Author Topic: OWON XSA1032-TG 3.2G Spectrum Analyzer  (Read 6285 times)

0 Members and 1 Guest are viewing this topic.

Offline cncjerry

  • Supporter
  • ****
  • Posts: 1308
Re: OWON XSA1032-TG 3.2G Spectrum Analyzer
« Reply #25 on: May 13, 2023, 11:44:47 pm »
Was this SA really $600?  I've seen them listed for like $2,600.
 

Offline TomCTopic starter

  • Frequent Contributor
  • **
  • Posts: 670
  • Country: us
Re: OWON XSA1032-TG 3.2G Spectrum Analyzer
« Reply #26 on: May 14, 2023, 03:02:29 am »
Was this SA really $600?  I've seen them listed for like $2,600.
Yes, I couldn't believe it was legit when I saw the prices they were selling for at Walmart's marketplace. Several chinese vendors had them listed anywhere between $425 and about $2600. See reply #12 where I posted the receipt:

https://www.eevblog.com/forum/testgear/owon-xsa1032-tg-3-2g-spectrum-analyzer/msg4854467/#msg4854467
 


Offline DaneLaw

  • Frequent Contributor
  • **
  • Posts: 549
  • Country: dk
Re: OWON XSA1032-TG 3.2G Spectrum Analyzer
« Reply #28 on: May 14, 2023, 10:21:48 pm »
These Amazon sellers linked above, dang 
-
Detailed Seller Information
Business Name: yibinxihongshishipinyouxiangongsi
Business Address:
gaoxianqingfuzhenjixiangxiang84hao
yibinshi
sichuansheng
645154
CN
-
Detailed Seller Information
Business Name: putianchengxiangfutaishangmaoyouxianzerengongsi
Business Address:
chengxiangquxialinjiedaolihuadongdadao8hao
putianwandaguangchang11haolou2ti805shi
putianshi
fujiansheng
351199
CN
-
Detailed Seller Information
Business Name: Hong yu guang zhou shang mao you xian gong si
Business Address:
Tian he qu hua guan lu 1934hao 905fang zi bian B07
Guang zhou shi
Guang dong sheng
510665
CN
-


- but then again, if you're in the US, and using Jeff Jorgensen's big A-store, you are likely pretty covered for not being sc3wed over, ain't it like 20 or 25%  Amazon takes when selling on their platform.
But it seems it's no dice, outside the US, as it won't even show the price, if your region aint in the US, you have to manually put in a US postal code, and no' I definitely don't put in 90210 hence some 90s series' like its the only US postalcode I remember from scratch. 
but with OPs mentioned price at sub 600 in mind, it seems like a decent SA, not least because it's the 3.2GHz variant.
are the lesser XSA10XX models. upgradeable, or how does Owon role, or does the hardware differs.?.

 

Offline TomCTopic starter

  • Frequent Contributor
  • **
  • Posts: 670
  • Country: us
Re: OWON XSA1032-TG 3.2G Spectrum Analyzer
« Reply #29 on: May 14, 2023, 10:54:48 pm »
I believe there are differences in the hardware. At least, for other OWON devices I own, that I've opened to look inside, that's been the case. I get what you are saying, about the non-traditional vendors, If I hadn't had some guarantee that the major retailer would have my back, I wouldn't risk dealing with them. Not because they are necessarily dishonest, but without a track record it's impossible to be comfortably sure. :)
 

Offline JeremyC

  • Regular Contributor
  • *
  • Posts: 148
  • Country: us
Re: OWON XSA1032-TG 3.2G Spectrum Analyzer
« Reply #30 on: May 15, 2023, 12:11:37 am »
I still see some places where they are selling this SA for less than MSRP, not quite as inexpensive as what I got a few months ago, but under $1000. Funny thing is that these examples at Amazon call it a handheld, in reality, it weights about 11 pounds (5kg). Maybe they made too many of them and let a batch go to non-traditional vendors dirt cheap. Who knows? :o

https://www.amazon.com/Oumefar-Precision-Generator-Resolution-regulations/dp/B08QJQX4BD/ref=sr_1_86?crid=S55K50VOUZ7C&keywords=XSA1032&qid=1684086658&sprefix=xsa1032%2Caps%2C146&sr=8-86&ufe=app_do%3Aamzn1.fos.5137e923-c7be-4142-979c-7c68b6c26f63

https://www.amazon.com/Portable-Tracking-Generator-Precision-regulations/dp/B08QJR3KFR/ref=sr_1_85?crid=S55K50VOUZ7C&keywords=XSA1032&qid=1684086658&sprefix=xsa1032%2Caps%2C146&sr=8-85

https://www.amazon.com/Portable-Generator-Resolution-Precision-regulations/dp/B093CZJ3ZH/ref=sr_1_84?crid=S55K50VOUZ7C&keywords=XSA1032&qid=1684086658&sprefix=xsa1032%2Caps%2C146&sr=8-84&th=1

https://www.amazon.com/Portable-Tracking-Generator-Resolution-regulations/dp/B08QJQ54XW/ref=sr_1_83?crid=S55K50VOUZ7C&keywords=XSA1032&qid=1684086658&sprefix=xsa1032%2Caps%2C146&sr=8-83&th=1

This entire thread looks to me like a SCAM!
Devices at Amazon in these links have the OWON brand grayed out from the label, take a close look!
In the initial post https://www.eevblog.com/forum/testgear/owon-xsa1032-tg-3-2g-spectrum-analyzer/msg4848032/#msg4848032 it is the same, there OWON is grayed out from the label.

Compare those pictures with https://www.testequipmentdepot.com/owon-xsa1032-tg-9khz-32-ghz-spectrum-analyzer-with-tracking-generator-kit.html

« Last Edit: May 15, 2023, 12:17:05 am by JeremyC »
 

Offline TomCTopic starter

  • Frequent Contributor
  • **
  • Posts: 670
  • Country: us
Re: OWON XSA1032-TG 3.2G Spectrum Analyzer
« Reply #31 on: May 15, 2023, 04:34:40 am »
To all the colleagues that have been following this thread. The attachments are photos of the equipment in my lab that I've been using to conduct the experiments and tests described in previous posts. I hope this satisfies any curiosity or suspicion of any type of nefarious purposes :phew:. With that out of the way, I would like to continue sharing with like-minded electronics enthusiasts, my experiences, discoveries, and doubts, as I try to learn and attempt to master new concepts and equipment  :-\.
 
The following users thanked this post: DaneLaw, nenea dani

Offline TomCTopic starter

  • Frequent Contributor
  • **
  • Posts: 670
  • Country: us
Re: OWON XSA1032-TG 3.2G Spectrum Analyzer
« Reply #32 on: May 15, 2023, 09:20:07 am »
Residual FM Test

I've been curious about this spec since it isn't even mentioned on the XSA1032 manual or datasheet. So I got some information on what it means and how to test it (attachments 2 and 3), but I'm not sure I fully understand the steps. Particularly, when they say to use the peak-to-peak search function. I experimented with Peak Max Search and Peak Min Search, but it doesn't give consistent results. So finally I decided to instead use a Delta Marker spanning the full Sweep time. This is  more stable, especially after I increased the Sweep time to 400ms, but the dB reading still constantly changes, although more gradually. For the calculations I picked the highest number I saw after a few minutes, but again, it's possible that this is not much better than a random number, since I'm not, or don't know how to follow the instructions to the letter.

For the Demodulation Sensitivity I'm using the values from attachment 4:

   2800/18.49 = 151.4 Hz/dB

Notice that I'm using an RBW of 5kHz, even though the specs for the Rigol SA calls for 1kHz. I had to go with a minimum of 5kHz, because on this SA any less than that causes the Zero Span function to gray out, and the next steps calls for setting the SA to Zero Span.

For the Frequency Deviation I'm using the setup shown on attachment 5, and waiting a couple minutes while eyeing the dB value to record the highest value that pops-up:

   0.19 dB

So the residual FM comes to: 151.4 x 0.19 = 28.8 Hz

Of course, this number is more likely than not off. Even if the procedure I used turns out to be close enough to what's required, the Residual FM of the signal generator I'm using (Windfreak SynthUSBII) is not specified. So, as stated in the instructions, it may influence the final result.

I'd be grateful if anyone with better knowledge of this test helps clarify the instructions!
 
The following users thanked this post: nenea dani

Offline TomCTopic starter

  • Frequent Contributor
  • **
  • Posts: 670
  • Country: us
Re: OWON XSA1032-TG 3.2G Spectrum Analyzer
« Reply #33 on: May 18, 2023, 08:13:05 am »
Second Harmonic Distortion Test Part 1

The following is my interpretation of this test after reading a number of articles regarding this subject, hope I've got it right:

This test is used to determine the dBc value of the second harmonic generated by the SA while a fundamental signal is being applied at the SA's input. To get a valid reading the signal generator should be capable of producing a signal nearly free of harmonics, particularly the 2nd harmonic. To help achieve a harmonic free signal before applying it to the SA's input, an appropriate low pass filter can be used between the generator and the SA. Ideally, the low pass filter would only block the harmonics, particularly the 2nd harmonic.

The SA manufacturer's specs may be given as a dBc value when the specified fundamental is at the specified dBm level, for example, on the XSA1032 the Second harmonic distortion is specified as -65dBc when a -10dBm harmonic free ≥ 50 MHz fundamental is applied to the SA's input. This means, that although the input signal contains no harmonics, up to a -65dBc 2nd harmonic may appear on the SA's screen that was solely generated by the SA's internal circuits.

Alternatively, the SA manufacturer's specs may be given as an SHI value when the specified fundamental is at the specified dBm level, for example, on the DSA832E the Second Harmonic Intercept (SHI) is specified as +40dBm when a -20dBm harmonic free ≥ 50 MHz fundamental is applied to the SA's input.

The SHI value is a theoretical dBm level at which the fundamental's and 2nd harmonic's dBm levels coincide. Normally, this can't be achieved in practice. However, is a known fact, that as the fundamental's dBm level increases the 2nd harmonic dBm level also increases with a 2:1 ratio. Using this fact, the level where both signals would theoretically coincide can be easily calculated mathematically or graphically. I added information to the original XS1032 & DSA832E specs showing how to mathematically convert an SHI to a dBc value or a dBc value to an SHI, see attachment 1.

   Attachments 2 and 3 comprise one of the more relevant articles I read regarding this subject.

   Attachment 4 is graphical representation I put together illustrating how the XSA1032 & DSA832E SHI's can be obtained graphically.


I don't have a generator that meets the requirements necessary to properly evaluate the XSA1032 Second Harmonic Distortion, but on the second part I'll post the results obtained with the generators I have.
« Last Edit: May 18, 2023, 09:02:36 pm by TomC »
 

Offline TomCTopic starter

  • Frequent Contributor
  • **
  • Posts: 670
  • Country: us
Re: OWON XSA1032-TG 3.2G Spectrum Analyzer
« Reply #34 on: May 19, 2023, 07:20:14 am »
Second Harmonic Distortion Test Part 2

To test the 2nd Harmonic Distortion on my XSA1032-TG, first I'm going to use my SynthUSBII signal generator. Attachment 5 shows the harmonic content specs for the main IC chip used in this generator. Since this generator's output is a square wave, the 3rd harmonic spec is a larger value than the second. But even the second harmonic spec is larger than the value required to get a valid reading of the SA's harmonic distortion. To perform the test the generator's output was set to 50MHz at approximately -10dBm (the output level is not calibrated).

For Attachments 6 & 7 the SA's span was set to 160MHz to allow to see the relative values of the fundamental and the 2nd & 3rd harmonics. The 2nd harmonic should be at marker 2, but with the current settings the noise floor is too high for it to be visible. I can't lower the noise floor because the SA won't let me set the Ref Level and the Attenuation to the values required for the test.

For attachment 8, to zero in on the 2nd harmonic, the SA's span was set to 100kHz and the center frequency to marker 2. Here the 2nd harmonic is visible because the noise floor is much lower. Since only a small signal is in the sweep's range, I was able (allowed by the SA) to set the Ref Level and the Attenuation to the values required for the test. The 2nd harmonic dBm value is around -62, since the fundamental is around -9dBm, the dBc value comes to -62 -(-9) = -53dBc. This is not even close to the -65dBc cited on the XSA1032-TG specifications, but that requires that the input signal be nearly harmonics free, which it isn't in this case.

Next I want to test the 2nd Harmonic Distortion on my XSA1032-TG using my AG1022F AWG. Attachment 9 shows the harmonic distortion specs for this generator which are much better than the ADF4351 specs. However, I'll have to perform the test with a 25MHz fundamental instead of the ≥ 50 MHz that the XSA1032 specs call for. Unfortunately, 25MHz is the highest frequency that my AG1022F can output. To perform the test the generator's output was set to 25MHz at 200mVpp. On a 50-ohm system 200mVpp is equivalent to around -10dBm.

For Attachments 10 & 11 the SA's span was set to 100MHz to allow to see the relative values of the fundamental and the 2nd & 3rd harmonics. The harmonics should be at markers 2 & 3, but with the current settings the noise floor is too high for them to be visible. I can't lower the noise floor because the SA won't let me set the Ref Level and the Attenuation to the values required for the test.

For attachment 12, to zero in on the 2nd harmonic, the SA's span was set to 100kHz and the center frequency to marker 2. Here again the 2nd harmonic is visible because the noise floor is much lower. Again, since only a small signal is in the sweep's range, I was able (allowed by the SA) to set the Ref Level and the Attenuation to the values required for the test. The 2nd harmonic dBm value is around -70, since the fundamental is around -10dBm, the dBc value comes to -70 -(-10) = -60dBc. This is closer to the -65dBc cited on the XSA1032-TG specifications, but still, the input signal is not nearly harmonics free as required, perhaps if I use an appropriate low pass filter I could get closer to a valid reading. Unfortunately, I don't have one at this time.

 
The following users thanked this post: nenea dani

Offline TomCTopic starter

  • Frequent Contributor
  • **
  • Posts: 670
  • Country: us
Re: OWON XSA1032-TG 3.2G Spectrum Analyzer
« Reply #35 on: May 20, 2023, 10:08:02 am »
Third-order intermodulation - Intercept (TOI) Test

I read several interesting articles regarding this subject and found that although different looking formulas are used to derive the TOI, they are just that, different looking. In essence, they are all describing the same thing. Attachments 1-3 are a sampling of the most relevant articles I used as reference.

The SA manufacturer's specs for the test are in attachment 0, the test description that I used to perform the test is on attachment 4. The manufacturer's specs call for a ≥ 50 MHz two tone input at -20dBm with a 100kHz separation between the tones (signals). I couldn't use 50MHz since I'm using my AG1022F generator that maxes out at 25MHz. This is a dual output generator, so I set channel 1 to 25MHz and Channel 2 to 24.9MHz.

The test instructions (attachment 4) call for a "Power Coupler", also known as a combiner, to connect the generators to the SA. I don't have a combiner, instead I used a BNC T coupler and two cables to make the connection (see attachment 6). This is not optimal but is allowed for lower frequencies (although 25MHz isn't that low). To keep the 50-ohm system as intact as possible, I changed the setup of the generator's outputs to 25-ohms. This should result in a 50-ohm combined output to match the SA's 50-ohm input.

Attachment 5 is a screen from one of my favorite dB calculators. It shows that for a -20dBm output about 63mVpp are required. I started out my generator's outputs at that level, but to get the -20dBm required by the specs I had to increase the level a little, -70dBm on channel 1 and -67dBm on channel 2.

Attachments 7 and 8 show the spectrums of the two fundamentals and the intermodulation products 2f₂-f₁ and 2f₁-f₂ as indicated in the test instructions. There are other intermodulation products, but these ones are the closest to the fundamentals and for that reason the most problematic (can't be easily filtered) when present on a DUT. From the marker readings the TOI can be calculated. In the attachments, as I mentioned before, there are a number of different looking formulas for this, but they all give the same answer. So I'm going to use the one in the test Instructions.

   TOI = Rl + [Ir₃/2]

   Rl is the power level of the fundamentals, -20dBm.

   Ir₃ is the distance between either one of the fundamentals and largest (smallest level) of the two intermodulation products (IMDs). I'm going to round that to 50dBM.

   So we have -20 + [50/2] = +5dBm

Well, that's half the +10dBm that we are supposed to have according to the specs but given that the test conditions were far from optimal, I think I'll regard it as a pass!
 

Offline TomCTopic starter

  • Frequent Contributor
  • **
  • Posts: 670
  • Country: us
Re: OWON XSA1032-TG 3.2G Spectrum Analyzer
« Reply #36 on: May 22, 2023, 09:51:07 am »
1 dB Gain Compression Test

I wasn't able to perform this test as specified. The main reason, as I see it, is that the SA has safeguards that prevent the user from manually setting certain parameters. As I understand it, the test calls for 0dBm attenuation and a ≥ 50 MHz input signal that shouldn't show the 1dB compression, nominally, before it's level reaches +2dBm (see attachment 1). So to see the signal, the reference level has to be higher than at least +2dBm, but the SA won't let me set the attenuation to 0dB unless the reference level is set a lot lower, -20dBm or lower.

In addition, although there is an abundance of information for performing this test on an amplifier, I couldn't find specific information on using the test to characterize the SA itself. The Fluke publication I've been using as a guide for other tests, has a write-up on a "Compression Test", but is a different test that deals with compression of an adjacent smaller signal when a larger signal is present. Attachment 2 is the complete Fluke pub in case some of you want to read it. Attachment 3 is a write up I found that in my opinion comes the closest to explaining the procedure and reasoning behind the test I wanted to perform.

The rest of the attachments deal with the procedure I envisioned to perform the test. Attachment 4 is the conversion table I used to set my generator at the appropriate dBm levels, my AG1022F only accepts pp values or rms values. Attachments 5 and 6 show the 25MHz fundamental at dBm values of 1 to 5dBm, I used different Traces and scattered them on the screen for better visibility by changing the Center Frequency before doing the Max Hold on each trace. The one thing I couldn't do is set the Attenuator to 0dB.

With the 25dB attenuator setting I didn't expect to see any compression, but if you look closely at the last two peaks, it appears as if compression is starting to creep in. I thought maybe this meant that there was some kind of non-linearity in the display, so I changed the reference level to 8dBm to see if there was a difference (attachments 7 and 8 ). But no, it's the same. I hooked up my DSO to check the signal mVpp values and it agreed with the SA. Seems that either there is loss on the cables, or my generator is not quite outputting the calibrated values.

So that's the best I could do! But please, if any one of you has any insights or opinions on how to do this test correctly, post it.
« Last Edit: May 22, 2023, 09:53:42 am by TomC »
 
The following users thanked this post: nenea dani

Offline TomCTopic starter

  • Frequent Contributor
  • **
  • Posts: 670
  • Country: us
Re: OWON XSA1032-TG 3.2G Spectrum Analyzer
« Reply #37 on: May 25, 2023, 09:01:08 am »
Resolution Bandwidth Test Part 1

I found this test very interesting. Besides the test specs and test instructions (attachments 1 and 2), I read an articles that compares the performance of the analog filters used on older SAs, to the current digital filters (Attachment 3). On Part 2, I'll try to duplicate the example showing how the new digital filters allow the spectrums of very close adjacent carriers of different amplitudes to be viewed and analyzed and do so with much faster repetition rates than what could be achieved with the older generations.

The spec identifies the different RBW filters available (10Hz to 500kHz (1-10 steps by sequence), 1MHz, 3MHz). I did test the selectivity of each one of these filters, but I've only attached a sampling of the spectrums I obtained to illustrate the main things I found (Attachments 4-9). To check the "Bandwidth Selectivity", I didn't follow the Instructions on attachment 2 to the letter because this SA has an NdB Marker Function that allows the bandwidth at the dB level of interest to be determined automatically.

According to the spec, the "Resolution Filter Shape Factor (60 dB : 3 dB)", this is what the instructions call the "Bandwidth Selectivity", should be "<5: 1 typical". The instructions show that this is calculated using: f60dB / f3dB, for this to be less than 5, the 60dB bandwidth can't exceed the 3dB value by more than 5 times. It also states that the typical ratio is 1. I'm afraid that I never saw 1 on any of my measurements, 1 would mean that the 60dB and 3dB bandwidths are the same. As far as the ratio being <5, that checked out fairly good with some caveats. The details are as follows:
  Edit: I didn't look at the punctuation close enough on "<5:1 typical", It's a colon, not a semicolon as I
           thought, so it claims the ratio is <5:1 typical, which is OK, not 1 typical as I mistakenly interpreted it.

   RBW = 10Hz: Attachments 4 and 5 show the 3dB and 60dB bandwidths. 143 / 10 = 14.3 not <5. So for RBW=10Hz the selectivity
   is not as good as stated. With the span   set to 500Hz, we have 50Hz per division. Notice that about half way between -60dBm
   and -70dBm the bandwidth is about 50Hz, that would be the 35dB bandwidth of the filter and the point where we actually
   have a 5 ratio. So the filter can still resolve different frequencies as specified but only for larger signals.

For all the remaining spectrums I'm not showing the 3dB bandwidth because I found out it's the same as the RBW setting or very very close. The next few possible RBW setting have an improved signal shape, and are closer to the stated spec, but still somewhat off. Starting with RBW = 50Hz specs are met for several possible settings.

   RBW=50Hz" Attachment 6 shows the 60dB bandwidth, the 3DB bandwidth is the same number as the setting (50dB). 247 / 50 <5.

For the settings prior to 300Hz the specs are met, but starting at 300Hz the noise floor starts to get in the way. To get a 60dB reading with the current settings, the noise floor has to be less than -90dBm. But right at RBW = 300Hz this is no longer the case, and it gets worse from this point on, since the noise floor gets higher with every further increase in RBW.

   RBW = 300Hz: Attachment 7 shows the 60dB bandwidth, the 3dB bandwidth is 300, 1700 / 300 = 5.66, not quite <5.

Even though the 60dB bandwidth can't be accurately determined on subsequent RBW settings due to the noise floor, the shape of the filter is still good, and if the bandwidth is measured just above the noise floor, the ratio between this higher level bandwidth and the 3dB level bandwidth is still <5. Attachments 8 and 9 are examples of this situation.
« Last Edit: May 26, 2023, 03:07:41 am by TomC »
 
The following users thanked this post: DaneLaw

Offline TomCTopic starter

  • Frequent Contributor
  • **
  • Posts: 670
  • Country: us
Re: OWON XSA1032-TG 3.2G Spectrum Analyzer
« Reply #38 on: May 26, 2023, 06:26:22 am »
Resolution Bandwidth Test Part 2

To get some idea of how the different RBW filters perform in real live, here i'm trying to recreate the example described on the "Agilent AN 1318 Optimizing Spectrum Analyzer Measurement Speed" article I attached to the previous post. Namely, to check the SA's performance when trying to visualize two CW signals that are 240Hz and 20dBs apart. In the article they use signals around 1 GHz at -35dBm and -55dBm, due to equipment limitations I'm using around 25MHz at -30dBm and -50dBm.

Attachment 10 contains the spectrums shown in the article as displayed by an Agilent 8563E SA at RBW=30Hz and RBW=100Hz with a 5kHz span. The equivalent spectrums on my XSA1032-TG are attachments 14 and 15.

Attachments 11 to 13 show the same signal using RBW=10Hz, here I tried span 500Hz, as well as the 5kHz span they used on the article. Notice that even at RBW 10Hz, the sweep time is nowhere near the 16.7 seconds it takes for RBW=30Hz at 5kHz span when using the older Agilent 8594E (Figure 5 in the "Agilent AN 1318 ..." article).

Attachments 16 and 17 show the same signals using the RBW=30Hz and 100Hz used in the article, but using a 1kHz span instead of the 5kHz span. I included these because it seems that the Delta measurements are a bit more accurate at this setting.
« Last Edit: May 26, 2023, 06:34:24 am by TomC »
 