I was in doubt if I should post it here or in the post for the SVA but sinds the calibration is still a hot topic for the upgraded SSA I decided for this forum. If I was wrong then please move or delete this post.
I came along this post:
https://www.instructables.com/id/23GHz-SMA-Calibration-Kit/
I build it on a N-connector to see how good that was only to discover that the SVA showed a poor open. My ohmmeter confirmed 50 Ohms? After close up inspection I notisted that the pin was much tinner, probebly a 75 Ohm connector.
I still think the site is good and I will try again when I have the right connectors.
Making your own calibration set is possible, but you need to keep high precision here.
In the case of the N connector there are greater distances and it should be remembered that the cable connected to the resistor must have an impedance of 50ohm (impedance not resistance). In principle, the resistor should also be non-inductive.
However, by using very short connections and a small resistor, this condition can be omitted (Can be made with small SMA connectors)...
Therefore, such a set will be easier to make using SMA connectors and using the SMA to N adapter.
Let's be more precise on this argument. It is positively possible to replicate a test kit, no doubt, therefore there are two main problems connected with DIY products:
. as already stated previously, to get precision of results... you have to have high precision, big wallet instrumentation otherwise how do you think to get a kit fulfilling reference characteristics?
. The most important problem is... repeatability. The electrical characteristics of each single component of the kit must not significantly vary with else slight variations of test conditions or result in significant wear of the parts after few test sessions.
The mechanical stability of kit's components are the big culprit. The fact that at every connection/disconnection a mechanical degradation occurs is inevitable.
Only a professional can with much experience dimension, select and couple the materials used so that either after say 50 measure sessions each component shows nearly the same electrical characteristics as at its very first run.
If a test kit shows different results under the same testing conditions, sorry, it becomes a nonsense trying to calibrate an instrument with it.
This is the real reason why serious calibration kits cost in the thousands and more, stability of results reflecting in repeatability of measures without near no divergence in values obtained.
That said, to make measures and calibrations as trustworthy as possible...:
. use as few transitions (between cables, adapters, connectors etc.) as feasible
. use good quality materials (e.g. Amphenol, Suhner etc.), no Chinese or whatever low cost material even if with very shiny and goldy look.
. no, I underline this in red, NO flexible or semi-rigid cables, use possibly rigid prebuild cables
. create a structural rigid setup connected to the ports of your instrument, possibly firmly fixed to a little tablet of RF insulating material e.g. PRFV, ceramic, PTFE etc.. I use a simple low cost porcelain stoneware tile from my last renovation and got optimal results up to and over 200Ghz.
Have precise measures.
My 10cents.
Andreas