Very good, so I also have the opportunity to add my repair case.
I bought my unit 1 year ago as defect.
The Ohm range was not working, repair description here:
https://www.eevblog.com/forum/repair/hp-34401a-error-612-613-615-617-618-619-621/msg3571485/#msg3571485The unit was sensitive to tapping on its case; it made a reset as if I would change its cards.
I also found out, that all of the clamps on the mains plug were quite loose. I assumed brief disruption of power, or that the instrument is sensitive to arcing on these contacts, or similar EMC noise.
I pulled all clamps and tightened them carefully with pliers. Still the clamps were not very well seated, as they seem not to fit correctly to the mains jacks contacts. Anyhow, all connections felt more rigid now, and the tapping sensitivity seemed to be much better, although it still happened from time to time.
The battery was depleted, so I cut it off and soldered another one on the remaining pins. Clock and memory were fine again.
![](https://www.eevblog.com/forum/testgear/my-34970a-keeps-rebooting-as-soon-as-i-power-it-on/?action=dlattach;attach=1537162;image)
I also replaced the firmware from faulty V9 to V13, and nearly forgot the problem.
![](https://www.eevblog.com/forum/testgear/my-34970a-keeps-rebooting-as-soon-as-i-power-it-on/?action=dlattach;attach=1537168;image)
Last week, when I used the instrument in my hot room upstairs and not in my cool basement lab, this problem re-occured in a very frequent manner, so I first disconnected the DMM, the backplane, disassembled the instrument completely and by tapping on the boards in the end localized the fault to be somewhere on A1, the main controller board.
Power supplies, regulators and capacitors were ok, and by chance I found the culprit. The solder joint of the batteries minus pole made a loose connection.
The reset and battery supervisor IC, ADM691, also might create resets, when the battery is changed.. that's a vague hint from the data sheet.
![](https://www.eevblog.com/forum/testgear/my-34970a-keeps-rebooting-as-soon-as-i-power-it-on/?action=dlattach;attach=1537174;image)
The assembly problem is that the battery weighs several grams, but the diameter of the via and the solder pad for this single minus pole is very tiny, and the battery probably has to be hand-soldered. As one does not want to heat up the battery, the soldering process might have been too short, so that not enough solder was sucked into the via to the top side, for making a good mechanical connection.
After 20 years, and by the higher temperatures and vibrations, this created a solder joint crack.
So I made a proper solder joint by feeding some excess solder, maybe over heating the battery a bit.
I also changed R183, from 8k25 to 300k, as the battery type I'm using is not intended for recharging.
![](https://www.eevblog.com/forum/testgear/my-34970a-keeps-rebooting-as-soon-as-i-power-it-on/?action=dlattach;attach=1537204;image)
The instrument works rock stable now.
Frank