf file.)