Author Topic: Siglent SDG6000X series 200-500 MHz AWG's  (Read 101251 times)
Offline JanL
Re: Siglent SDG6000X series 200-500 MHz AWG's
« Reply #475 on: November 30, 2023, 03:20:29 pm »
Hello there, I stumbled across the SDG6022X as an signal generator for a reasonable price. Looking at the data sheet I am missing information about rise/fall times for different output voltages. Does anybody know what times to expect for 5, 10, 20V outputs?
Thanks, Jan
Online Performa01
  • Country: at
« Reply #476 on: November 30, 2023, 04:36:47 pm »
I only have SDG6052X here, but this shouldn’t make any difference:
Firstly, I won’t be able to show you the rise time at 20 Vpp, because this instrument, like any other AWG in this class, can only output 10 Vpp into a 50 ohm load. There’s certainly no way to get a fast pulse with single digit nanosecond risetime into a high-Z input without significant degradation. So we have to stick to 50 ohms.
The fastest specified rise-time of the SDG6000X is 2 ns. At least my specimen is a little faster and manages to stay below 1.7 ns at 5 Vpp:
SDG6052X_5Vpp_T2ns
For 10 Vpp it was a bit of a hassle already, because that level is too high for a 50 ohm DSO input. I had to fit a 6 dB inline attenuator and consequently you’ll still see only a 5 Vpp pulse in the screenshot. But it is 10 Vpp originally – you can also see that the rise time got quite a bit slower now (~2.3 ns).
SDG6052X_10Vpp_T2ns_A6dB
« Last Edit: November 30, 2023, 04:50:26 pm by Performa01 »
The following users thanked this post: egonotto, Martin72
Offline JanL
« Reply #477 on: December 01, 2023, 08:03:50 am »
Thank you very much. It helps a lot with my decision.
Online Performa01
  • Country: at
« Reply #478 on: May 16, 2024, 06:46:42 pm »
Attached is an example for the SDG6052X after a power loss.
It was set to a 100 Hz, 1 Vpp sine wave.
Utility – System – PowerOn was set to “Last”.
Utility – Output Setup – Power On State was set to “On”
Utility – Output Setup – Amplitude was set to 20.000 Vpp, i.e. no limit set.
That’s how a well behaved AWG handles mains power outage and -return.
SDG6052X Sine_100Hz_1Vpp_Power Loss
« Last Edit: May 17, 2024, 07:53:15 am by Performa01 »
The following users thanked this post: egonotto, Martin72, Detlev
Online Martin72
  • Testfield Technician
« Reply #479 on: May 16, 2024, 07:02:49 pm »
What I ask myself all the time is whether the "gap" between SDG2000 and SDG6000 will ever be filled....
(Kierkegaard)
Offline nctnico
« Reply #480 on: May 16, 2024, 07:25:25 pm »
Was this measured with peak-detect enabled? At 4Ms/s it is easy to lose a spike.
Offline 2N3055
  • Country: hr
« Reply #481 on: May 16, 2024, 08:34:20 pm »
There is no spike. Simple as that.
Offline nctnico
« Reply #482 on: May 16, 2024, 08:51:36 pm »
The scope screenshot doesn't show that information so if you have no useful information to add, it is better to wait for Tautech to acknowledge he enabled peak-detect in order to do this kind of measurement correctly or to redo the measurement with peak-detect enabled and show a new screenshot.
« Last Edit: May 16, 2024, 08:55:30 pm by nctnico »
Offline 2N3055
  • Country: hr
« Reply #483 on: May 16, 2024, 09:11:38 pm »
You just have to be condescending aren't you?
I MEASURED. There is no spike. With Peak detect, without, with different scopes.
I have that AWG for 3+ years...
I would not post that if I didn't have veritable information and was sure...
Haven't you learn that by now?
The following users thanked this post: Performa01, egonotto, Jacon, KungFuJosh
Offline Njk
  • Country: ru
« Reply #484 on: May 16, 2024, 10:15:16 pm »
It's a trivial case that every dealer can show. But the fact that it is able to keep the zero offset does not prove that it is also able to preserve a user-defined DC offset (i.e. the logic levels). While the latter is obviously of more importance
Offline tautechTopic starter
« Reply #485 on: May 16, 2024, 10:25:31 pm »
Not my screenshot above.
However I might have done it a little different using a Normal Trigger waiting for the signal to return after power ON.
But not today, too much to arrange for EMEX the week after next.....
Out of office and unavailable for a few days.
Offline nctnico
« Reply #486 on: May 16, 2024, 10:38:26 pm »
Well, you should have written that to make your point clear to everyone. That is useful information.