Author Topic: DIY Cold junction Compensation on a DMM6500 - 1 Kg of copper  (Read 4012 times)
Offline Kleinstein
Re: DIY Cold junction Compensation on a DMM6500 - 1 Kg of copper
« Reply #25 on: December 01, 2019, 08:10:39 pm »
The cold junction compensation is only really caring about the thermo-power from the virtual reference temperature (e.g. 0 C) and the temperature of the actual junction. Just imaging using the simulated cold junction to also measure an actual thermocouple at 0 C in parallel.
One could either use the simulation or the actual TC to 0 C.
Really important is only the thermal EMF around the cold junction temperatures, so around room temperature. The part in between is just to set the actual offset.
The piece of copper is really huge. Thermal equalization scales with the distance square, so smaller can really be faster. It should still be much faster than the water version. Symmetry can reduce the effective size to about half.
The following users thanked this post: HendriXML
Offline HendriXMLTopic starter
    • KiCad-BOM-reporter
« Reply #26 on: December 01, 2019, 11:36:27 pm »
I now understand why I was wrong. One of the starting statements above was not complete:
A temperature difference results in a voltage difference, how much is dependent on the alloy
Should be
A temperature difference results in a voltage difference, how much is dependent on the alloy and the cold junction temp.
I assumed that any same temp difference would lead to the same voltage difference. But that's not the case - which seems also more logical…
To use the translation tables one needs to know the cold junction temp even for calculating a temperature difference.
« Last Edit: December 01, 2019, 11:46:27 pm by HendriXML »
“I ‘d like to reincarnate as a dung beetle, ‘cause there’s nothing wrong with a shitty life, real misery comes from high expectations”
« Reply #27 on: December 02, 2019, 12:15:41 am »
I would like it to be smaller and of different size too, but heating up this amount may also be more fun because it asks for more max power. But as said the shape can be used to have some controlled and dampening oscillating happening in the heat up fase.
