Author Topic: Siglent SDS2000X Plus - Bugs / Missing Features / Feature Requests  (Read 268852 times)
Offline roberthh
Re: Siglent SDS2000X Plus - Bugs / Missing Features / Feature Requests
« Reply #150 on: October 13, 2020, 11:09:39 am »
SDS2104x-plus, 1.3.5R10
BUG: Inconsistency between displayed and used settings.
Trying to do a simple test with UART decode I trapped over a pitfall, which I would consider a BUG. Setting the protocol signals, the menu showed fine C1, threshold 1.57V. But it decoded wrong. It turned out, that the threshold was wrong. Even if the menu said 1.57V, the value used for decode was 0. It only came right after I had reset it again to 1.57V. This is a problem which occurs at many places, that the settings shown in the menu do not reflect the actual settings used for the measurement.
I did not test yet how easy of difficult it is to repeat  the error may not be easy to repeat. One may have to change between protocols or go through a power cycle.
P.S.: At first I thought it was a problem in decoding a custom baud rate, but it was not.
@cobra514: when setting the protocol signals, try to set the threshold level again by varying it a little bit.
Edit: It might be that the issue I have seen with 8/10 bit setting and I2C decode is also related to this topic. Using the same test set-up again and deliberately adjusting the threshold for SCL and SDA again, there is not difference between 8 and 10 bit mode. I can however set the threshold for SDA to an value, where 8 bit decodes fine, but 10 bit not. This level was around 2.5 V in my set-up.
« Last Edit: October 13, 2020, 12:19:54 pm by roberthh »
Offline tv84
  • Country: pt
« Reply #151 on: October 13, 2020, 11:21:11 am »
If real, this type of bug is a real PITA. One looses confidence...  :palm:
Online DL2XY
« Reply #152 on: October 13, 2020, 11:51:41 am »
These are the screenshots of my tests.
DUT is a UHF data receiver unit. It has 3 PLL IC`s (MAX2871) connectet to MCU via SPI. Each PLL has 6 write only register (32bit).
The signal is the initialization sequence requiered: R5 to R0 PLL1 (CS1),R5 to R0 PLL2 (CS2) and R5 to R0 PLL3 (CS3). For testing i repeatet this sequence at 10ms intervall.
The least 3 bits of  32bit  words are defining the register address.
One scan, digital input, triggered on data (CS2,R3)
1088154-0
Long memory test, digital input, only CS2 decoded, triggered on data (R3),
1088158-1
Zoomed to R3, digital input
1088162-2
Analog input, triggered on CLK timeout,  3 PLL's* 6 registers
1088166-3
Zoomed to R3
1088170-4
18 full scans
1088174-5
Analog inputs with CS, trigger on data (R3 CS2), one full scan.
1088178-6
« Last Edit: October 13, 2020, 11:58:35 am by DL2XY »
Offline Cobra514
« Reply #153 on: October 13, 2020, 02:22:59 pm »
@Cobra514: Looking at the pictures of your first series, the ones that decode right have the CS low/high slope in the captured trace, the one that fail do not have it. What is the CS type setting in the Protocol Signals tab? For a low level CS it has to be ~CS (or CLK Timeout, if the CS transition is too far off or if there is no CS). Due to the CLK synchronous nature w/o timeout the decoder cannot tell from CLK and Data alone where a data item starts. So it needs some other indication.
If the CS setting changes the decoding, still the question arises about the difference between odd/even numbers.
Hi roberthh.  Yes, good question.  The reason the CS low/high transition is not visible in some of the pictures is simply due to timebase difference.  But it is present in all of the tests.  After I send a command with my SPI generator, it raises the CS line to prepare it to do a high/low transition.
My generator is set up for ~CS (active low) and per the Siglent User Manual: "The ~CS signal needs a complete falling edge on the screen to be regarded as active."
So therefore I was careful to ensure that the complete falling (and subsequent rising) edge of ~CS is on the screen during the decode.
There should be no problem here.
Offline Cobra514
« Reply #154 on: October 13, 2020, 02:46:23 pm »
@DL2XY - Thank you for taking some time to post.
Probably the closest conditions shown there to one of the conditions I have tried would be xxxxpng_5, where the analog channels are used AND the timebase is under 50uS/div AND performing 32-bit decodes.
I actually would not see an issue with decodes as well with the condition you have posted there.
In my situation, I only see a failure on the decode of the last nibble sent and decoded.  I see in  your plot that there is subsequent SPI data.
In other words, if I send (n) 32-bit messages, (n-1) of them will be correct and the (nth) will be decoded incorrectly (for timebase <100us only as well).
Not sure what it means, but for 8-bit decodes I usually miss the last bit of the LSB.  For 32-bit decodes I sometimes miss the entire last nibble and decode "0" regardless of value.
Offline TK
« Reply #155 on: October 13, 2020, 02:53:43 pm »
Siglent scopes are bad at keeping serial trigger and decode information after reboot, so every time you reboot, you need to check the serial trigger and decoding settings like channels assigned to signals and the thresholds
Offline roberthh
« Reply #156 on: October 13, 2020, 03:01:18 pm »
Did you make tests with clk-timeout as CS setting?
Offline Cobra514
« Reply #157 on: October 13, 2020, 03:03:59 pm »
Yes, thank you - I did try that.  I actually highlighted the threshold "box" and used the universal knob to vary threshold all the way from GND up to the top of the signal.  My findings were consistent except when at the vary top or bottom of the signal.
Offline roberthh
« Reply #158 on: October 13, 2020, 03:20:04 pm »
That is not a problem as long as it is visible. The issue I have seen was that the menu for (here I2C signals) showed different values than the actual setting, until I changed deliberately each and every entry in that menu again, irrespective of what was already shown.
Offline Cobra514
« Reply #159 on: October 13, 2020, 03:20:34 pm »
@roberthh
My apologies, you mentioned this before.  Yes, I did try the clk timeout as well.
I attached 2 pictures of subsequent transmissions (seconds apart) using clk timeout.  Both performed with the same data sent.  One OK, One failed decode (last bit).
Both performed at the same timebase as well in this case.
Each time I click send on my generator I get randomly AC or AD (AD is correct).
I tried the "history" feature as well in this one.
...and if you look closely, the rising edge of ~CS is visible at the left hand side of screen in both screens (since this was a minor difference noted before).
The two patterns are identical save for some mV level noise.  The decodes are different.
*EDIT* As expected, for timebase >100us/div the decode worked perfectly for me every time
« Last Edit: October 13, 2020, 04:17:00 pm by Cobra514 »
Offline roberthh
« Reply #160 on: October 13, 2020, 06:19:11 pm »
@Cobra514 Hi. I used a micro to simulate a SPI with similar wave than yours. Clock and data transition at exactly the same time.  The result: The last bits are decoded wrong. Instead a 0x33 it shows 0x30, instead of 0x37 it shows 0x30 too. Since a sequence of 1 bits is decoded as 0, that explains why 0xff gets 0x00. My simple USB based logic analyzer works well.
BUT: That was all with a pattern, where the quiet levels of Clk and Data is high. If I change it to Clk & data low when inactive, all decodes well.
Edit: It is the miso line which makes the difference.
Edit 2: It is sufficient that the miso line is low for a short while only. It worked with a 40 ns low state. I could not test further with the little micro I used for pattern generation.
In any case, the active phase of clock was the falling one. And I do not see a setting to cater for a low of high quiet state of the bus. Attached:
spi_high_033.png:  Wrong decoding
spi_low_033.png: Proper decoding
LA2016_0033.png: Decoding with the USB LA
spi_datalow_0013.png. Proper decoding with data returning to 0
spi_data_pulse_0017.png: Proper decoding with a 50 ns low miso pulse at the end
« Last Edit: October 13, 2020, 07:10:40 pm by roberthh »
Offline Cobra514
« Reply #161 on: October 13, 2020, 06:53:58 pm »
@roberthh,
That is very good work, indeed.  Impressive.
For once I am not the only one seeing this :phew:
Seems it should be capable of decoding either way.
When I free up again, I might have to try all the options for CPOL, CPHA, and SS (active high and low).
Online DL2XY
« Reply #162 on: October 13, 2020, 09:51:29 pm »
Maybe the problem results from incorrect decoder settings. You are decoding 8-bit wise but the bitstream is one 32-bit block!
The transfer ends with the rising transition of CS, so if you put out 32 bits while CS is active you should decode it as a 32 bit word.
Offline TK
« Reply #163 on: October 13, 2020, 09:52:37 pm »
How did you simulate the SPI signals?  Bit banging?  The signals do not seem to be the same on the different capture examples
Offline Cobra514
« Reply #164 on: October 13, 2020, 10:46:13 pm »
No.  That's not the problem.
See failed decode attached.  Note on upper right of screen that "Data Length" = 8 is selected.
Doesn't get much simpler than this...
A single byte (0x11) sent and (0x10) decoded.
MOSI does end in a high state, however.
« Reply #165 on: October 14, 2020, 12:27:04 am »
Yes but and a big but, what protocol has both rising and falling edges ?
CS = active low
MOSI = active low
CLK = active high
Offline TK
« Reply #166 on: October 14, 2020, 01:06:13 am »
CS = active low
MOSI = active low
CLK = active high
it is not both rising and falling edges, it has to do with clock polarity and phase and SPI has 4 modes based on the combination of these 2 parameters
Offline TK
« Reply #167 on: October 14, 2020, 01:15:10 am »
What is the SPI decoding setting on the scope, including threshold levels for all signals?
« Reply #168 on: October 14, 2020, 01:16:03 am »
CS = active low
MOSI = active low
CLK = active high
it is not both rising and falling edges, it has to do with clock polarity and phase and SPI has 4 modes based on the combination of these 2 parameters (Attachment Link)
Right, thank you for that however if we study the screenshots edge alignment of all 3 traces they don't make sense.  :-//
CS has no relation to the real CLK other than it's active for way in excess of the packet and the start of CLK is after the start of MOSI.
Certainly conditions that are most unusual.
Garbage in = garbage out.
« Last Edit: October 14, 2020, 01:18:58 am by tautech »
Offline TK
« Reply #169 on: October 14, 2020, 01:19:58 am »
CS = active low
MOSI = active low
CLK = active high
CS has no relation to when the CLK / MOSI / MISO signals start toggling.  You set CS and you can wait as long as you want to start transmitting.
When the sampling occurs with CLK going low to high, the MOSI line needs to toggle before the first clock toggle, so it is correct that MOSI moves first
Offline TK
« Reply #170 on: October 14, 2020, 01:21:47 am »
CS = active low
MOSI = active low
CLK = active high
What are the blue arrows on the top?  It looks like decoding is ending sampling on the second blue arrow, and it is the 7th clock, missing the last clock toggle where bit is 1, but as decode stopped, it shows 0.  You can also see that the white rectangle indicating decoded packet ends on the 7th clock
EDIT: I don't think the blue arrows are limiting the decoding, but it is clearly sampling on clock edge low to high, and the last bit is 0 when the 8th clock starts to transition low to high
« Last Edit: October 14, 2020, 02:05:17 am by TK »
Offline Cobra514
« Reply #171 on: October 14, 2020, 01:56:56 am »
CS = active low
MOSI = active low
CLK = active high
The dark blue arrow is merely indicating the trigger.  In that shot I had it detect a SPI pattern (so it needs to see some activity before it triggers.
I have similar plots for the exact same data (and error) where the trigger is at the falling edge of CLK (because I set the trigger for falling edge of CLK).
Last bit is not 0.  This is set for MSB.  Look again.
By the way.  If I set the timebase to 100us/div or higher, it decodes correctly.
Offline TK
« Reply #172 on: October 14, 2020, 02:09:25 am »
CS = active low
MOSI = active low
CLK = active high
I can see what you mean.  I don't like how Siglent serial decoding works, but this bug is the most bizarre one I saw in any scope.  BTW, how are you generating the SPI signals?  Is there a pull-up on the MOSI line?
Offline Cobra514
« Reply #173 on: October 14, 2020, 02:26:02 am »
CS = active low
MOSI = active low
CLK = active high
Covered in previous posts. I know - there are many.
Offline Cobra514
« Reply #174 on: October 14, 2020, 02:30:52 am »
I'm simply going to move on from this decode topic.
I have spent too much time on this and have other more important tasks to work on.
Thanks to everybody who looked into it.
Hopefully just something I'm doing on my end here...
...and I can always just use the timebase workaround (although there are times I'd rather not).