Retests today, after running the 'Do Self Cal'. Tests run using an oscilloscope calibrator: Wavetek 9500 and 9520 Active Head.
CH 1First are 3 Vpp signals with V/div set to
500 mV/div to establish a baseline:
50 kHz ref200 MHzNext are 5 mVpp signals with V/div set to
1 mV/div:
50 kHz ref* See note below.
200 MHzFinal two are 4.44 mVpp signals (lowest amplitude available) with V/div set to
500 µV/div:
50 kHz ref200 MHzCH 2First are 3 Vpp signals with V/div set to
500 mV/div to establish a baseline:
50 kHz ref200 MHzNext are 5 mVpp signals with V/div set to
1 mV/div:
50 kHz ref200 MHzFinal two are 4.44 mVpp signals (lowest amplitude available) with V/div set to
500 µV/div:
50 kHz ref200 MHz* A note on the triggering. The triggering was a bit erratic (switching between rising and falling) at the 50 kHz ref frequency for the 1 mV/div and 500 µV/div settings. I deliberately set the trigger point to zero, so this may be a harsh case.