It is coming along great. I got it all apart and it did take some work to get those shields off. I cleaned up the switch decks with an eraser and then sprayed them with the deoxit afterwards. I went ahead and sprayed the other pots that had access as well. I went through the calibration steps and all is going well so far except:
The timebase talks about VR801/VC801, but my PCB as three adjustments VR803, VR804, VR805 instead:
VR803 is for the fast timebases
VR804 begins at 50us
VR805 begins at 5ms
These were not too hard to calibrate.
I had a problem on step 13 of the trigger adjustment - is talks about adjusting VR309 to get 0V on the collector of Q610, but my meter shows +23VDC. Changing VR309 does not do anything. Then you repeat it for CH2, changing VR310. Still not working - still +23VDC. Any ideas?
ALSO - the corner where the transformer is gets quite warm without the case on - I wonder if it gets as warm with the case in place.
The scope is working so much better now...