Pulse response and bandwidth for optional 500MHz in interleaved channel modeThe 500MHz option is available for all the ones who would like to measure that high occasionally, but don’t need it on all channels at the same time.
Here is the frequency response:
![Siglent SDS2000X Plus 873204-0](https://www.eevblog.com/forum/index.php?action=dlattach;topic=218238.0;attach=873204;image)
SDS2354Xplus 2GSa 8bit 1GHz
As you can see, my pre-production sample does a fairly decent job:
-1.0dB @ 430MHz
-1.7dB @ 500MHz
-3.0dB @ 570MHz
Here’s the pulse response to a 60MHz squarewave with fast edges:
![Siglent SDS2000X Plus 873200-1](https://www.eevblog.com/forum/index.php?action=dlattach;topic=218238.0;attach=873200;image)
SDS2354X+_2GSa_Pulse_60MHz_D
In this screenshot, we can spot a number of nice features:
• The 7-digit frequency counter (how many other 2000 series DSOs have that?) right below the Siglent logo. The actual signal frequency has been measured as 59.9999459MHz, so the deviation is only 0.265ppm. Not bad for a DSO in this class.
• A nice display with working intensity grading
• Stable measurements (except overshoot), indicated by low standard deviation in the statistics.
• Optional histograms for the measurement statistics that can be enlarged for further analysis if so desired.
• Rise time measurement of ~720ps average, which would hint on a bandwidth just below 500MHz for an input channel with Gaussian filter shape, but that’s certainly not a valid method to determine the bandwidth of a DSO. We already know it’s actually 570MHz.
Finally let’s see the frequency domain of that signal:
![Siglent SDS2000X Plus 873208-2](https://www.eevblog.com/forum/index.php?action=dlattach;topic=218238.0;attach=873208;image)
SDS2354X+_2GSa_Pulse_60MHz_FFT
We can see that the harmonics of the signal extend well beyond 1GHz and it is not a perfect rectangle, as the even harmonics are still quite strong.