Author Topic: Siglent Bench DMMs SDM3065X, SDM3055 and SDM3045X  (Read 455404 times)
Offline KungFuJosh
  • TEAS is real.
Re: Siglent Bench DMMs SDM3065X, SDM3055 and SDM3045X
« Reply #1900 on: August 17, 2024, 05:39:29 pm »
Following up from the other thread talking about SDM3065X stability:
After 1 hour warm up time with both the meter and the reference on.
10.00460 (claimed by reference seller - not certified)
10.0039? on SDM3065X
I will post again after some hours.
ETA: TestController log rate is set at 0.1s.
Offline TheDefpom
  • YouTuber Nerd - I Fix Stuff
    • The Defpom's Channel
« Reply #1901 on: August 17, 2024, 07:17:41 pm »
It is FAIRLY settled after an hour, it is still warming up for about 4 hours, after that although there is still some movement but it is small.
Cheers Scott
Offline KungFuJosh
  • TEAS is real.
« Reply #1902 on: August 17, 2024, 09:12:25 pm »
Okay then. I've passed the 4 hour mark now. So I'll post the info now, then reset the data and let it run for 4 more hours and see if there's a difference.
Note that there were some minor changes in temp and RH.
« Last Edit: August 17, 2024, 11:11:11 pm by KungFuJosh »
Offline bdunham7
« Reply #1903 on: August 17, 2024, 09:44:50 pm »
ETA: If we pretend to believe the stated value of 10.0046, then the bottom number to stay within would be 10.000197976. Seems safe. 🤷
10.0046 would be displayed as 10.00460 on a 6.5-digit DMM and is 4.6mV or 460ppm over the nominal 10V.  If your combined tolerance is specified as 48ppm of 10V or 480µV (40ppm  * 10V + 4ppm * 20V) then your lower reading limit would be 10.00460 - .000480 or 10.00412.
Offline KungFuJosh
  • TEAS is real.
« Reply #1904 on: August 17, 2024, 10:57:10 pm »
Apparently I'm still confused. If the resulting accuracy for this range is:
0.0040 + 0.0004 = 10 x (0.0040 / 100) + 10 x (0.0004 / 100) = +/-0.00044
So you're saying that the specified .00044 is an offset number instead of a percentage number? Damn it. Okay, I guess that makes more sense than another percentage.
Thank you. Learning is fun? lol
I guess there's no point in worrying about the 10.00460 since it's a bullshit number. Still pretty close anyway.
« Last Edit: August 17, 2024, 10:59:21 pm by KungFuJosh »
Online Martin72
  • Testfield Technician
« Reply #1905 on: August 18, 2024, 12:17:07 am »
You can imagine why I want to have my reference (yes, it's not forgotten) calibrated externally...
And even then it wouldn't be “perfect”.
In the calibration protocol of my SDM3065X, the measurements were done like this:
Fluke voltage reference set to the range, measured the actual current value with a Keysight 8.5 digit meter, then measured with the SDM3065X and its displayed value is compared to that from the Keysight.
That would be the way.
Then the reference value would be almost irrelevant, as you measure with a much more accurate meter.
(Kierkegaard)
Offline KungFuJosh
  • TEAS is real.
« Reply #1906 on: August 18, 2024, 12:50:09 am »
Weeeeeeellll, I think this test is pointless. My ref isn't stable enough, nor is the environment. PP was solid at 3ppm for the first two hours. Then after I sat in the room for a bit, it jumped to 3.6ppm. Then when the air conditioner shut off, it jumped again up to 4.2ppm.
None of that is bad of course, but for a tighter tolerance I'd need an enclosed ref and better environment control.
You can see the jump in the chart after the AC shut off.
« Last Edit: August 18, 2024, 01:12:01 am by KungFuJosh »
Online mawyatt
« Reply #1907 on: August 18, 2024, 12:55:46 am »
Think you are showing the same long term plot twice, they both look identical!!
Best,
~Wyatt Labs by Mike~
Offline KungFuJosh
  • TEAS is real.
« Reply #1908 on: August 18, 2024, 01:00:01 am »
Nuh uh! 😉
The air conditioner came back on, and the voltage went back down (or at least started to before I shut it off).
« Last Edit: August 18, 2024, 01:11:30 am by KungFuJosh »
Online mawyatt
« Reply #1909 on: August 18, 2024, 12:43:17 pm »
Just noted that in all the SDM3065X plots this is showing SD (assume Standard Deviation) as 0.000000V. Don't think the SDM3065X, nor the Reference are quite that good noise wise!
Best,
Offline KungFuJosh
  • TEAS is real.
« Reply #1910 on: August 18, 2024, 02:16:24 pm »
I assumed that was a bug when I noticed it, but also wasn't going to stop a 4 hour test to upgrade TC. 😉
Nope. Another PEBKAC error. Apparently I needed to go to the Current Values tab, right-click and increase the number of digits and/or choose SI for SD to work.
« Last Edit: August 18, 2024, 04:58:14 pm by KungFuJosh »
Offline KungFuJosh
  • TEAS is real.
« Reply #1911 on: August 18, 2024, 03:56:56 pm »
Tomorrow I'll test both from cold start. Or should I warm up the ref first, to separate that? It makes sense to me to warm up the ref first remove the ref's drift from the test.
Ideally you'd do it both ways--with the ref warmed up and meter cold then vice versa.  But try it the first way first.
Looks like 30 minutes for mine if uV resolution is needed. Immediately useable if only needing mV.
I assume if I need super accuracy I should stick with what Scott said and warmup for 4 hours. Or get solar panels and never turn off the meter. 😉
Offline KungFuJosh
  • TEAS is real.
« Reply #1912 on: August 18, 2024, 07:03:43 pm »
Looks like the ref is pretty good after 5 minutes or so, depending on the environmental stability.
I also decided to point a hot air wand at it. 100C at 50% power. Alternated a couple times between 3 to 6 inch distance.
Offline bdunham7
« Reply #1913 on: August 19, 2024, 01:25:01 am »
Actually those results are pretty good, in fact better than the typical 34401A. I wouldn't worry about warmup times now that you've characterized it.  Anything after the 5-minute mark is a very small proportion of the specified tolerance so TC and long-term drift will dominate any measurement uncertainty.  IMO for 99% of uses you can use it a soon as it boots and for the rare occasion where you want maximal stability 5-10 minutes is plenty.  If you are doing single-ppm relative measurements or calibrating the meter, then you might go for a 2 hour warmup.
Now you need to independently verify its absolute accuracy!   :)
Offline KungFuJosh
  • TEAS is real.
« Reply #1914 on: August 19, 2024, 02:27:50 am »
It's still in cal, and I don't have a better reference than the one I used. It also agrees with my other meters. Good enough, or you got some fun suggestions? 😉
Offline KungFuJosh
  • TEAS is real.
« Reply #1915 on: August 30, 2024, 11:54:50 pm »
I'm a schmuck and decided to spend the money on a better ref that I don't need. I bought the VREF10-001 from Doug. He clearly does good work.
I captured the 1st, 2nd, and 3rd hours after 4 hours of warmup. The first hour was probably the most accurate, as I was back and forth working on my bench in the following two hours. Even with my interference (including soldering 1 foot away), the stability was excellent.
It was also much less influenced by environmental change as you can see by the minor shift related to the temperature change. With the previous (cheapo) ref, that would have been much more significant. The bigger temperature shift after about 1.5 hours was when I closed my office door and the AC became more effective.
« Last Edit: August 31, 2024, 12:15:21 am by KungFuJosh »
Online Martin72
« Reply #1916 on: August 31, 2024, 12:00:41 am »
(Kierkegaard)
Offline KungFuJosh
  • TEAS is real.
« Reply #1917 on: August 31, 2024, 02:28:59 pm »
I was impressed with the results so I emailed Doug to thank him. In his reply he said "It is interesting that the Avg reading was within 1uV during all 3 tests;  the uV noise on the Reference output is low frequency noise inherent to the LT1021 Reference chip- it looks like your DVM does an excellent job of averaging."
Does the SDM3065X do any averaging outside of the statistics box?
Online tautechTopic starter
« Reply #1918 on: August 31, 2024, 04:22:32 pm »
What NPLC settings are engaged ?
Any Math engaged ?
Offline KungFuJosh
  • TEAS is real.
« Reply #1919 on: August 31, 2024, 04:52:47 pm »
Any Math engaged ?
10PLC and the statistics box was open.
« Reply #1920 on: August 31, 2024, 04:57:24 pm »
Any Math engaged ?
Yeah you said.
No Min / Max / Average as well ?
Offline KungFuJosh
  • TEAS is real.
« Reply #1921 on: August 31, 2024, 05:00:45 pm »
That's in the statistics box. I haven't setup anything else. It's always like this old screenshot:
Offline vulubalulu
  • Country: af
« Reply #1922 on: September 16, 2024, 12:01:08 am »
I was scouting the internet for a benchtop dmm for instantaneous autoranging for voltage and resistance. I saw a video of a sdm3045X being really quick and it beeps shortly when it detects voltage, i like that feature too. But I feel like it's overkill for my application. I just repair non-audio electronics as a hobby. At most I measure AC that comes out of the socket and usually it's DC voltage ranging from 1v to 50v. I don't really need so many counts either, i just want it fast. the 45x looks really neat but at almost 450€ the price point is high. Are there alternatives you can recommend. Please advise.
« Reply #1923 on: September 16, 2024, 12:59:15 am »
369 Euro plus whatever taxes is the list price.
https://www.siglenteu.com/digital-multimeters/sdm3045x-digital-multimeter/
Offline vulubalulu
  • Country: af
« Reply #1924 on: September 16, 2024, 01:58:40 am »
the 369 Euros are without tax. it comes down to 450