Never use it.
Can't a startup script be written to only act after a certain number of days that isn't another process running?
That's exactly what the Windows scheduler is for. It's not worth optimizing for a single clearing of registry keys every 15 days.
Software by day, hardware by night; blueAcro.com
Offline swperk
« Reply #326 on: July 04, 2014, 03:57:56 pm »
I just updated mine from 1.08 to 1.09. The upgrade went quickly and without a hitch, and all the options remain intact.
Model: DSA815
Serial Number: DSA8A1536XXXXX
Version of Main Board: 00.04
Version of Radio Frequency Board FPGA: 00.05
Version of Digital Board FPGA: 00.04
Version of Firmware: 00.01.09
Version of Boot: 00.01.03
Offline videobruce
« Reply #327 on: July 04, 2014, 04:40:33 pm »
Who said anything about "optimizing"??
swperk:
Any idea what was changed or fixed?
Offline sv1dpp
« Reply #328 on: July 04, 2014, 07:23:46 pm »
Hi all,
after firmware update to 1.09, the device did not BOOT displaying the message LDR failed boot error.....So I had to revert to the previous FW 1.08 booting from a USB flash disk to return the device in normal contition. Maybe the new FW 1.09 has something wrong.
Regards
Spiros
Offline sv1dpp
« Reply #329 on: July 04, 2014, 07:37:40 pm »
Hi again,
after 3 attempts the upgrade is OK to 1.09. The USB flash was wrong.
regards
Spiros
Offline Velund
« Reply #330 on: July 23, 2014, 11:20:32 pm »
Hm...
Does anyone know a safe way to save a backup copy of cal. memory to a file?
Offline 9a4wy
« Reply #331 on: July 29, 2014, 11:57:22 am »
Does anyone know how to restore Lost TG Calibration Data, or the procedure for for TG Calibration?
This is the result of Error 470/471, Calibration Memory Lost.
Any thoughts or assistance would be greatly appreciated.
   Thank you in advance for any information.
how did you enter into this problem ??
please explain...tnx
K
Offline videobruce
« Reply #332 on: July 29, 2014, 08:35:18 pm »
Here is the answer to my question about obtaining a service manual from Rigol. Is anyone aware of this situation??
Hi Bruce,
Our repair process for spectrum analyzers is in the midst of some changes. Currently repair is done overseas back at he factory. This can extend lead times to 12 weeks or more. We hope to establish domestic repair soon but wont have more detail for another month or so. It's not so much that the repair is complicated it's the calibration. So currently there are three options:
1) We can repair the scope for $250 plus shipping.
2) We can sell you a demo/refurb unit.
3) We can provide a discount on a new purchase.
Best regards,
Steve Huss
Applications Engineer
To: Steve_Huss@rigoltech.com
Date: 07/26/2014 05:42AM
Subject: Re: Case # 00004641: Service manual
If that is the case. how is it repaired??
Bruce
On Fri, 25 Jul 2014 18:32:24 -0400, <Steve_Huss@rigoltech.com> wrote:
Hi Bruce,
I think the short answer is we do not have a service manual available. At least I have not been able to track one down.
Best regards,
Steve Huss
Rigol Technologies
10200 SW Allen Blvd. Suite C
Beaverton, OR 970056
503-336-9102
Date: 07/13/2014 06:49A
Case #: 00004641
Subject: Service manual
Description: DSA1030A
Is a service manual available for this model (which is now out of warranty)?
Thanks,
Bruce
Offline 9a4wy
« Reply #333 on: July 30, 2014, 08:30:01 am »
So I just asked a couple of guys at work to 'human' translate the first line of the 1.09 changelog and it states that a change has been made to prevent hacking. Thus it appears that this version (if the unit already hasn't been riglol'ed) may likely be a scheme to prevent riglol generated licenses being installed.
Presumably new units that ship with 1.09 may be riglol 'proof' - will have to see what happens with new shipped units with the new firmware hit the street... Presumably a downgrade to 1.08 would allow for riglol to work and then upgrade back to 1.09.
Anyhow, just a possible scenario for future owners/purchasers to be aware of.
cheers,
george.
It is very well possible that also a new boot loader is put on the new units that prevent loading FW 1.08 (or older versions).
It is already impossible to load the 1.08 firmware to a 1.09 machine using the described method of the STORAGE button.
Using the PRESET button at start-up to load the 1.08  is still possible with the 1.02 boot loader.
So if you want to have a fully loaded analyser you need to be quick. I personally like the 10Hz RBW option. And would have paid for this, if it was not hackable....
Unfortunate, the 10Hz RBW is not offered by RIGOL
Did you see TG option dissapears with "downgrade" to 01.08??
posibly then install keys and then reflash it back to 01.09 but what will happen with TG option??I did'n want to do that ::)
K
Offline videobruce
« Reply #334 on: July 30, 2014, 10:07:39 am »
Note also that Rigol told us how to get into the Service/Maintenance Mode to Calibrate the DSA's internal 10 MHz Frequency Standard (with a PDF document).
So would you like to provide a link or a copy of this service literature for everyone that would like one??  :-//
Offline videobruce
« Reply #335 on: July 30, 2014, 09:33:59 pm »
It shows as a .pdf, but it downloads as a .zip that 7Zip can't open.
How about just a .pdf file?
Offline eurofox
    • Music
« Reply #336 on: July 30, 2014, 09:39:20 pm »
Just click and you download a pdf not a zip  |O
eurofox
Offline mrflibble
« Reply #337 on: July 30, 2014, 09:46:19 pm »
Oh alright ... *clickey*
Yup, confirmed. The attachment posted by ted572 is a plain PDF file. And double confirmed, yup, it's the known rigol document on adjusting the 10 Mhz osc.
Offline videobruce
« Reply #338 on: July 31, 2014, 02:21:23 am »
Well, it downloads as a .zip. The extension is getting changed.
Offline miguelvp
« Reply #339 on: July 31, 2014, 02:47:16 am »
It's you, I get the pdf file here as well.
Offline georges80
« Reply #340 on: July 31, 2014, 02:47:37 am »
Well, it downloads as a PDF and displays as a PDF for me on my win 7 laptop.
I'd suggest something is not configured correctly on your system.
cheers,
george.
Offline mrflibble
« Reply #341 on: July 31, 2014, 03:43:36 am »
Indeed. It might be something like a recently installed 7zip thingy that feels so important that it has hijacked all mime types and file extensions. Not sure though. :-//
Offline videobruce
« Reply #342 on: July 31, 2014, 04:13:43 pm »
The extension on the downloaded file (as shown in the attachemnt) is .zip, NOT .pdf
Offline HighVoltage
« Reply #343 on: July 31, 2014, 04:23:46 pm »
The file shows AND downloads to me as a .pdf file and Acrobat reader opens it automatically after download.
Offline mrflibble
« Reply #344 on: July 31, 2014, 04:39:41 pm »
Nope. HTTP headers don't lie, random windoze installs do. ;)
Relevant parts of http header when downloading that attachement:
HTTP/1.1 200 OK
Content-Type: application/octet-stream
Content-Disposition: attachment; filename*=UTF-8''DSA815%2010Hz%20Clock%20Frequency%20Adjustment%20.pdf
Mmmh, seeing that header, it could also be you are using internet explorer XYZ or something. Mickeysoft is famous for having their own ideas on how to interpret standards. Anyways, I would expect right-click: save as ... whatever.pdf to work properly in such a case.  :-//
Offline videobruce
« Reply #345 on: July 31, 2014, 04:49:35 pm »
I don't use Idiot Exploiter. I have used Opera almost from around 2000.
Offline eurofox
    • Music
« Reply #346 on: July 31, 2014, 04:57:56 pm »
Maybe an nsa or ru virus in your computer  :-DD :-DD :-DD
eurofox
Offline ted572
  • Radio Communications Equipment/System Design Engr.
Offline videobruce
« Reply #348 on: September 13, 2014, 10:17:33 pm »
ted572;
Thanks for the Guide. (Much nicer if it was a full Service Manual  :( )
« Last Edit: September 14, 2014, 11:53:02 am by videobruce »
Offline ted572
« Reply #349 on: September 13, 2014, 11:26:40 pm »
Attached is the 'DSA815 10MHz Clock Frequency (Freq. Std.) Adjustment' Guide: