The Probe is an agilent Probe with high frequency, didn't know exactly 250 or 300 MHz.
Have also used my 4 Coax-Cables with 50 Ohm terminated. All the same.
Was Sitting all the night when I see error to figure out what's wrong but it seems it is the scope. But what could cause the Problem ?
I have also ooened the scope and compare it with Fotos from Daves teardown of the sds1202x but it looks okay at first overview.
Do you
know really this SDG flatness up to 100MHz. This is SDG2042X.
What you
know about actual real signal level in oscilloscope input terminal. Do you have any instrument for check it.
This because I do not trust this SDG2042X level accuracy with 100MHz (and specially when load reactance is what ever)
One other thing. Do you have good quality feed thry 50ohm terminator. If you have, do test so that you use scope input high impedance with this terminator. Then other test without it and scope internal 50ohm as termination. As short 50ohm coaxial as possible. Also if you have example 10dB atenuator, put it to scope input when use scope internal 50ohm.
Then use (just for reference) 300mV-pp sinewave for test.
In some message you say Hantek cheap scope show it better, you think. How you know it show more right. it is other and good question.
If you can not see difference when you measure with this agilent probe connected directly to SDGoutput BNC and then good coaxial terminated with 50ohm, sorry but it is hard to belive. I expect least some visible difference. Not only between 10 and 100MHz points but also in whole response shape. (with peak mode, roll, and SDG sync you can produce whole response curve.
Until you know what is signal level really in scope input we can not say anything sure.
Is it possible you can use real trusted leveled RF generator.
I have measured several SDS1102X scopes and everyone -3dB point related to 1MHz reference level is always over 140MHz with all V/div settings in two bottom V band. (whole V range is divided to 3 bands)
And for sidemark, I have measured lot of Hantek's (as I previously have sold these, and I teest every single scope what I sell, everyone. I know what its freq response hassle is - really. So if Hantek tell signal flatness is better, it can be true or it can be wrong. Of course it is possible it is more right but, with tens of Hantek experience, sorry but I do not believe anything what this crap show until I really know individual Hnatek what I have measured then I can somehow trust how much it lie. (all scopes lie, indepeendent of brand and price, one less than other etc..
IF this SDS1102X freq response is really out of specification then just send it to repair, it have warranty.
I believe you have not adjusted this front end "accidentally" when you have opened it. Btw, why you open it if you think something is wrong and there is warranty. And even more strange because you are in Germany and Siglent service is also there.
Specifications:
Bandwidth Flatness
DC ~ 10%(BW): ± 1 dB
10% ~ 50%(BW): ± 2 dB
50% ~ 100%(BW): + 2 dB / -3 dB
But first need measure it really.