I have an SDG2042X (upgraded to 120 Mhz). I used it as a signal source. Synchronized the outputs at 50 Ohm load setting. Connected scope's CH1(DC1M) to CH1 of AWG using pass-through 50 Ohm terminator at the output of AWG. CH4 (DC50) of the scope is directly connected to CH2 of the AWG.
...
Using 1Hz-120MHz sweep setting at 632 mVpp (~0dBm) I created a sweep signal. Analyzed the 3 probes again.
Nice measrements!
Here come some remarks:
Contrary to popular believe, probe bandwidth is nothing like e.g. scope bandwidth. There is no clear 3 dB corner frequency, where the probe output begins to fall at a constant rate. Essentially, probe bandwidth only means that the manufacturer is not willing to guarantee anything for frequencies above.
It has been demonstrated quite often by now, that even a humble 100 MHz probe can outperform a 500 MHz probe on certain instruments and in certain circumstances.
As has been pointed out already, the true performance characteristics in any practical setup will be determined by many parameters, of which the specified probe bandwidth is certainly the least important. The input capacitance is about the only obvious quality criterion – for practical use – not for the industry-standard test using a 25 ohms source.
What
is important is the frequency compensation. All probes have a compensation trimmer for the LF (low frequency) compensation. This is for adjusting for best pulse edge fidelity with the internal 1 kHz square wave calibration signal. But this only affects the probe gain at higher frequencies without changing the frequency response. In other words, if you ignore everything below about 1 MHz, then the low frequency compensation is nothing but a gain control.
A passive probe is a rather complex design, and it needs more compensation measures to work properly at high frequencies. This is because of the input impedance of the DSO, which is certainly much more complex than just 1 megohm in parallel with 17 pF. So you can have either probes that are optimized for the oscilloscope model or the universal ones, that provide two or even three trimmers for optimizing the frequency response on different instruments.
Your measurements show that the HF-compensation happens to be just perfect for the combination of SDS2000X+ and the Diligent probe, while it is rather mediocre for the low end Siglent probes – up to the point where the rise time measurements are seriously impacted.
The reason for this would be that the PP510 and PP215 have been designed for the SDS1000X-E series DSOs, where they both perform pretty good. See reply #56 in the thread linked below:
https://www.eevblog.com/forum/testgear/siglent-sds1104x-e-in-depth-review/msg1434665/#msg1434665You can also see from my tests that the performance of the PP510 and PP215 is within about 0.5 dB up to 100 MHz and pretty much identical above that. Even though my tests have been carried out with early samples of these probes, I can’t believe that performance has changed significantly since then. Maybe you could repeat your test, but instead of using the 1 kHz square wave for calibrating the probes, you could try to use a 100 MHz sine wave and adjust both probes for 0 dBm response in your scenario. After that, the results should be pretty much identical for both probe models.
So the PP510 and PP215 probes have been optimized for the SDS1000X-E series and are certainly not the preferred solution for higher class oscilloscopes like the SDS2000X Plus. Yet in practical terms this hardly matters because … see next paragraph.
All this is just an academic discussion, because in any real-world scenario a passive high impedance probe like the ones discussed here isn’t going to be very useful for high frequencies anyway. Input impedances in the realm of 10 pF mean that any probing above 100 MHz requires a source impedance below 160 ohms, just to keep the measurement error below 50%. Once you’re at source impedances that low, a passive low impedance probe (which essentially is just a resistive match to a 50 ohms coax line) usually is the much better solution.