However, it comes up as a 16902A
and works without issue
Interesting that it reports itself as a 16902A. The 16900A, 16902A, and 16903A were never available with the M-890 motherboard, only later versions of the 16901A, the 16902B, and 16800A models.
As far as I know normally you can only use the 16800A versions of the 16910A and 16911A analyzer modules with the 16800A models. My guess is that there is something in the analyzer module EEPROMs that identify them as 16800A versions. As your analyzer now reports itself as a 16902A with a 16911A analyzer module, I wonder if it would recognize and work with a 1675x analyzer module.
Is your 16802A analyzer module already licensed at the max 32M-Sample / 500MHz ?
This is the later Serial # 16802A , after opening it up, it was very clean, (except the destroyed HDD mount from the scrapper selling it ) after fixing the HDD mount plate ( unmangling it ) , i put in a 40gb SSD.. and its working great.
The Card when i removed it had 16911 and is identical to a 16911 i already had, but it only the base , but i already unlocked it to 32M/500mhz.
So this works on a 16802A to unlock it into a 16902A ( I don't have Option 103 ) so no touch screen, and the EXT Display Switching doesn't seem to work.
I have also tested a 16752A and it worked fine, fully detected, fully operational...
( per post # 63 from ingowien: )
[HKEY_LOCAL_MACHINE\SYSTEM\AgMachineInfo]
"AgFrameOverrideKey1"=dword:483c4aa9
"AgFrameOverrideKey2"=dword:3900384e
"AgFrameOverrideKey3"=dword:5af7983e
"AgFrameOverrideKey4"=dword:82100f65
"AgFrameModelNumberOverride"="16900"
16911A that came in the 16802A Board ID Bits: A:1,2,4,5,6 Jumpered, 3 Open, B: 1,2,3,4,6 Jumpered, 5 Open
16911A that came out of a 16903A Board ID Bits: A:1,2,4,5,6 Jumpered, 3 Open, B: 1,2,3,4,5 Jumpered, 6 Open
Board has NO identifiers that its for a 16800 Series Frame.
I wish i could figure out where this stuck bit is: ( this is for my spare 16911 card, not the one that came in the 16802 )
16911A Logic Analyzer(A) running...
Analyzer Chip Memory Bus Test running...
Slot A, Analysis Chip Data Bit Failed, nChip 0, nPort 0, exp=00000001, act=00001001
Slot A, Analysis Chip Data Bit Failed, nChip 0, nPort 0, exp=00000002, act=00001002
Slot A, Analysis Chip Data Bit Failed, nChip 0, nPort 0, exp=00000004, act=00001004
Slot A, Analysis Chip Data Bit Failed, nChip 0, nPort 0, exp=00000008, act=00001008
Slot A, Analysis Chip Data Bit Failed, nChip 0, nPort 0, exp=00000010, act=00001010
Slot A, Analysis Chip Data Bit Failed, nChip 0, nPort 0, exp=00000020, act=00001020
Slot A, Analysis Chip Data Bit Failed, nChip 0, nPort 0, exp=00000040, act=00001040
Slot A, Analysis Chip Data Bit Failed, nChip 0, nPort 0, exp=00000080, act=00001080
Slot A, Analysis Chip Data Bit Failed, nChip 0, nPort 0, exp=00000100, act=00001100
Slot A, Analysis Chip Data Bit Failed, nChip 0, nPort 0, exp=00000200, act=00001200
Slot A, Analysis Chip Data Bit Failed, nChip 0, nPort 0, exp=00000400, act=00001400
Slot A, Analysis Chip Data Bit Failed, nChip 0, nPort 0, exp=00000800, act=00001800
Slot A, Analysis Chip Data Bit Failed, nChip 0, nPort 0, exp=00002000, act=00003000
Slot A, Analysis Chip Data Bit Failed, nChip 0, nPort 0, exp=00004000, act=00005000
Slot A, Analysis Chip Data Bit Failed, nChip 0, nPort 0, exp=00008000, act=00009000
Slot A, Analysis Chip Data Bit Failed, nChip 0, nPort 0, exp=00010000, act=00011000
Slot A, Analysis Chip Data Bit Failed, nChip 0, nPort 0, exp=00020000, act=00021000
Slot A, Analysis Chip Data Bit Failed, nChip 0, nPort 0, exp=00040000, act=00041000
Slot A, Analysis Chip Data Bit Failed, nChip 0, nPort 0, exp=00080000, act=00081000
Slot A, Analysis Chip Data Bit Failed, nChip 0, nPort 0, exp=00100000, act=00101000
Slot A, Analysis Chip Data Bit Failed, nChip 0, nPort 0, exp=00200000, act=00201000
Slot A, Analysis Chip Data Bit Failed, nChip 0, nPort 0, exp=00400000, act=00401000
Slot A, Analysis Chip Data Bit Failed, nChip 0, nPort 0, exp=00800000, act=00801000
Slot A, Analysis Chip Data Bit Failed, nChip 0, nPort 0, exp=01000000, act=01001000
Slot A, Analysis Chip Data Bit Failed, nChip 0, nPort 0, exp=02000000, act=02001000
Slot A, Analysis Chip Data Bit Failed, nChip 0, nPort 0, exp=04000000, act=04001000
Slot A, Analysis Chip Data Bit Failed, nChip 0, nPort 0, exp=08000000, act=08001000
Slot A, Analysis Chip Data Bit Failed, nChip 0, nPort 0, exp=10000000, act=10001000
Slot A, Analysis Chip Data Bit Failed, nChip 0, nPort 0, exp=20000000, act=20001000
Slot A, Analysis Chip Data Bit Failed, nChip 0, nPort 0, exp=40000000, act=40001000
Slot A, Analysis Chip Data Bit Failed, nChip 0, nPort 0, exp=80000000, act=80001000
Slot A, Analysis Chip Data Bit Failed, nChip 0, nPort 0, exp=FFFFEFFF, act=FFFFFFFF
Slot A, Analysis Chip Data Bit Error, nChip 0, nPort 0, bits=00001000
...Analyzer Chip Memory Bus Test ended. Result: Failed***
...16911A Logic Analyzer(A) ended. Result: Failed***
I could recover a 16911 Card... I already replaced the Ram with no affect.