How common is it in a scope, where you sample and hold 2 channels together....and they are not positioned properly with ref to each other in time...eg one is incorrectly time shifted? Is this common?
Say you have two signals, where they definitely pulse up at the same time.....but in the scope shot, they are not sync'd....because the scope is in error. Is this common?
Please also consider when say one channel is a big signal on 200x and 100V/sec, and the other channel is say 1V/div and on 20x magnification....is this error more likely then?
« Last Edit: October 24, 2023, 06:28:22 pm by Faringdon »
'Perfection' is the enemy of 'perfectly satisfactory'
Online ataradov
    • Personal site
Re: Scope waveforms "out of sync"
« Reply #1 on: October 24, 2023, 06:24:57 pm »
What? No, this is not common. This is not even a thing at all.
There may be some propagation delay differences between the channels, but they are not going to be significant compared to the sample rates.
Alex
The following users thanked this post: Faringdon
Offline Gyro
« Reply #2 on: October 24, 2023, 06:32:36 pm »
Connect both probes to the same node and post a picture of the waveform and probes connected together. Only then can the bad workman stop blaming his tools and work out why the signals he thinks "definitely pulse up at the same time" in fact don't!
EDIT: This may be related to https://www.eevblog.com/forum/beginners/high-voltage-seen-when-scoping-with-diff-probe/ who knows.
« Last Edit: October 24, 2023, 07:48:22 pm by Gyro »
Best Regards, Chris
The following users thanked this post: boB, Faringdon
Offline wasedadoc
« Reply #3 on: October 24, 2023, 06:37:21 pm »
It can happen.  See around time 10:23 in https://www.youtube.com/watch?v=9cbK5o_7D6U&ab_channel=TechCornerTVElectronics%2CProjects%26Tutorials.  Review of a small oscilloscope.
Online ataradov
    • Personal site
« Reply #4 on: October 24, 2023, 06:45:03 pm »
Ok, I assume we are talking about actual tools, not random junk for $20 from Aliexpress. Obviously absolutely anything can happen there.
Alex
The following users thanked this post: 2N3055, Faringdon
Online 2N3055
  • Country: hr
« Reply #5 on: October 24, 2023, 06:52:36 pm »
If you are using different probes (especially if one of them is active DIFF probe and other one is passive one) then both of these probes will have different propagation delay. Also if cables are not same length. Or using combination of voltage and current probes...
Proper scopes have Channel Deskew function in channel setup to compensate for that.
The following users thanked this post: Performa01, pdenisowski, Faringdon
Offline ebastler
« Reply #6 on: October 24, 2023, 06:56:55 pm »
@Faringdon: What's the scope model? Could you share a screenshot which shows the settings and the problem?
Offline gf
« Reply #7 on: October 24, 2023, 08:07:30 pm »
The Hantek2000, for example, has such a problem (1/2 sample offset between the two channels). In single-channel mode, the two halves of the AD9288 are interleaved (in order to double the sample rate), so of course the two clocks must be interleaved as well. However, it seems that the clocks are still interleaved in dual-channel mode :(
Offline Gyro
« Reply #8 on: October 24, 2023, 08:17:11 pm »
From the other thread, I suspect that the OP is working at low frequency, so ADC interleaving shouldn't be an issue.
Best Regards, Chris
Online ataradov
    • Personal site
« Reply #9 on: October 24, 2023, 08:20:36 pm »
This is what I understood under "insignificant". If 1/2 sample starts to matter, you need to get a better scope.
But as usual with OP, we are not likely to see any answers. He will just make another pointless thread.
Alex
The following users thanked this post: tggzzz, CustomEngineerer, Gyro, Faringdon
Online pdenisowski
  • Product Management Engineer, Rohde & Schwarz
    • Test and Measurement Fundamentals Playlist on the R&S YouTube channel
« Reply #10 on: October 24, 2023, 08:24:52 pm »
I was just about to say the same thing :)  Deskew fixtures and compensation exist -- even in mid to low range scopes -- for a reason.
If you're probing the same point with two identical voltage probes and see offset/skewed waveforms, I would definitely find that odd.
Test and Measurement Fundamentals video series on the Rohde & Schwarz YouTube channel:  https://www.youtube.com/playlist?list=PLKxVoO5jUTlvsVtDcqrVn0ybqBVlLj2z8