Difference between HP / Agilent 435A and 435B Power Meters
« on: July 22, 2024, 03:10:00 pm »
Hello HP/Agilent fans and experts,

Can someone please explain me what is the difference between the  HP/Agilent 435A and 435B Power Meters?
Both power meters are of analog type, looking identical from the outside. For at least one sensor (HP 8484A) the HP recommends use of 435B power meter version only. I checked HP site and did not found any useful info about that.

Thanks for your help!


 

Offline johnk0gcj@gmail.com

  • Contributor
  • Posts: 12
  • Country: us
Re: Difference between HP / Agilent 435A and 435B Power Meters
« Reply #1 on: July 22, 2024, 06:02:18 pm »

Good question. I've had a couple 435A's over the years and had seen but never had 435's, and had
assumed that the difference was the interchangeable scales on the 'A'. However, I recently obtained
another 'A' model, and it has the non-interchangeable scale. At first I thought someone had swapped
knobs with a non-working 435 until I realized that it has the marker where it belongs, just above the
outside circumference of the scale. The two 435a's are identical except one is a 1415A prefix with the
old style range indicator and the other is a 1629A prefix with the interchangeable scales. Apparently
the difference is other than what I've been able to detect without getting inside. Maybe when I get more
curious I'll do an inside examination.

 
The following users thanked this post: TimYTTT

Offline johnk0gcj@gmail.com

  • Contributor
  • Posts: 12
  • Country: us
Re: Difference between HP / Agilent 435A and 435B Power Meters
« Reply #2 on: July 22, 2024, 06:19:29 pm »

After giving this a little thought I remembered from years ago that the 435's select the correct
indication by determining which power sensor is connected. If I recall correctly it uses 2 or 3
pins and corresponding wires in the cable to indicate which sensor. The interchangeable scales
remove the need to do the math.
 
 
The following users thanked this post: TimYTTT

Offline metrologist

  • Super Contributor
  • ***
  • Posts: 2234
  • Country: 00
 
The following users thanked this post: TimYTTT

Offline johnk0gcj@gmail.com

  • Contributor
  • Posts: 12
  • Country: us
Re: Difference between HP / Agilent 435A and 435B Power Meters
« Reply #4 on: July 24, 2024, 02:40:18 am »

Thanks a lot. Now you've goaded me into doing some more
research to find the difference between the A and B.

 

Offline TimYTTTTopic starter

  • Contributor
  • Posts: 15
  • Country: ch
Re: Difference between HP / Agilent 435A and 435B Power Meters
« Reply #5 on: July 24, 2024, 03:10:33 pm »
First thanks to all of you for your inputs.
In the meantime I did some comparison based on the HP product leaflets for both meters. The differences that I detected are not significant:
1. Frequency Range
A: 100 kHz to 18 GHz (sensor dependent)
B: 100 kHz to 26.5 GHz (sensor dependent)
Conclusion: Anyhow, difference between A and B, being sensor dependent, is questionable.

2. Response Time (0 to 99% of reading)
A:
2 seconds on 3 μW range
0.75 second on 10 μW range
0.40 second on 30 μW range and
100 milliseconds on all other ranges
B:
Range 1 (most sensistive range) 
      <10.0 seconds
Range 2   <3.8 seconds
Range 3   <1.3 seconds
Range 4 to 10   <500 milliseconds
Conclusion: For me not so relevant difference

Enclosed is the document with the feature comparison. I do not understand why HP introduced the B variant of the Power meter. Maybe just the commercial reasons?


« Last Edit: July 24, 2024, 08:53:43 pm by TimYTTT »
 

Offline metrologist

  • Super Contributor
  • ***
  • Posts: 2234
  • Country: 00
Re: Difference between HP / Agilent 435A and 435B Power Meters
« Reply #6 on: July 24, 2024, 05:47:41 pm »
The reason for the B model I highly expect is the painful process of chasing gigahertz. The 8485A is the only supported sensor to 26.5 GHz
 
The following users thanked this post: TimYTTT

Offline johnk0gcj@gmail.com

  • Contributor
  • Posts: 12
  • Country: us
Re: Difference between HP / Agilent 435A and 435B Power Meters
« Reply #7 on: July 25, 2024, 02:19:39 am »

That sounds logical.
 

Offline TimYTTTTopic starter

  • Contributor
  • Posts: 15
  • Country: ch
Re: Difference between HP / Agilent 435A and 435B Power Meters
« Reply #8 on: July 27, 2024, 09:24:24 am »
--------------------------------------
metrologist: The reason for the B model I highly expect is the painful process of chasing gigahertz. The 8485A is the only supported sensor to 26.5 GHz
--------------------------------------
You are right. I went thru the HP-Catalogs from 1980 to 1997 and found following info:

435A was supported till 1980 and discontinued 1981.
 Frequency range: 0.1 MHz to 26.5 GHz

435B was introduced 1981 and discontinued in 1998.
Frequency ranges:
-0.1 MHz - 26.5 GHz  (from 1981 to 1985)
-0.1 MHz - 50 GHz (from 1986 to 1992)
-0.1 MHz - 110 GHz (from 1993 to 1997)
The range changes are probably related to new sensors and not to the 435B itself.


« Last Edit: July 27, 2024, 09:07:38 pm by TimYTTT »
 