« Reply #150 on: June 15, 2022, 04:36:13 pm »
Shortly playing a bit with the instrumentcontrol via web..
The tektronix mode is still on board (in the menu), DVM looks nice and what I like:
The axis-labeling seems to have no more than 2 digits after the comma, not 4 like the sds2k+ got.
Testing shortly the scpi command line, with interesting response @tv84..
More on the weekend, probably saturday.
(Kierkegaard)
Offline tv84
  • Country: pt
« Reply #151 on: June 15, 2022, 04:44:40 pm »
 >:D  So, UNNAMED_LIC-V2 as expected since the software is shared with 5000 and 6000.
All can forget the public keygen. One needs the non-public.  :popcorn:
Online Martin72
« Reply #152 on: June 18, 2022, 09:32:57 pm »
Tomorrow, when time will allow it, I´ll "test" the scope in the afternoon by using my siglent demo-board.
Then check if tv84 is right and try to install keys from the gen*.
Last will be trying to get access via telnet.
If I got the time.... :( 8)
*)  Generating keys.... When typing SDS2000XHD or SDS2000X HD, different keys will be generated - Or will be 2000X_HD the right....
(Kierkegaard)
Offline Performa01
  • Country: at
« Reply #153 on: June 19, 2022, 02:41:25 pm »
If you want a demonstration of the true 12-bit acquisition in an SDS2000X HD, I can share my test results for measuring low signal levels using the FFT function.
Look at the table attached to this posting:
Level Measurement Results
It compares the SDS2504X HD at its various acquisition modes (Normal, ERES 2.0, Avg.16) with the SDS2354X Plus in 8 and 10 bit modes using a variable 10 MHz test signal.
The upper part of the table contains the tests at 100 mV/div vertical gain and we can see that the first order dynamic range for <1 dB amplitude error is in excess of 100 dB for the HD instrument, while we have to tolerate slightly more error even for the limited dynamic range of only 50 dB on the 8-bit X Plus.
Now some might ask why the difference is so huge? An improvement by 24 dB was to be expected in theory, but here we seem to have at least 50 dB. Let’s do some math:
One division is 30 LSB on the X Plus. At 100 mV/div, one LSB is therefore 3.33 mVpp.
Below this level, the 8-bit ADC doesn’t see a continuous signal anymore and reads increasingly low.
One division is 425 LSB on the X HD. At 100 mV/div, one LSB is therefore 235.3 µVpp.
This level is (just) low enough, so that the frontend noise can act as a dither and the ADC keeps working just fine. Since the FFT is a resolution enhancement technique in itself (buzz word: “process gain”) it can indeed deliver higher resolutions, even in an 8-bit system, but an appropriate dither is required for this to work.
Now for the lower part of the table, where a 40 dB higher sensitivity at 1 mV/div is used. One LSB is 33.3 µVpp for the X plus and 2.35 µVpp for the X HD. In both cases the frontend noise can act as dither just nicely.
But now we have the opposite problem: the noise reduction and resolution enhancement of the FFT is limited, so at one point the frontend noise just starts obscuring the measurement results.
The first order dynamic range for <1 dB amplitude error is now about 60 dB for both instruments, yet we can see that the X HD still has a slight edge at even lower levels.
We should appreciate the fact that with these FFT settings and independent of the vertical gain(!), at least in case of the X HD, we are able to measure down to -100 dBm with better than 1 dB accuracy – that is a level of 2.24 µVrms. And if we do this using the 100 mV/div range, where the maximum level can be 800 mVpp = 282.84 mVrms = 2.04 dBm, then we’re talking about more than 100 dB dynamic range.
Look at the attached example of a -100 dBm measurement at 100 mV/div with the SDS2504X HD:
SDS2504X HD_LVL_10MHz_100mV_-100dBm_Normal
This test also revealed that all further measures for resolution enhancement and/or noise reduction do not provide any advantage over the simple “Normal” acquisition. Especially averaging makes things a lot worse.
I want to use this opportunity to remind you of the resolution enhancement by ERES, where every half bit provides that “half bit” enhancement only in terms of ENOB, but of course still one full bit in terms of resolution enhancement. No wonder, because apart some theoretical numbers, something like half a bit cannot exist in reality. So, bear in mind that ERES 2.0 produces 4 additional bits of resolution, hence we get a total of 16 bits on the X HD. Anything higher than that (like ERES 3.0) doesn’t make sense, since the internal data paths and memory widths are limited to 16 bits.
« Last Edit: June 19, 2022, 02:48:50 pm by Performa01 »
The following users thanked this post: egonotto, Martin72, teddychn
Online Martin72
« Reply #154 on: June 19, 2022, 04:13:01 pm »
Yepp, proofed...Won´t function on this scope. 8)
Telnet via putty don´t work also, if I didn´t made a mistake.
(Kierkegaard)
Offline voltsandjolts
« Reply #155 on: June 19, 2022, 05:25:47 pm »
Are there any teardown pictures anywhere?
I don't see any in this thread yet...Martin ;D
Online Martin72
« Reply #156 on: June 19, 2022, 05:43:25 pm »
I can post a snip - As I´ve looked into the fan-hole today... ;)
Because:
This scope is super-silent, no question  and my working room can be silent too, today it was silent in the room..
Then I´ve heard a silent clac.....clac-clac...clac...noise.
Once you´ve heard it, you will hear it any time.
It´s the fan, his gears seems to be slightly defective or something´s coming too near to his flaps.
Nevertheless, I´ve informed batronix together with the offer by fixing it by myself - When warranty will not lost in this case.
When they agree, we will get pics. 8)
If not and I have to send it back, it could be last several weeks - I was the only one who got one from Batronix, they only got three.
One for demonstration, one for the "embedded world 2022" in Nurenberg (next week) for a give away and the last is mine.
Edit: "snip" shows a label near a bigger hestsink with qr-code and you can read "E9" and "Atom" on it...
« Last Edit: June 19, 2022, 05:49:22 pm by Martin72 »
(Kierkegaard)
Offline tv84
  • Country: pt
« Reply #157 on: June 19, 2022, 06:00:45 pm »
In the SDS6000 software we can find:  (this one must have the same as they seem to share the code)
Product_types:
ZODIAC
ZODIAC+
LP
NEZHA
SDS2004HSR
SDS2000HSR_1G
SDS2002HSR
SDS2004X_PLUS
SDS2002X_PLUS
ATOM_4CH
ATOM_2CH
1- SDS1002X_E
2 - SDS1004X_E
3 - SLA1016
4 - SDS2000X_E
5 - ARCHER
Product_versions:
UNNAMED
4CH_500M
4CH_1G
2CH_500M
2CH_1G
6 - ZDL
9 - ZDLP
« Last Edit: June 19, 2022, 06:23:42 pm by tv84 »
Offline INLET1011
  • Country: jp
« Reply #158 on: June 22, 2022, 04:56:09 pm »
Does SDS2000XHD require a dynamic key, and we can't unlock it?
My hope dashed... :'(
I wanted MSO and Bode plot options, but I need additional 785€.
The advantage over RTB2004 with full options has been shaken.
(Of course, both are expensive for hobby)
Online Martin72
« Reply #159 on: June 22, 2022, 10:21:29 pm »
Usually it won´t last long until the first promo-offers will appear, like e.g. (nearly)all options including.
For people like me who are the first buying a brand new model it´s always a kick in the ass.
A kind of punishment being so dumb and got one before the offers comes up.  ;)
There was only one exception:
As the promo-offers for the rigol mso5000 came up, I got a free bundle from the rigol support, although I´ve already bought the scope over a year before.
(Kierkegaard)
Offline hpw
« Reply #160 on: June 23, 2022, 09:08:43 am »
@Martin72: "For people like me who are the first buying a brand new model it´s always a kick in the ass."
The other name would be Banana wear... the other pitta, as it gets now about one year without any FW updates. Even the Siglent support without any timeline when things gets fixed on FW.
This new DSO is not a bargain any more, some nice to have features, 12 Bit could be done cheaper by a LNA if this allows.
Hp
Offline 2N3055
  • Country: hr
« Reply #161 on: June 23, 2022, 09:33:18 am »
Hp
Could you please explain this ?
Offline hpw
« Reply #162 on: June 23, 2022, 09:40:33 am »
Hp
As 12 bit gets lower sensitivity ... using an external LNA (Low Noise Amplifier) you may control the BW as you like for your application and have a gain of your choice as up to 60dB as using on PN or Audio measurements.
Offline 2N3055
  • Country: hr
« Reply #163 on: June 23, 2022, 10:10:22 am »
Hp
Ah, now I understand.It is not like that.
12 bit is not primarily about sensitivity. It is about instantaneous dynamic range. A difference between smallest and largest signal you can look at the same time. Fact that 12 bit scope will see smaller signal at most sensitive range is kind of a bonus, but here we are already getting to the point of noise being a problem.
So for looking into very small signals (with no large signals present) you will always need some kind of LNA for best results..
That being said, this new SDS2000X HD has both 12 bit resolution (meaning better resolution to resolve small signals in the range) AND low noise front end, that is easily one of the best on market in that segment.  That results in fact that the point where you have to use LNA is pushed almost an order of magnitude lower than before. For instance, you could easily verify linear PSU that claims 70µV RMS noise can be easily measured, provided you have good probing....
The following users thanked this post: Electro Fan, voltsandjolts
Offline hpw
« Reply #164 on: June 23, 2022, 10:58:45 am »
@ 2N3055 ... yes probing see my other thread as currently  :palm:
Offline swmcl
« Reply #165 on: June 29, 2022, 10:47:01 pm »
I'm guessing the high price of the HD model is seen by many as too much to pay for a little more resolution ?  The thread seems to have died ...
Offline Bad_Driver
« Reply #166 on: June 30, 2022, 01:49:42 pm »
Yes, it‘s expensive!
Do I need it? No.
Do I want it? Yes!
Can I afford it? ….depends of upcoming liberation progress which for simple minds like me will be handy enough  :=\
The following users thanked this post: doppelgrau
Offline Performa01
  • Country: at
« Reply #167 on: June 30, 2022, 06:24:42 pm »
The SDS2000X HD is a very nice instrument. I like it a lot and I think this has also to do with its design - it has the same screen size as the SDS2000X Plus, yet is more compact and looks like the cute little sister of the SDS6000.
It does not only provide genuine high resolution because of the true 12 bit acquisition, but also a higher than usual accuracy like <0.5% gain error.
I've checked this and think the real world test result speaks for itself. A 1 kHz quare wave with 3 V amplitude (specified error is <1%) from an SDG7102A with OCXO option is fed into channel 4 of the SDS2504X HD. Bandwidht limit 20 MHz for lowest noise and best accuracyhas been enabled. See attached screenshot.
Online Martin72
« Reply #168 on: June 30, 2022, 07:07:14 pm »
Actually it´s hard to find one who got them in stock - Batronix told me that I´d simply luck to get one.
And:
We´re talking about a scope with an entry-price of appx 3500€, over 2000 bucks more than for the "topseller" sds2104X+.
And it´s hard to hack..
And for this price you can get a SDS5000X, starting with 350Mhz and 5GSa/s.
I gues for not a few the 1400€ for the sds2104X+ was absolute top of the edge for their hobby budget.
So much less people are thinking about to buy a scope for 3500.
And these one are having the choice between the SDS5034X and the SDS2104X HD.
For me the decision was very simple.
I just want to have it, I´ve told it months ago, if it will be avaible, it will be "immediately" mine.
In the last weeks I couldn´t play much with it, but tomorrow is the last day before my summer holidays... 8)
(Kierkegaard)
The following users thanked this post: Performa01, egonotto
Online Martin72
« Reply #169 on: July 03, 2022, 10:11:14 pm »
Oh and another new feature to check out.....Probe Check. You can find it in the probe menu when setting up probe attenuation settings.
It checks for adequate/correct compensation and if OK assigns the probe a Pass.
This has also been introduced to SDS5000X.
Oh yes, it´s a cool thing to watch, let me think about it to do a video for the very first time.. 8)
(Kierkegaard)
Offline rf-loopTopic starter
« Reply #170 on: July 06, 2022, 03:36:17 pm »
A little about the Bode Plot (FRA) measurement dynamics of the SDS2000X HD model.
Some manufacturers, for example Keysight and R&S, state the FRA dynamic range from 0dBm down.
It is quite a good way because the dynamic range from the 0dBm level upwards is obvious in other ways.
One essential point is whether the FRA function uses automatic continuous level adjustment during the frequency sweep.
In this case, the entire level range available to the oscilloscope can be used.
The 0dBm level is also practical due to the fact that with the input sensitivity of the oscilloscope set to 100mV/div,
the signal in question is almost the entire vertical range.
In many oscilloscopes, like the Siglent, the vertical display range is 8div.
(of course the full display range would be around 2dBm, 800mVpp)
Siglent's maximum FRA level (multiplier 1x) is of course 80Vpp (10V/div).
This corresponds to a level of 42dBm.
(When I use dBm in this context, unless otherwise noted, the impedance is 50 ohms.)
Of course, the use of such levels requires the use of an external 50 ohm load.
SDS2000X HD's maximum level with its own internal 50 ohm impedance is practically around 22dBm (8Vpp.
In this case max V/div is limited to 1V/div.
Even though max 5Vrms with 50 ohm inputs is mentioned, it is not available for measurements.
Everyone can simply calculate / evaluate the part of the dynamic range above the 0dBm level according to the current usage situation.
The end result is that we measure the dynamics from the 0dBm level down.
Siglent Bode Plot (FRA) have two modes: Channel Gain: Auto  and Channel Gain: Hold. (aka manual)
When Auto is in use, during sweep FRA continuously (for every single freq step in sweep) adjust input channels sensitivity from 500uV/div up to 10V/div using also fine adjust.
When Hold is in use, during sweep FRA keep input channels as is when Bode Plot analyzer function is activated. If user want adjust, whole Bode Plot need shut off. (least in this version)
Naturally only Auto mode give full dynamic range (what can be up to over 120 - 140dB (with some notes))
Auto mode is more slow than Channel Gain Hold mode.
In many simple cases Hold mode is well enough for measure something what do not need high dynamic range.
But, in Hold mode, before even start Bode Plot function user need set input V/div so that DUT out signal level do not clip in any point during frequency sweep.
There is also possible to use test signal (DUT In) level profile during sweep but I do not mess and mix it to this simple dynamic test (even when it can use also as tool for some dynamic advantages, but in these cases also measurement methot need be Vin/Vout, instead of Vout).
For look dynamic range I use attenuator method.
It is a simple sure and reliable way to test this. At the same time, you can also see how the phase measurement reacts.
Transit time of the type of attenuator used is (especially at the frequencies used here) very precisely the same regardless of the attenuation level between 0dB and 145dB.
The mechanical step attenuator used, if anyone is interested, is the HP 33321SC/333221SD pair what can step 0 - 145dB with 5dB steps.
Test signal is 0dBm, which goes to the attenuator and from there to input 4 of the oscilloscope.
Ch4 has a high-quality external 50 ohm load due to other experiments I did with slightly higher powers.
Input channels 2 and 3 are not connected to anything.
They reflect the FRA noise floor.
The signal of the reference channel has been attenuated. In this context, it does not affect the test result.
(with higher powers, instead, the crosstalk of the channels must be taken into account in some cases)
Test settings main things.
Signal 0dBm to attenuator what is here as "dummy" DUT.
Center frequency 455kHz (this is because it is one common IF freq and with high quality filters need very high dynamic range if need characterize also stop band.)
Span 10kHz
Input channels 20MHz BW
FRA Amplitude mode in this test is: Vout
Unit: dBm (note there is used external high quality feed thru 50 ohm for DUT out channel (Ch4)
In all test here, FRA sweep start with specified attenuated level then around 454kHz it step to 0dB and then back to attenuation.
First, Channel Gain: Auto
CH3 is there only for display FRA base noise level when 500µV/div
-85dB
-90dB
-95dB
-100dB
How high is -100dBm signal here (sine wave of course).
It is ~6.3 µVpp,  2.24 µVrms, -113dBV  Think about it...
Still level well detectable over noise level.
-105dB
Even with this level if you just disconnect test signal you can detect how it drop to full noise floor.
(-105dBm signal (average) level distance to displayed base noise average level is visually estimated still roughly 10dB)
Here -105dBm signal level is 3.6 µVpp  and we can see it is still roughly 10dB over base noise average.
If displayed noise average is roughly -115dBm ...  keep in mind that 19bit ideal ADC dynamic range is 114dB.
But with Channel Gain Auto our total maximum level correspond 42dBm and this level distance to displayed noise average is roughly 156dB.
Note also test frequency. If we go to audio frequencies, noise level rise, same happen specially when go to higher freq. Example around 60MHz FRA base noise average is roughly -100dBm.
What is FRA base noise level at other frequencies.  This together with previous tests with 455kHz can use for roughly estimate available and usable dynamic range for other frequencies.
Here tested from 10Hz to 60MHz. Inputs BW 200MHz, inputs open, no signal. (note that with 2000X HD model internal generator max is  25MHz and max level is 13.5dBm)
In image level top is 40dBm so that vertical include FRA whole one sweep dynamic range from noise to max when Channel Gain mode is Auto.  (absolute max before signal clipping is bit more than 40dBm and this kind of levels need also external 50 ohm load and of course DUT what have this output level)
Next using Channel Gain: Hold (what ever signal level is when FRA start and during sweep it keep same V/div)
Now CH2 is there only for display FRA noise level when 100mV/div
CH3 is there only for display FRA noise level with 500µV/div (just for compare, not affect anything here)
Test signal level is of course 0dBm to step attenuator.
-60dB
Note: C2 is now displaying relevant base noise level with 100 mV/div input.
-65dB
-70dB
-75dB
How high is -75dBm signal here (sine wave of course).
It is ~112 µVpp,  39.8 µVrms, -88 dBV  Think about it, we have 12bit ADC and 100mV/div setting.
(ADC f.s. is here some (small) amount over display vertical range what is 800mV. ADC f.s.)
Let's think with 850mV (not perfect). (850000/4096)  ~208µV  one ADC step!
Still level is well detectable over noise level. We can still least somehow detect levels even 10dB more down as can see in next images.
-80dB
-85dB
(-85dBm signal (average) level distance to displayed base noise average level is visually estimated still roughly 10dB)
Here -85dBm signal level is 35.6 µVpp. (with input set 100mV/div it is really small) and we can see it is still roughly 10dB over base noise average.
If displayed noise average is roughly -95dBm and with 100mV/div maximum level is 2dBm... keep in mind that 16bit ideal ADC dynamic range is 96dB.
As can see Siglent dynamic range is quite good. It have price. Speed.
How this high dynamic range is possible. One important thing is frequency selectivity.
It do not look signal using wide band. Frequency selectivity (RBW) also change depending frequency.
Small extra info: Example freq band 200kHz to 500kHz it use filter what have -3dB width 10kHz and -80dBc ~30kHz
Filter center is tuned during sweep. From 500kHz to 1MHz band filter width is more (but proportionally to freq. same)
There is lot of frequency bands (first is proportionally more wide, 10Hz to 200Hz, then 200 to 500Hz, 500 to 1kHz, 1k to 2k, and so on... and they use different RBW.)
This also reduce signal source spurs and harmonics disturb and also same if DUT generate spurs and harmonics.
Specially when testing frequency changing DUT's, example mixer or more complex circuits (what need 2 channel external SDG when other cannel have same offset whjat DUT produce)
What is FRA (Bode Plot) dynamic range depends of course what noise level (level random errors) we can accept. So it is highly subjective.
Of course, if someone defines what are the limit values according to which the dynamic measurement range is declared, and when everyone uses the same criteria, then of course it is different. For example, limits for the noise of the measured value. But even in that matter, a few other factors affecting the matter should also be defined.
Also it (noise level) depends frequency. Best dynamic range (most low noise level) is around between 50kHz to around 1MHz.
When example R&S tell FRA dynamic range  is ≥70dB or Keysight tell same, over ≥80dB. Do we know if they are comparable. This plain dB value do not tell much.
It is very different if we tell that something is detectable from base noise or something is enough reliable measurable. What is enough reliable/accurate...  it need also define.
Lot of things do not need at all big dynamic range.
So there can use Channel Gain Hold but user need set input channel V/div for maximal DUT out top level without clipping.
Also speed depends how oscilloscope is commincating with generator. Also internal and extrernal generator may have some difference.
Example SDG1000X with USB connection looks like faster than LAN connection and also faster than internal generator even it is also USB bus controlled.
Then with Ghannel Gain Auto, speed also depends signal levels. There is not single truth if someone tell speed. It is valid perhaps only in just same setup and with all same things.
Only single or half truth is, Siglent oscilloscopes FRA speed is slow.
But if some other do not have at all needed dynamics but Siglent have, then this other speed is "infinite" (incapable of the task) and Siglent speed is as it is.
Perhaps, least I hope, we can some day see also "fast mode" in this FRA with all its advantages and especially disadvantages.
Of course some may say there is lot of other kind of instruments specially for do all these FRA things and more, of course there is, and lot of. Also in price only sky is limiting.
I remember my first Rohde&Schwarz SWOB (Polyscop) in decades ago. Expensive analog FRA for up to 1G (so also very different) and still not very stellar, but one good thing, no one did take and borrow it away because it was so heavy (and need also know how to use).
My recommendation is that osscilloscope Bode Plot dynamic range is best to define as dynamic range down from 0dBm level.
Also here if someone want compare different machines.
But then if need print this value to data sheet. There is big problem.
As can see here also in these images one salesman brochure may claim example up to >105dB down from 0dBm and then some other may say that no no it is 80dB.
And both are right.
I have not seen perfect definition for BodePlot dynamic range and what is commond standard or least normal practice and widely in use.
This is just usual oscilloscope and then it have BodePlot. No need numbers for its dynamic range, perhaps Siglent can tell about BodePlot dynamic range simply declaring it -- "adequate".
ETA: for peoples who are not so familiar with dBm here conversion (roughly) to Vpp for these levels used here in previous tests
    0 dBm 632.5 mVpp
  -60 dBm 632.5 µVpp
  -65 dBm 355.7 µVpp
  -70 dBm 200.0 µVpp
  -75 dBm 112.5 µVpp
  -80 dBm  63.2 µVpp
  -85 dBm  35.6 µVpp
  -90 dBm  20.0 µVpp
  -95 dBm  11.2 µVpp
 -100 dBm   6.3 µVpp
 -105 dBm   3.6 µVpp
ETA 2: Added 10Hz - 60MHz  FRA (Bode Plot) base noise level image and some clarifications and typemistakes corr.
« Last Edit: July 07, 2022, 04:49:28 am by rf-loop »
The following users thanked this post: Performa01, Electro Fan, Martin72, mawyatt
Online Martin72
« Reply #171 on: July 06, 2022, 04:38:36 pm »
Preparations for tomorrow, bode plot.. ;)
Won´t be so fancy.
(Kierkegaard)
Online Martin72
« Reply #172 on: July 07, 2022, 01:19:45 pm »
Is there something I must obey when connecting the external Gen ?
In Bode-Plot Menu, you can choose the source, I´ve choose it to "USB".
Pressing "Test" button, scope confirms tha awg is succesfully connected.
But when I start Bode, internal Wavegen Button lit up, on the sdg1062X nothing will happen....
Martin
Edit pics attached, testing time of internal awg reduced although usb was selected and tested, but not used from bode - the litting awg button is for the internal only I guess.
« Last Edit: July 07, 2022, 01:53:07 pm by Martin72 »
(Kierkegaard)
Offline Performa01
  • Country: at
« Reply #173 on: July 07, 2022, 02:47:10 pm »
That's baffling. What FW version do you run?
I never experienced a weird behaviour like that. It should be straight forward:
 - make sure, the AWG is configured for standard operation with sine, maybe channel coupling as required for the reference channel but no I/Q mode or maybe other incompatible configuration!
 - connect the AWG via USB
 - select "USB" in the Bode Plot "Configure" screen. for the internal AWG we would select "Internal"
 - Tap on "Test" and see if you get "AWG connected successfully"
 - start bode plot
It worked for me everytime, on every scope, from SDS1104X-E up to SDS6000 with all FW versions so far. Just tested it once again with my SDS2504X HD and FW 1.2.0.0
« Last Edit: July 07, 2022, 03:48:03 pm by Performa01 »
Online Martin72
« Reply #174 on: July 07, 2022, 03:03:48 pm »
Firmware is 1.2.0.0
Test says ok, when turning on bode plot, internal awg button lit (and remains litting when you exit bode)
See attached "real" pics, in bode mode you can´t make screenshots (Bug imho).
(Kierkegaard)