Dear Fixpoint,
Thank you for your extensive comments, appreciate it. I learned several things from what you wrote.
But on some other, major issues, my experiences / observations are, however, not the same as yours.
The NGE103B is an entry-level instrument and does not feature regulated load sharing.
Your comment implies that it was unreasonable of me to expect, in this price class, a parallel mode in a PSU that would not require the user to manually have to achieve load balancing for the device to work. To find out whether that was true, I now searched more specifically for parallel mode implementations of devices at the same or lower price level. A
video of the Keysight E36313A, the main alternative I considered, with a similar list price of the NGE with similar options, suggest that that PSU works precisely as per my expectations, with no requirements for the user to balance loads across channels. Also has relays that set up the internal wiring. Checking a bit further, the Keithley Model 2231A-30-3 (considerably cheaper than the NGE103B)
also works this way. Heck, even the much, much cheaper Siglent SPD3303X-E performs automatic load sharing between its channels in parallel mode (see the following this video at 41:55 -
https://www.youtube.com/watch?v=MwSO9cu5c1o. Also this one has the internal relays to set up parallel wiring.
The idea is the following: Suppose you need up to 7A at 5V. Then, you should set all channels to 5V and limit each of them to 2.4A, so that the device is forced to spread the 7A more evenly.
So, I just performed a test following your line of reasoning. Each channel is advertised to deliver up 32V, up to 3 amps, where the total power per channel needs not to surpass 33.6 Watt. So, at 12V, the max. channel load is 2.8 amps (which is indeed exactly what the user interface shows us when 12V is selected).
So I connected a load that at 12 Volt draws 7.35 amps. At 7.35/3=2.45 amps per channel, that is below the max. rating. I set each channel at 12V, and, in line with your suggestion, I then set the current for each channel lower than the max rating, at 2.6 amps. When I turn the outputs on, and initially all three channels are still nicely 10% below their maximum load rating (<30 of a maximum of 33.6 Watt). But one channel gives up after 25 seconds (turns into OPP mode) and subsequently the other channels jump into current limiting mode and drop from 12 to 5.4 Volts.
Again, we are clearly still in the devices specs here...
I'd be really interested to see whether you (or someone else) could replicate this simple test, to see whether its just my unit that keeps turning of channels while used within the specifications, or whether you find the same. For the load, for convenience, I just took four halogen 12V light bulbs....
I must say that low(er) loads, the current limiting method you suggest does work, whereas using the method listed first by R&S in their manual (
“By increasing the voltage slightly, the load distribution can be manipulated”) I could not get stable channel load balancing at all and channels kept turning off even at lower loads.
what you are saying implies that you didn't pay enough attention.
I understand that, depending from the perspective you come from, this 1200 Euro power supply is an entry-level instrument. But if a PSU is prominently advertised to be able to deliver up to 9 amps (within the limitations of the overall channel power), then the user should be able to trust on it that it does so, in a reasonably uncomplicated manner, and following the information in the user manual. Not that channels turn of after 30 seconds. I want the device to meet the specifications as advertised.
I don't think that is a matter of not me not paying enough attention.
Since you can edit each digit of the settings individually, you can make quick adjustments of the settings without a numerical keypad. Many adjustments are even quicker this way than with a numerical keypad. You should make use of that feature.
Yes, the digit entry feature is useful. But this entry mode is not available when you most need it - for instance if want to make a number of points in the TrueArb table. No idea why this entry mode does not work here, but the only way to populate this table is painstakingly dialling in all voltages, currents and times with the rotery dial. If you know a way how to activate the direction keys in the TrueArb table mode, then please share with us how that is done!
The UI is far from intuitive. Not having dedicated channel on/off buttons
>>> What?? Of course it has those buttons, and they are clearly labeled and prominently placed! Just have a look.
I really don’t see dedicated channel on/off buttons. There are three buttons, labelled “Ch 1”, “Ch 2” and “Ch 3” respectively, but dependent on the state the device is in, these buttons serve to select channels for voltage setting, for current setting, for tracking selection, for "fuse" settings,
or for channel on/off. So these are not dedicated channel on/off buttons.
Concerning the manual, well, it may be well true that this device is closely related to other devices that R&S makes. But this particular manual is for the NPE100B series devices only. So I find it strange that it refers to the touch screen or mouse controls, while the NPE100B series certainly has neither of these.
There is still enough space between the unit and the next one in the stack, this shouldn't be any problem.
I would be happy if I this was the case. But on this subject, the R&S user manual (p. 19) states
“The R&S NGE100B is fan-cooled and must be installed with sufficient space on the top to ensure a free flow of air. Required minimum distance: 1 rack unit (RU).” One rack unit (generally known as 1HE) equals ~4.4 cm. The distance between stacked NGE103B devices is only around 1.0 cm so way below the required 4.4 cm.
For a change, I can tell you a REAL nuisance with the NGE103B: It's fan control algorithm is not very good. After running for some time (it does not matter whether there is a load or not), it enters a control loop that starts and stops the fan periodically even when there is no load (at room temperature). R&S have told me that they have this problem in their backlog, however I don't know whether/when they will fix it.
Thanks for mention this - even if I'm looking for solutions, no more problems
I now left the unit on for a most of the day, without any load, and indeed it gets into a loop of turning the fan on and off.