But I think this alone proves that the unit is defective or what do you think?
I find that hard to say. Let us remind us again what the R&S specs are:
Voltage ripple and noise (20 Hz to 20 MHz):
- typ. < 1.5 mV (RMS)
- < 20 mV (peak-to-peak) (meas.)
Firstly, like noted before, good noise and ripple measurements are difficult to make. Potential issues are:
- Induced signals resulting in common mode noise (the reason some are suggesting differential probes to address this)
- Mismatching loads resulting in reflections and ringing; note the PSU output impedance is in the mOhm range (milli), and that of the scope is typically 1 MOhms (Mega) (there are impedance matching solutions to address this)
- Ensuring measurements are under different voltage and load conditions that those used for the specifications (unfortunately often not provided by the manufacturer),
- Impact of temperature
- Possible impact of other devices used in the measurement chain. For instance, an electronic load may be cause of problems (but may be the only practical way to measure a PSU a specific series of voltages and loads). Or, if a wire-wound resistor is used as load, its parasitic inductance may have an effect
Your new measurement setup is probably better than your first one, bt may not address all above issues.
Having that said, lets now look at your Picoscope screenshots. I do not know that instrument well, but the screenshots you shared seem to show only peak-to-peak measurements, not RMS measurements. That limits us in answering your question. I also cannot see whether the bandwidth of the measurement was limited to 20MHz (which is the way the specs are specified). Especially because there are switching elements in this PSU, this bandwidth may really make a difference. The peak to peak measurements on your screenshot seem to be close to the specs, though.
Yesterday, I did some more ripple and noise measurements on my NGE103B (and acknowledging these are probably far from perfect for reasons mentioned above). My results are roughly speaking in line with those
posted by Fixpoint a short while ago, and those
posted by YetAnotherTechie in this threat. Depending on the voltage and load, my measurements are close to the manufacturer’s specifications: not much better, not much worse either.
Despite the possible imperfections, let me share my measurements:
First measurement uses a 19Ohm resistor as a load, PSU set at 10V, and a resulting 0.54A current flowing. I do this to be able to see whether the electronic load, used for the following measurements, makes a lot of difference. PP and RMS results are close to R&S specifications. (See pictures Ripple 1 and Ripple 4)
The second measurement is basically the same, but now the 19Ohm load is an electronic load. The waveform is not identical (may have several reasons) but the resulting PP and RMS results are again close to R&S specifications. Using an electronic load instead of a resistor does not seem to affect findings a lot (see picture Ripple 2). (Note that Dave, in a video above in the threat, finds that the electronic load is a source of measurement problems in a similar setting.)
Thirdly a measurement at full load (10V, 3A, which is the maximum for that voltage because of the 33W per channel). Now we go above R&S specifications by a factor of 2-3, but again, I cannot be confident the measurement is perfect so cannot claim lack of compliance. (see picture Ripple 3)
In comparison, the PDVS2mini in battery mode shows 1.6mV p-p noise.
Yes, I see that is lower than the noise you measure yourself from your NGE. But it may not be a fair comparison. The NGE is a power supply, designed to significant current over a wide range of voltages. To do so in a compact housing R&S chose to use a switching regulator (offering efficiency and compactness) followed by a linear control circuitry (for stabilisation). Having switching elements inside (like many modern designs do) inevitably will result in some noise on the outputs. Comparing that to a device solely design to be a close-to-perfect voltage source may not be fair.
IMHO, the bottom line is that the NGE100 series may not be one's first choice when shopping for a low ripple PSU, its specs in this regards are certainly not stellar. But whether a specific unit meets those specs is not that easy to measure, and the measurements of NGEs we have seen up to now seem to suggest they are close to or meeting the specs, despite probably measurement imperfections.