FW 6.01.01.18 to 6.01.01.19
HW B-64-5.1 ( seems to have changed to B-64-5.2 with P19 )
SN SDS10BA216****
Trigger Level line going off with change of FW. #8,#13
Off by only 0.2 Div. P18 -> P19
Missing Dots Persistance, with 1CH 100ns & 50ns/Div, with 2CHs just missing from 100ns/Div. #2,#13
Missing Dots Persistance/LongMem 25ns/Div and 2.5ns/Div. #22
Unstable ETS mode, lockups, freezes. When switching to ETS, the freeze can be so fast that it doesn't even have time to change the menu from "Real Time" to "Equ Time".
Jumping Waveforms on ETS STOP. #15
STOP works differently on Normal/Long Memory. #22,#30
LongMem corrupted displays. #24,#30,#34
LongMem Option disappeared. #35
Math sensitivity goes to 1pV/Div. #42
Making any changes to Math, including just returning to the Math menu, resets the Math sensitivity to that of the scope channels, but it doesn't seem to take into account the probes' X10 setting.
Trigger X position sometimes moving with X expansion on STOPed waveforms. #45
ETS waveform STOPs on the opposite trigger edge polarity. #51
Channel Skew adjustment jumping/not saving. #52
EXTtrig ignores Channel Skew values. #52
EXTrig "Set to 50%" sets to 0V. #52
Intermittent Saving of REF 'A' or 'B' waveforms, broke P18 -> P19. #69
- Improved by itself, just an occasional unexplained "Function isn't useable."
Unexplained lockups/freezes when using Scan Mode on slow timebases 1 to 10 Sec/div.
10S/Div. the worst, will lock up on power up after the splash screen if I don't quickly turn the TB up to 50ms/div.
In Scan Mode touching almost any control causes the trace to restart, about the only control you can touch without losing the captured scan is RUN/STOP.
Minor oddities #50
The messed up ALT Time Units Fixed in P19.
Pressing STOP while in Math,Dots+Persistence mode giving the whiteout. #41
Maybe Fixed in P19.