Author Topic: Is there a danger to shortening a tracking generator of a spectrum analyzer?  (Read 1220 times)

0 Members and 1 Guest are viewing this topic.

Offline lausTopic starter

  • Contributor
  • Posts: 25
  • Country: us
I would like to do sweeps with the spectrum analyzer on a voltage balun.  Checking with a multimeter at DC there is a short between all points.  Looking at the schematic it is obvious why.

Do I need to worry that the tracking generator is being short circuited to ground at some frequency or are spectrum analyzers designed to handle shorts?

I'm using a Siglent SSA3021X and have done some sweeps on a home made common mode choke to see the signal vs common mode response.  Now I would like to compare that to a commercially made voltage balun. But I'm hesitant about breaking the spectrum analyzer.
 

Offline aristarchus

  • Regular Contributor
  • *
  • Posts: 107
  • Country: 00
Your balun provides an impedance to the tracking generator output, thus it always has a resistive component at a given frequency.
Regardless, I guess that the tracking generator output can withstand shorts if not continuously at least for brief periods, after all their power is in the range of 0dBm.

Better ask @tautech though.
 
The following users thanked this post: laus

Online coppice

  • Super Contributor
  • ***
  • Posts: 9238
  • Country: gb
Most engineers are pretty clumsy oafs. If instruments are not designed so their low power output signals can tolerate considerable abuse, they'll be sent for repair a lot and get the product a bad reputation for being fragile. Even a lot of high power outputs are designed to detect overload and shut down quite aggressively. Don't let this encourage you to be sloppy, though. No circuit is so well protected that you can't find a way to blow things up.
 
The following users thanked this post: laus

Online Wolfgang

  • Super Contributor
  • ***
  • Posts: 1812
  • Country: de
  • Its great if it finally works !
    • Electronic Projects for Fun
I would like to do sweeps with the spectrum analyzer on a voltage balun.  Checking with a multimeter at DC there is a short between all points.  Looking at the schematic it is obvious why.

Do I need to worry that the tracking generator is being short circuited to ground at some frequency or are spectrum analyzers designed to handle shorts?

I'm using a Siglent SSA3021X and have done some sweeps on a home made common mode choke to see the signal vs common mode response.  Now I would like to compare that to a commercially made voltage balun. But I'm hesitant about breaking the spectrum analyzer.

Not even a Siglent should be blown up by this. A tracking generator in a spectrum analyzer or a VNA port must be able to work with *all* passive impedances.
 
The following users thanked this post: laus

Online TimFox

  • Super Contributor
  • ***
  • Posts: 8160
  • Country: us
  • Retired, now restoring antique test equipment
Tracking generators normally achieve 50 ohm source impedance with a suitable series resistor from the source.  You can check this by connecting an external voltmeter (or similar) with and without 50 ohm termination:  unterminated should give 2x voltage, or 6 dB more on a voltage measurement.  The source should be capable of withstanding a short after that resistor.  To blow up a tracking generator or similar output from a signal generator normally requires feeding "real" power from an external source by accident.  Power amplifiers, however, can be damaged by improper load impedance, and should use protection circuits to avoid smoke.
 

Offline tautech

  • Super Contributor
  • ***
  • Posts: 29160
  • Country: nz
  • Taupaki Technologies Ltd. Siglent Distributor NZ.
    • Taupaki Technologies Ltd.
To blow up a tracking generator or similar output from a signal generator normally requires feeding "real" power from an external source by accident.
Yep, this ^
Stated clearly in the SSA3000X datasheet:

TG Output maximum reverse level Mean power: 30 dBm, DC: ±50 VDC
Avid Rabid Hobbyist.
Some stuff seen @ Siglent HQ cannot be shared.
 
The following users thanked this post: laus

Offline lausTopic starter

  • Contributor
  • Posts: 25
  • Country: us
Thanks everyone.  I went ahead and did the measurements.  Nothing bad has happened.  :-+

Tracking generators normally achieve 50 ohm source impedance with a suitable series resistor from the source.  You can check this by connecting an external voltmeter (or similar) with and without 50 ohm termination:  unterminated should give 2x voltage, or 6 dB more on a voltage measurement.  The source should be capable of withstanding a short after that resistor.  To blow up a tracking generator or similar output from a signal generator normally requires feeding "real" power from an external source by accident.  Power amplifiers, however, can be damaged by improper load impedance, and should use protection circuits to avoid smoke.

I tried to follow your instructions and do a voltage measurement of the tracking generator output with a digital multimeter and oscilloscope without much success.  The voltage fluctuates constantly.  I guess due to the TG sweeping across the frequency range.  I didn't have any more luck when reducing the frequency range or set it to a fixed frequency.  I just couldn't get any meaningful reading.
I probably just didn't setup things properly to do the measurement.
 

Online TimFox

  • Super Contributor
  • ***
  • Posts: 8160
  • Country: us
  • Retired, now restoring antique test equipment
You may be seeing the TG output "mute" during the retrace from end frequency back to stop frequency.  Otherwise, that would be strange--can you slow down the sweep to a very slow sweep time?
 