I may have found a problem that could lead to the scope turning on by itself. The ground is floating for some strange reason. I was getting all sorts of AC in the signals and could not figure out why. I measured 57 VAC from scope ground to bench ground! not good at all.
I figured maybe the last owner of this thing may have a loose ground wire in the chassis, so I have taken it apart. The AC input is (not surprising) in a fully shielded housing that is soldered to the PCB with a big piece of sheet metal soldered around it. Essentially, and all day affair to gt to it. I am, however, able to get probes on it. And there is no connection at all from the GND pin on the AC jack to the GND pin on the PCB side. Not even high-impedance, nothing.
Not sure how or why this would be. The question now is would it be just fine if I connect an earth wire to the common bench ground to 'fix' this. Also, anyone have an idea what would have caused the GND to be disconnected. I can only think it would be a manufacturing flaw of some sort.