You may find the following of interest as well where we look at several types of gates ran over various temperatures and voltages.
https://www.eevblog.com/forum/projects/waveforms-in-a-74ls04-ring-oscillator/
Offline PartialDischarge
« Reply #451 on: November 08, 2022, 03:58:24 pm »
Somewhere in a thread user ttggzz suggested the use of LVC gates in parallel, I made a board to test the idea and I'm getting less than 500ps rise and fall times with very nice shape, tested with a 1GHz scope:
« Last Edit: November 08, 2022, 04:02:17 pm by PartialDischarge »
The following users thanked this post: egonotto, radiolistener, teddychn
Offline tggzzz
« Reply #452 on: November 08, 2022, 04:35:35 pm »
Here perhaps: https://www.eevblog.com/forum/testgear/show-us-your-square-wave/msg1902941/#msg1902941
Good to see the results can be replicated. It looks like the scopes I was using (not mine, very fast test) used the 20%-80% times, which would be faster than the 10%-90% times.
I used 143ohms in series since the gate output is ~7ohms. I don't think that would make too much difference in practice - decoupling would make more difference.
« Last Edit: November 08, 2022, 04:38:49 pm by tggzzz »
« Reply #453 on: November 08, 2022, 05:41:38 pm »
I tried with a 74VHC gate but it is slower, I'm very happy with the results since 2.5V 500ps is very useful for many applications, it is a fairly cheap solution and aberrations are minimal which I consider important
Offline tggzzz
« Reply #454 on: November 08, 2022, 08:27:51 pm »
Agreed.
You can also reduce the output swing by reducing the PSU voltage, but the transition time increases. Try reducing Vcc  to 1V or so, and seeing if you get a 0.5V swing with 3ns risetime (and some "pre-undershoot" for want of a better term). You can go even lower, with some entertaining[1] aberrations :) The aberrations are gone by 1.7V or so, as expected.
[1] I'm easily entertained :)
Offline tchiwam
« Reply #455 on: January 07, 2023, 08:33:51 pm »
4 outputs on top of each other ;)
100us pulse 1 PPS rising edge ~3GHz BW
Offline tggzzz
« Reply #456 on: January 07, 2023, 08:41:02 pm »
4 outputs of what?
Why not use averaging?
PPS? And those figures don't seem to match the figures in the photograph.
Offline tchiwam
« Reply #457 on: January 07, 2023, 09:21:58 pm »
I was comparing 4 of the 8 putputs from my PPS distribution. That's s the rising edge of a 100us pulse.
The measurements on the side are out of the window so they are wonky except for the cursors.
Offline bozidarms
  • Country: at
« Reply #458 on: February 05, 2023, 03:54:17 pm »
on CSA830A, my new toy :D  - 20,8ps -  rise and fall time!
« Last Edit: February 05, 2023, 10:02:02 pm by bozidarms »
Offline HighVoltage
« Reply #459 on: February 05, 2023, 04:11:11 pm »
@bozidarms
20.8 ps rise and fall times ?
What generator do you use for that?
And what scope and probe?
Offline bozidarms
  • Country: at
« Reply #460 on: February 05, 2023, 04:26:48 pm »
It`s one Tektronix CSA803A with SD-24 TDR/sampling head.
However, that is my 3. SD-24, which has characteristics far more better than proclamated.
The other two wasn't so god.
That is a test pulse from SD-24 itself on SMA out, which is only shorted.
Measurement is at 10% - 90%.
« Last Edit: February 05, 2023, 05:20:45 pm by bozidarms »
The following users thanked this post: egonotto, tchiwam
Online Aldo22
« Reply #461 on: March 08, 2023, 03:42:38 pm »
- kind of curious to see what is the lowest price device is that can produce a really good looking square wave at 1, 5, 10, 20, and 30 MHz?
I don't know if "really good looking" but it's very low priced.  ;)
The signal comes from a ESP32 microcontroller  for ~ 5$.
The oscilloscope is a handheld ZEEWEII DSO2512G for ~ 80$.
Please note:
The 40MHz measurement was made in the overclocking mode of the device (Menu Aux + ⇧ [long])
with 250MHz probes and a ground spring.
« Last Edit: March 12, 2023, 08:24:34 am by Aldo22 »
Online the Chris
« Reply #462 on: November 10, 2023, 07:38:40 am »
My pulse generator from Leo Bodnar arrived yesterday, so I would like to contribute with my slow-but-I-am-loving-it EDUX1052A. Measurements were taken with an RS Pro 50Ohm inline terminator that I ordered from Leo alongside the pulser.
I got 5.8767ns over 112k measurements. Considering the specified 50MHz and therefore assuming a classic -3dB drop, the BW = 0.35 / r_t should be valid, resulting in 60MHz bandwidth.
This is in line with what I estimated using a 10MHz to 60Mhz sinus sweep of my UTG962E, with pretty much no drop under 10MHz (not shown) and about ~1.2V peak in amplitude at 60MHz using a N2142A 75MHz probe in 10x mode.
Thanks Leo for the fast shipping and now including the customs and taxes loads for EU members!
EDIT: 5ns/div measurement added
« Last Edit: November 10, 2023, 08:02:30 am by the Chris »
The following users thanked this post: egonotto, Martin72
Offline David Hess
  • DavidH
« Reply #463 on: November 11, 2023, 02:12:48 am »
I wonder where that aberration came from; it is much greater than I would expect considering the much faster settling time of the pulse generator compared to a 60 MHz oscilloscope.
My guess is that the oscilloscope does *not* have a Gaussian or single pole response.
The following users thanked this post: egonotto, the Chris
Offline Sighound36
« Reply #464 on: November 11, 2023, 01:42:51 pm »
Below are four examples of square wave forms I looked at earlier, one using the Siglent SDG2042A 10Mhz 4V pk<>pk / 10Mhz ultra low phase noise -124dbm@1hz to -175dbm@10Mhz/ Leo Bodnar 40pS rise time generator/ 44.1Khz word clock.
Using Wavepro 4Ghz HD scope 50 Ohm T-flex 405 18Ghz cabling, full scope BW of 4Ghz NO ERS (software enhancement) up to 20GS/s rate. Scope up to temprature and calibrated.
« Last Edit: November 11, 2023, 02:08:09 pm by Sighound36 »
Offline Sighound36
« Reply #465 on: November 11, 2023, 02:11:33 pm »
As we can only post 10 attachments second post last of the 44.1Khz images plus I thought if would be nice to show the eye pattern for each of these square wave sources as well
Offline joeqsmith
« Reply #466 on: November 11, 2023, 04:52:03 pm »
I followed  Leo Bodnar's threads about that pulsar but never bought one.   All of my scopes are too slow to verify it.
Many years ago, I stumbled onto an employee of LeCroy who had purchased some of their junk at an internal auction.  They had the test fixtures for the 7200A.  Another gem was they had some of the 7262 plugins.  These are the only ones I have ever seen.   This plugin includes an ECL output for TDR measurements and has an analog BW of 4GHz.   I posted some data for it recently where it measures 118ps.
https://www.eevblog.com/forum/testgear/12-ghz-active-probe-project/msg5007958/#msg5007958
I posted about my conversations with LeCroy about the risetime and bandwidth:
https://www.eevblog.com/forum/testgear/show-us-your-square-wave/msg593071/#msg593071
Typically RT*BW = 0.40 to 0.45 for modern high bandwidth scopes
4GHz * 118ps = 0.472.   Much higher than 0.35.
If I attach another old relic, Tektronix S-52 which has a risetime of less than 25ps, the scope measures 111ps.   Or 0.444.
This scope was my first DSO, outside of my HP scope/logic analyzer.   Originally owned by the USAF.   Released in 1989.   She's old but fun to play with.
The following users thanked this post: DH7DN
Online Grandchuck
« Reply #467 on: November 11, 2023, 08:22:01 pm »
Here is a result from a Tektronix Scope Evaluation board and a Siglent DS2202.  The board has a 4-bit counter with a risetime of around 2 nsec with some overshoot.
The following users thanked this post: joeqsmith
Offline tggzzz
« Reply #468 on: November 11, 2023, 09:15:58 pm »
How was the probe's shield connected to the board's ground? A 6inch ground lead is a classic cause of overshoot.
Online Grandchuck
« Reply #469 on: November 11, 2023, 09:50:35 pm »
Your guess is close ... it is a 5 inch lead.  They put the ground pad as far from TP1 as possible???
I included a shot from the manual that came with the board.
Offline joeqsmith
« Reply #470 on: November 11, 2023, 09:55:30 pm »
That's funny to even see one of these boards.   I picked up what appears to be the same one directly from Tektronix when I bought my first scope (Hitachi) back in the 80s.   I think they were giving these away back then as a promotion.
Connecting from pin1 to pin4 using a LeCroy PP061 resistive probe with a pretty much dead battery,  scope measure 1.65ns without the big overshoot.  The BW is much higher than what you have available and there is no ringing.  I also suspect your connection from the board to the scope.
The following users thanked this post: Grandchuck
Offline tggzzz
« Reply #471 on: November 11, 2023, 10:04:46 pm »
Have a look at https://entertaininghacks.wordpress.com/2015/04/23/scope-probe-accessory-improves-signal-fidelity/
As a quick test I would wrap a length of wire around the probe body, with a short straight spike to the gnd hole on the left. In other words, a variant of the "spiked spring" supplied with the probe.
Offline joeqsmith
« Reply #472 on: November 12, 2023, 12:35:14 am »
Recently someone was asking about how to use a scope and signal generator to measure the self resonance frequency of an inductor.    I provided several examples, the later using a square wave to cause the inductor to ring at at the SRF.
https://www.eevblog.com/forum/beginners/measuring-the-self-resonant-frequency-of-an-inductor/
As long as I have the pulser out... shown is a Mini-circuits 6.3 - 15GHz high pass.  Data sheet may be found at:
https://www.minicircuits.com/pdfs/VHF-6010+.pdf
I attached this to the old Tektronix pulser and then into my LeCroy 8500A.  This scope has a BW of 5GHz.  They claim a typical rise time of 90ps.   Obviously the filter is outside the limit of what the scope can measure.  Ignore that.  The signal was stable and I used RIS to get a clearer picture what is going on.   What I thought was interesting is it appears to have a dominant SRF of about 6.2GHz.
In the end, who smacks a filter like this anyway....
Online Grandchuck
« Reply #473 on: November 12, 2023, 03:48:08 pm »
tggzzz and Joe are great motivators.  Look and see what happens.
Now, why would Tektronix do that?  One would think that their pictured result would exemplify best practices???
Offline joeqsmith
« Reply #474 on: November 12, 2023, 05:15:50 pm »
Well ... Back in the 80's my mind set was a PC with a clock frequency of 50MHz, never going to happen.    I will never need a scope with more than 100MHz BW.  ....  Our high end scope where I worked was a large 100MHz Tektronix DSO.  I was using a Tektronix Polaroid camera to record data...   I'm not surprised at all by the demo boards layout.  Wouldn't surprise me to find out it was an intern's project.
The trigger and missing pulse test was a problem for some of our scopes at that time.   :-DD
I keep a spook of wire on-hand to make these springs from scratch.  1) I start by tightly wrapping the probe's ground.  2) I remove the spring and give it a half turn.  Round need nose pliers work fine.  This gets enough tension on the spring to lock to the probe.   3) Ground strap is made from copper foil.  Sometimes I use braid.   All depends...  4) Solder the spring to the strap.  5) For insulation, I use Kapton.  With braid, some heat shrink or woven high temp fabric.   Depends.    6) Solder the ground strap to the board we leaded solder.  I won't use that lead free stuff for home use.    7) attach probe and measure.
Still not a great setup, but consider these are pretty slow edges we are dealing with.  Here was a home made probe where we were playing with much faster edges.   I was attempting to get the ground inductance and loop as small as possible.
https://www.eevblog.com/forum/testgear/12-ghz-active-probe-project/msg5006290/#msg5006290