Author Topic: Siglent SDS800X HD 12 bit DSO's  (Read 825135 times)
alvidrezluke, nsalzman and 6 Guests are viewing this topic.
Offline BillyO
Re: Siglent SDS800X HD 12 bit DSO's
« Reply #1350 on: May 07, 2024, 10:35:41 pm »
Just having a look at the PB470 "70MHz" probes that come with the SDS804X-HD.  They don't come with ground springs.  There is no excuse for this.  However I found their performance to be .. sufficient, to say the least providing a 215MHz BW.  The following screen shots were taken using an SDS2504 which has been tested to over 600MHz, so not down very much at 215MHz.
@10MHz Vpp is 1.78mV which implies -3db=1.26mV
10 MHz
2172058-0
70 MHz
2172034-1
140 MHz
2172040-2
210 MHz
2172046-3
215 MHz
2172052-4
« Last Edit: May 07, 2024, 10:38:12 pm by BillyO »
Want to see an old guy fumble around re-learning a career left 40 years ago?  Well, look no further .. https://www.youtube.com/@uni-byte
The following users thanked this post: Mortymore
Offline gitm
« Reply #1351 on: May 07, 2024, 11:06:44 pm »
My PB470 probes also didn’t come with ground springs, but they’re pretty cheap at $1 apiece.
RE: bandwidth, using the probe tip to BNC adapter I was getting in excess of 300 MHz using PB470 on my “SDS824”. ;-)
Offline BillyO
« Reply #1352 on: May 07, 2024, 11:16:09 pm »
I was using a probe to BNC adapter.   Considering the SDS824 is down 3db @ ~260MHz, how attenuated was the 300MHz signal compared to 10MHz?  What was the signal generator?
Offline gitm
« Reply #1353 on: May 07, 2024, 11:32:04 pm »
This was the post I discussed the bandwidth https://www.eevblog.com/forum/testgear/siglent-sds800x-hd-12-bit-dsos-coming/msg5404754/#msg5404754
The frequency response was obtained using a TinySA Ultra as the signal generator doing a sweep out to 400 Mhz, -36.6 dbV amplitude (41.8mVpp) -3db was at 338 MHz.
Let’s see if I can get the picture in-lining to work….
« Last Edit: May 07, 2024, 11:34:44 pm by gitm »
Online KungFuJosh
  • TEAS is real.
« Reply #1354 on: May 08, 2024, 12:57:15 am »
Highly unlikely.  The OS would need support for this.
True, but if there are enough resources, it could be possible. It would be a smart move for the scopes that could use the external HDMI dongle.
"I installed a skylight in my apartment yesterday... The people who live above me are furious." - Steven Wright
Offline BillyO
« Reply #1355 on: May 08, 2024, 02:19:52 am »
Agreed.
Offline BillyO
« Reply #1356 on: May 08, 2024, 02:30:40 am »
Thanks.
I also used a TinySA Ultra.  As good as it is the SDS824X-HD can't compete with the SDS2504X-P on BW.  I'm not sure how to interpret your results.  We need a few more data points.
Offline tautechTopic starter
« Reply #1357 on: May 08, 2024, 02:50:14 am »
Thanks.
100%
To add, any analyzer TG source is not renowned for flatness over a wide sweep as indicated by some unexplained dips and ramps.
The following users thanked this post: pdenisowski
Offline BillyO
« Reply #1358 on: May 08, 2024, 03:14:49 am »
100%
For my tests I didn't use it in tracking mode.  I had verified the output from 10 MHz to 215 MHz was with in 5mV over that range  on the SDS2504X-P, but by setting the frequency manually.
Offline AndyC_772
« Reply #1359 on: May 08, 2024, 05:11:10 pm »
Do we have a specification anywhere for the number of waveforms/sec that the SDS800X-HD (and its bigger brothers) can process?
I'm running a simple experiment right now with a real board I'm interested in probing. It's generating a somewhat repetitive signal at around 250 Hz, and I'm viewing the interesting portion at 200us/div. I have two scopes connected to the same signal, the SDS804X-HD and my Agilent MSO-X3104A.
The 'trig out' ports on both scopes are connected to a 3rd scope which I can use to measure how many triggers/sec each is achieving.
As expected, the Agilent triggers every single time, no issue whatsoever. Nothing is missed. It's sampling at 1 GSa/s.
With the Siglent, it's dependent on the horizontal memory depth. At 10k and 100k points, it captures most waveforms - but not quite all. About 1 waveform in 20 is missed on average.
At 1M points it misses more triggers than it captures - the rate is down to 115 Hz. With 10M points (the same 1Gsa/s as the Agilent) it's managing just 29 triggers/sec, which means it's missing nearly 90% of trigger events.
Yes, Acq Mode is set to 'Fast'.
I feel as though I've found one of the limitations inherent in what is, after all, an incredibly cheap scope. Not a major problem, provided I know in advance what to expect - but I don't think this performance would be adequate for a scope that had to be my everyday workhorse.
Is the 2000 X-HD or 3000 X-HD series significantly faster?
Offline BillyO
« Reply #1360 on: May 08, 2024, 05:53:16 pm »
You can get this information from the data sheets for all the scopes here: https://siglentna.com/resources/documents/digital-oscilloscopes
However, do you really think it's fair to compare a $500 scope to a $33,000 scope?
« Last Edit: May 08, 2024, 06:01:33 pm by BillyO »
Offline Performa01
  • Country: at
« Reply #1361 on: May 08, 2024, 07:32:56 pm »
It should be obvious that the waveform update rate depends on the record length and not the sample rate. Thus it’s not fair to compare 10 Mpts on the Siglent to a DSO that can only use a total of 2 Mpts in run mode.
Furthermore, Keysight MSOX-series is optimized for speed; in fact, everything else is subordinate to this. Functionality is implemented in hardware in the Megazoom ASIC, which is blazingly fast, but also rather inflexible. By contrast, on the Siglent there’s always some processing time for the data, which pauses the acquisition. And if speed and waveform update rate are really that important, anything other than a Keysight Megazoom DSO will be a disapointment.
The average waveform update rates for the SDS800X HD can be found here (reply #2, section “History & Sequence Mode”):
https://www.eevblog.com/forum/testgear/sds800x-hd-review-demonstration-thread/msg5293747/#msg5293747
Instead of complaining about the update rate, most engineers would be excited about the analytic capabilities they get with the SDS800X HD and the many things it can do, which are beyond an MSOX3000, such as:
True 12 bit resolution.
0.5 % accuracy.
Custom probes.
Permanently visible 7-digit trigger frequency counter.
True 500 µV/div full resolution sensitivity and not just 4 mV/div and software zoom below, which would not be so bad on a 12-bit instrument, but with only 8 bits it certainly is.
<2.4 nV/√Hz voltage noise above 1 MHz when 50 Ω terminated.
Deep memory up to 100 Mpts total, which is not halved in run mode.
History recording permanently active in the background, where past trigger events can be looked up and analyzed to our your hearts content.
Deep math and measurements up to 10 Mpts, not just 32-64 kpts.
Up to 4 math channels incl. FFT and formula editor.
FFT up to 2 Mpts.
Measurement statistics with Histicons that can be enlarged to histogram windows.
Track Plots.
Bode Plots, up to 3 channels and up to 120 MHz with frequency selective detector and AGC, providing more than 100 dB dynamic range.
Last not least: the price!
For anyone not caring for these features, Analytic (Siglent) scopes in general are not for them.
The following users thanked this post: hfleming, egonotto, tautech, 2N3055, Mortymore, Martin72, core, Markus2801A
Offline AndyC_772
« Reply #1362 on: May 08, 2024, 09:46:09 pm »
Thank you!
I'm just trying to quantitatively assess the strengths and weaknesses of the scope in the specific context I bought it for. That's all.
The Agilent scope is just being used as a control, to make sure that any unexpected behaviour I see is indeed down to the scope and not the device being measured. Standard, good scientific practice.
« Reply #1363 on: May 08, 2024, 11:44:46 pm »
Thank you!
Directly proportional to how you drive and understand how to get the best from it.
Set a small mem depth and fixed sampling rate and test again.
Offline Martin72
  • Testfield Technician
« Reply #1364 on: May 08, 2024, 11:47:08 pm »
You couldn't pay the 800 a higher compliment.
(Kierkegaard)
Offline BillyO
« Reply #1365 on: May 08, 2024, 11:59:57 pm »
« Reply #1366 on: May 09, 2024, 12:05:37 am »
May I remind you the 800X HD range starts from just $339 !
The following users thanked this post: egonotto, Markus2801A
Offline EvgenyG
    • Evgeny's Blog
« Reply #1367 on: May 09, 2024, 05:36:02 am »
Still not available in Aus :(. I don't understand why everybody else in the world has it but not us. Is it about the C-Tick? Anyway, I gave up on waiting for 800X stock or someone selling their 1104XE cheap and bought an Owon HDS242S today, which was on super special. That will get my tasks done for now, and once 800X goes on sale/special, I'll get one.
Offline EvgenyG
    • Evgeny's Blog
« Reply #1368 on: May 09, 2024, 08:28:40 am »
Oh man, I've jinxed it. The Owon seller cancelled the order and refunded. Perhaps that's a sign...
« Reply #1369 on: May 09, 2024, 08:46:46 am »
Still not available in Aus ..........
Hmmm, your Google Fu is failing you:
https://appvision.com.au/index.php?controller=search&s=SDS800X+HD
https://appvision.com.au/index.php?id_product=903&id_product_attribute=0&rewrite=siglent-sds1202x-e-digital-oscilloscope-200mhz-2-channel-1gss-with-serial-decode&controller=product
https://appvision.com.au/index.php?id_product=904&id_product_attribute=0&rewrite=siglent-sds1202x-e-digital-oscilloscope-200mhz-2-channel-1gss-with-serial-decode&controller=product
https://appvision.com.au/index.php?id_product=905&id_product_attribute=0&rewrite=siglent-sds1202x-e-digital-oscilloscope-200mhz-2-channel-1gss-with-serial-decode&controller=product
https://appvision.com.au/index.php?id_product=906&id_product_attribute=0&rewrite=siglent-sds1202x-e-digital-oscilloscope-200mhz-2-channel-1gss-with-serial-decode&controller=product
https://appvision.com.au/index.php?id_product=907&id_product_attribute=0&rewrite=siglent-sds1202x-e-digital-oscilloscope-200mhz-2-channel-1gss-with-serial-decode&controller=product
https://appvision.com.au/index.php?id_product=908&id_product_attribute=0&rewrite=siglent-sds1202x-e-digital-oscilloscope-200mhz-2-channel-1gss-with-serial-decode&controller=product
Just how Appvision cocked up the URL's so bad is beyond me  :palm: yet all are listed as in stock and same day dispatch......they actually have more models than we do.  :-[
« Last Edit: May 09, 2024, 08:50:49 am by tautech »
The following users thanked this post: egonotto, EvgenyG
Offline EvgenyG
    • Evgeny's Blog
« Reply #1370 on: May 09, 2024, 09:30:56 am »
Amazing. I checked yesterday before ordering that Owon unit, and none of the 800X units were not in stock at Appvision.
They must have received them today (or updated their website). Will be placing an order soon but will definitely check with them if they actually have stock. Yes they've got other higher models but I can't justify purchasing those for hobby use. I'll be ordering the 804X. Wavecom also received new units but not the 800X range.
« Reply #1371 on: May 09, 2024, 09:43:45 am »
Yep, best bang for buck.
Shipped one out today to a customer that has worked that out too.  ;)
Offline ebastler
« Reply #1372 on: May 12, 2024, 04:56:40 pm »
Quick question to those experienced with the current generation of SDS scopes: Is there a way to save "Measurement Trend" results, beyond taking a simple screenshot?
I would like to store the complete set of measurement values as a CSV or BIN file -- but that seems possible only for the physical channels and Math traces. Am I looking in the wrong place, or is this indeed not supported? Would be a pity to have such a nice long-time recorder for 1000s of measurement values, but then have the information reduced to the screenshot pixels only.
« Last Edit: May 12, 2024, 05:04:31 pm by ebastler »
Offline 2N3055
  • Country: hr
« Reply #1373 on: May 12, 2024, 05:52:59 pm »
No, not supported.
The following users thanked this post: egonotto, ebastler
Offline Performa01
  • Country: at
« Reply #1374 on: May 12, 2024, 06:15:25 pm »
This is currently supported on the SDS7000A only. Reason is that Trend and Track are math functions there, which in turn is a consequence of the vastly different multi-window strategy on these scopes.
